We have Siemens Magnetom AVANTO and I have a problem with storage of MR spectroscopy. I found some info here, but without any effect.
I have SIEMENSCSANONIMAGE 1.3.12.2.1107.5.9.1 sop and MRStorageSpectroscopy 1.2.840.10008.5.1.4.1.1.4.2 sop in my dgatesop.lst, but if I try to send spectroscopy series, scanner told me "No transferable object(s) found for PATIENT" and Conquest Archive do nothing.
I look to log and use gdb, but I have no message on it.
Do you have any idea to fix it?
Thanx Milan
SIEMENS Spectroscopy is not accepted
-
-
Hi,
if there is really nothing in the log for your described attempt, Siemens itself thinks it can't send it. I would therefore look in the configuration at the Siemens side.
Marcel
-
I have also Siemens Avanto and after adding sops in dgatesop.lst Conquest receives and transmits raw spectro data without problems.
Besides mentioned SOP's i have also:
RawDataStorage 1.2.840.10008.5.1.4.1.1.66 sopCheck this out and test this - but my CQ server accept this after that modifications - I double check tommorow again, but two weeks ago I had tested this and worked without problem.
Regards
Bart -
Thank you bart. It works now, only one warning in pacs_trouble.log - incostitent modelity SR//MR, but spectroscopy works.
Milan -
Hi,
this warning is normal (data with another modality is added to a study).
Marcel