To All,
I'm currently attempting to get files from our new scanner transferring through a conquest server but am having little luck. We have an older 'Allegra' scanner from Siemens that works fantastically without a hitch however we are installing a new 'Skyra' machine which is having trouble contacting the conquest server.
To test I've also setup a K-PACS workstation on my desktop with a few files from both the Allegra and Skyra systems. Allegra files transfer without a problem, however I get the following attempting to transfer Skyra files:
The above is all we get out of KPacs with the process log set to debug and 'log dicom network connections'.
The conquest server (at debuglevel set to 10) doesn't even indicate that it's been communicated with, however a TCPDump of the servers network ports indicated it is indeed communicating. I'm currently running the RC2 version of conquest, I can't get the RC3 to compile presently. It's running on a CentOS 5.x 64-bit server.
The skyra workstation does report an error when transferring stating the 'dicom service request is not acceptable from network node', this error is not seen on the alegra or with the skyra connecting to our ancient CTN server (which we are trying to replace with conquest) so we aren't sure what exactly is causing this response.
I have been told that the skyra uses a new private Siemens header, not the one used by the previous Allegra models but I'm not sure why this would matter at all. I'd be happy to provide any other information or a few files for comparison if it will help.