Hi Marcel!
Im'having this issue with Conquest 1.4.15 when receiving study from ultrasound machine.
Tried also after reinstalling original configuration from GUI
what could be?
Winxp SP2 (o SP3), Kaspersky deactivated
Regards,
Ciro
Hi Marcel!
Im'having this issue with Conquest 1.4.15 when receiving study from ultrasound machine.
Tried also after reinstalling original configuration from GUI
what could be?
Winxp SP2 (o SP3), Kaspersky deactivated
Regards,
Ciro
Hi,
can you give me the debug log of the error?
Marcel
Tomorrow I can give You the report!
But, if I put: SQLHost, SQLServer, Username and Password =" in dicom.ini
should I have the same problem? I don't need conquest DB because I'm using my own db, only images
Let me know...
regards,
ciro
No,
no DB, no issue with duplicate UID's.
Marcel
dear Marcel, without SQL username, pass, ....
the result is this:
[CONQUESTSRV1] User interface test: local server is running!
[CONQUESTSRV1]
[CONQUESTSRV1] UPACS THREAD 4: STARTED AT: Wed Jan 12 16:35:16 2011
[CONQUESTSRV1] Calling Application Title : "ECO "
[CONQUESTSRV1] Called Application Title : "ALPHA7 "
[CONQUESTSRV1] Application Context : "1.2.840.10008.3.1.1.1", PDU length: 27740
[CONQUESTSRV1] Presentation Context 0 "1.2.840.10008.1.1" 1
[CONQUESTSRV1] Presentation Context 1 "1.2.840.10008.5.1.4.1.1.6.1" 1
[CONQUESTSRV1] Presentation Context 2 "1.2.840.10008.5.1.4.1.1.3.1" 1
[CONQUESTSRV1] [recompress]: recompressed with mode = un (strip=0)
[CONQUESTSRV1] ***Error writing file: Z:\GESTIONE\IMG_ECO\P000041\OS20110112_000124\unknown_TESTDICO\1_2_1294846517.v2
[CONQUESTSRV1] UPACS THREAD 4: ENDED AT: Wed Jan 12 16:35:17 2011
[CONQUESTSRV1] UPACS THREAD 4: TOTAL RUNNING TIME: 1 SECONDS
no writing of the file. I have 1 US machine, 1 PC with conquest but the images are stored on another pc (server). on the pc with conquest Z: is the server shared folder of images.
I'm in trouble....I dont'know how to go on...
Hi,
can you manually write that file from the server? This suggest an issue with rights.
Marcel
I think so, but the strange thing that using (instead of export converters):
ImportConverters = 2
ImportConverter0 = save to D:\SOFTWARE\ECOGYN\IMG_ECO\P000041\EG20110112_000124\temp.dcm
ImportConverter1 = process with dgate.exe "--convert_to_jpg:D:\SOFTWARE\ECOGYN\IMG_ECO\P000041\EG20110112_000124\temp.dcm,1024,D:\SOFTWARE\ECOGYN\IMG_ECO\P000041\EG20110112_000124\test.jpg,//0";destroy
it works...but now I've the following problems:
- the name of the jpg should be random (like %b for export conv.) and not fixed one, tried with %o but is too long to manage for me...a short one could be ok. Can I use %V0020,0013.jpg,//0";destroy to have the name of the jpg only with the image number?
- if a line of dicom.ini is too long (like this) it goes carriage return and doesn't work, how can I put a command on 2 lines in dicom.ini???
- regarding export converters, I think that installing conquest on the server should solve the problem but... I have 3 US machines and 3 pc with conquest, all the images should arrive on a 4th pc (server). And don't work with export conv., sure I can put 3 conquest installations on the server in different folders, but should it work? and for the services?
let me know, thanks a lot...
ciro
Hi,
I assume that the process (service ?) has no rights to write to drive Z:. For a service you should use an UNC path.
You can install as many servers as you want one one machine, and also as service.
For the name of the JPG you can use a substring of %o, e..g, %o[30,40]
Use an editor which allows longer lines (notepad without word wrap on).
Marcel
using
%o, e..g, %o[30,40]
is sure that jpgs have different names? and using %V0020,0013.jpg????
For a service you should use an UNC path
do you mean a path starting like \\ ????
Hi,
the last characters of the UID are unique - experiment.
UNC is indeed \\server\share\folder
Marcel
using %V0020,0013.jpg is ok because with %o[30,40] the lenght is not equal.
But......if I use service, it crashes (win xp sp2), If I use GUI it works. I need it to work with service.
another thing: it works for 2 aloka's US but not for GE E8 US??? it receives the dcm file but no conversion in jpg is made (and no error returned!) why?
let me know....
thanks
ciro
Hi,
Services do not know about drive mappings, and the service account has less rights. This is the only difference. You can try set the user for the service (from windows) to yourself.
No clue about the other US machine enable debug logging (at high level) and inspect the log to see what goes wrong.
Marcel
ok, but if I use the GUI, can I sistematically run conquest in the tray?
because users sometimes forget to start conquest and the images are not stored....
From my EMR I can start conquest but I cannot set in tray mode, only maximized or minimized, is there a .ini file in which set conquest to run in tray???
let me know...
cirollo
Hi,
try this. In your shortcut start it with:
c:\dicomsercer\conquestdicomserver.exe /hide
Marcel
"No clue about the other US machine enable debug logging (at high level) and inspect the log to see what goes wrong."
tomorrow I'm again there with Voluson E8, how can I enable debug logging?
When I get the log I'll post here
let me know...
regards
cirollo
Server status page - click debug log - click up arrow next to it 4 times.
Marcel
this is the debug log
[CONQUESTSRV1] User interface test: local server is running!
[CONQUESTSRV1] set debug level from GUI
[CONQUESTSRV1] set debug log from GUI
[CONQUESTSRV1] set debug level from GUI
[CONQUESTSRV1] Connected by address: 0100007f
[CONQUESTSRV1] Testing transfer: '1.2.840.10008.1.2' against list #0 = '1.2.840.10008.1.2'
[CONQUESTSRV1] 0000,0002 18 UI AffectedSOPClassUID "1.2.840.10008.1.1"
[CONQUESTSRV1] 0000,0100 2 US CommandField 48
[CONQUESTSRV1] 0000,0110 2 US MessageID 7
[CONQUESTSRV1] 0000,0800 2 US DataSetType 257
[CONQUESTSRV1] 9999,0300 24 UN "set debug level from GUI"
[CONQUESTSRV1] 9999,0400 12 UN "debuglevel:2"
[CONQUESTSRV1] set debug level from GUI
[CONQUESTSRV1] Connected by address: 0100007f
[CONQUESTSRV1] Testing transfer: '1.2.840.10008.1.2' against list #0 = '1.2.840.10008.1.2'
[CONQUESTSRV1] 0000,0002 18 UI AffectedSOPClassUID "1.2.840.10008.1.1"
[CONQUESTSRV1] 0000,0100 2 US CommandField 48
[CONQUESTSRV1] 0000,0110 2 US MessageID 7
[CONQUESTSRV1] 0000,0800 2 US DataSetType 257
[CONQUESTSRV1] 9999,0300 24 UN "set debug level from GUI"
[CONQUESTSRV1] 9999,0400 12 UN "debuglevel:3"
[CONQUESTSRV1] set debug level from GUI
[CONQUESTSRV1] Connected by address: 0100007f
[CONQUESTSRV1] Testing transfer: '1.2.840.10008.1.2' against list #0 = '1.2.840.10008.1.2'
[CONQUESTSRV1] 0000,0002 18 UI AffectedSOPClassUID "1.2.840.10008.1.1"
[CONQUESTSRV1] 0000,0100 2 US CommandField 48
[CONQUESTSRV1] 0000,0110 2 US MessageID 7
[CONQUESTSRV1] 0000,0800 2 US DataSetType 257
[CONQUESTSRV1] 9999,0300 24 UN "set debug level from GUI"
[CONQUESTSRV1] 9999,0400 12 UN "debuglevel:4"
[CONQUESTSRV1] set debug level from GUI
[CONQUESTSRV1] Connected by address: 0100007f
[CONQUESTSRV1] Testing transfer: '1.2.840.10008.1.2' against list #0 = '1.2.840.10008.1.2'
[CONQUESTSRV1] 0000,0002 18 UI AffectedSOPClassUID "1.2.840.10008.1.1"
[CONQUESTSRV1] 0000,0100 2 US CommandField 48
[CONQUESTSRV1] 0000,0110 2 US MessageID 7
[CONQUESTSRV1] 0000,0800 2 US DataSetType 257
[CONQUESTSRV1] 9999,0300 24 UN "set debug level from GUI"
[CONQUESTSRV1] 9999,0400 12 UN "debuglevel:3"
[CONQUESTSRV1] set debug level from GUI
[CONQUESTSRV1] Connected by address: 0100007f
[CONQUESTSRV1] Testing transfer: '1.2.840.10008.1.2' against list #0 = '1.2.840.10008.1.2'
[CONQUESTSRV1] 0000,0002 18 UI AffectedSOPClassUID "1.2.840.10008.1.1"
[CONQUESTSRV1] 0000,0100 2 US CommandField 48
[CONQUESTSRV1] 0000,0110 2 US MessageID 7
[CONQUESTSRV1] 0000,0800 2 US DataSetType 257
[CONQUESTSRV1] 9999,0300 24 UN "set debug level from GUI"
[CONQUESTSRV1] 9999,0400 12 UN "debuglevel:4"
[CONQUESTSRV1] set debug level from GUI
[CONQUESTSRV1] Connected by address: 3b01a8c0
[CONQUESTSRV1] Testing transfer: '1.2.840.10008.1.2.1' against list #0 = '1.2.840.10008.1.2'
[CONQUESTSRV1] Testing transfer: '1.2.840.10008.1.2.1' against list #1 = '1.2.840.10008.1.2.4.50'
[CONQUESTSRV1] Testing transfer: '1.2.840.10008.1.2.1' against list #2 = '1.2.840.10008.1.2.4.51'
[CONQUESTSRV1] Testing transfer: '1.2.840.10008.1.2.1' against list #3 = '1.2.840.10008.1.2.4.53'
[CONQUESTSRV1] Testing transfer: '1.2.840.10008.1.2.1' against list #4 = '1.2.840.10008.1.2.4.55'
[CONQUESTSRV1] Testing transfer: '1.2.840.10008.1.2.1' against list #5 = '1.2.840.10008.1.2.4.57'
[CONQUESTSRV1] Testing transfer: '1.2.840.10008.1.2.1' against list #6 = '1.2.840.10008.1.2.4.70'
[CONQUESTSRV1] Testing transfer: '1.2.840.10008.1.2' against list #0 = '1.2.840.10008.1.2'
[CONQUESTSRV1] Testing transfer: '1.2.840.10008.1.2.4.50' against list #0 = '1.2.840.10008.1.2'
[CONQUESTSRV1] Testing transfer: '1.2.840.10008.1.2.4.50' against list #1 = '1.2.840.10008.1.2.4.50'
[CONQUESTSRV1] Testing transfer: '1.2.840.10008.1.2.1' against list #0 = '1.2.840.10008.1.2'
[CONQUESTSRV1] Testing transfer: '1.2.840.10008.1.2.1' against list #1 = '1.2.840.10008.1.2.4.50'
[CONQUESTSRV1] Testing transfer: '1.2.840.10008.1.2.1' against list #2 = '1.2.840.10008.1.2.4.51'
[CONQUESTSRV1] Testing transfer: '1.2.840.10008.1.2.1' against list #3 = '1.2.840.10008.1.2.4.53'
[CONQUESTSRV1] Testing transfer: '1.2.840.10008.1.2.1' against list #4 = '1.2.840.10008.1.2.4.55'
[CONQUESTSRV1] Testing transfer: '1.2.840.10008.1.2.1' against list #5 = '1.2.840.10008.1.2.4.57'
[CONQUESTSRV1] Testing transfer: '1.2.840.10008.1.2.1' against list #6 = '1.2.840.10008.1.2.4.70'
[CONQUESTSRV1] Testing transfer: '1.2.840.10008.1.2' against list #0 = '1.2.840.10008.1.2'
[CONQUESTSRV1] Testing transfer: '1.2.840.10008.1.2.1' against list #0 = '1.2.840.10008.1.2'
[CONQUESTSRV1] Testing transfer: '1.2.840.10008.1.2.1' against list #1 = '1.2.840.10008.1.2.4.50'
[CONQUESTSRV1] Testing transfer: '1.2.840.10008.1.2.1' against list #2 = '1.2.840.10008.1.2.4.51'
[CONQUESTSRV1] Testing transfer: '1.2.840.10008.1.2.1' against list #3 = '1.2.840.10008.1.2.4.53'
[CONQUESTSRV1] Testing transfer: '1.2.840.10008.1.2.1' against list #4 = '1.2.840.10008.1.2.4.55'
[CONQUESTSRV1] Testing transfer: '1.2.840.10008.1.2.1' against list #5 = '1.2.840.10008.1.2.4.57'
[CONQUESTSRV1] Testing transfer: '1.2.840.10008.1.2.1' against list #6 = '1.2.840.10008.1.2.4.70'
[CONQUESTSRV1] Testing transfer: '1.2.840.10008.1.2' against list #0 = '1.2.840.10008.1.2'
[CONQUESTSRV1] Testing transfer: '1.2.840.10008.1.2.1' against list #0 = '1.2.840.10008.1.2'
[CONQUESTSRV1] Testing transfer: '1.2.840.10008.1.2.1' against list #1 = '1.2.840.10008.1.2.4.50'
[CONQUESTSRV1] Testing transfer: '1.2.840.10008.1.2.1' against list #2 = '1.2.840.10008.1.2.4.51'
[CONQUESTSRV1] Testing transfer: '1.2.840.10008.1.2.1' against list #3 = '1.2.840.10008.1.2.4.53'
[CONQUESTSRV1] Testing transfer: '1.2.840.10008.1.2.1' against list #4 = '1.2.840.10008.1.2.4.55'
[CONQUESTSRV1] Testing transfer: '1.2.840.10008.1.2.1' against list #5 = '1.2.840.10008.1.2.4.57'
[CONQUESTSRV1] Testing transfer: '1.2.840.10008.1.2.1' against list #6 = '1.2.840.10008.1.2.4.70'
[CONQUESTSRV1] Testing transfer: '1.2.840.10008.1.2' against list #0 = '1.2.840.10008.1.2'
[CONQUESTSRV1]
[CONQUESTSRV1] UPACS THREAD 0: STARTED AT: Wed Jan 19 16:06:31 2011
[CONQUESTSRV1] A-ASSOCIATE-RQ Packet Dump
[CONQUESTSRV1] Calling Application Title : "Voluson "
[CONQUESTSRV1] Called Application Title : "VOLUSON "
[CONQUESTSRV1] Application Context : "1.2.840.10008.3.1.1.1", PDU length: 28672
[CONQUESTSRV1] Number of Proposed Presentation Contexts: 5
[CONQUESTSRV1] Presentation Context 0 "1.2.840.10008.5.1.4.1.1.6.1" 1
[CONQUESTSRV1] Presentation Context 1 "1.2.840.10008.5.1.4.1.1.3.1" 1
[CONQUESTSRV1] Presentation Context 2 "1.2.840.10008.1.1" 1
[CONQUESTSRV1] Presentation Context 3 "1.2.840.10008.5.1.4.1.1.6" 1
[CONQUESTSRV1] Presentation Context 4 "1.2.840.10008.5.1.4.1.1.7" 1
[CONQUESTSRV1] Server Command := 0001
[CONQUESTSRV1] Message ID := 0000
[CONQUESTSRV1] 0000,0002 28 UI AffectedSOPClassUID "1.2.840.10008.5.1.4.1.1.6.1"
[CONQUESTSRV1] 0000,0100 2 US CommandField 1
[CONQUESTSRV1] 0000,0110 2 US MessageID 0
[CONQUESTSRV1] 0000,0700 2 US Priority 0
[CONQUESTSRV1] 0000,0800 2 US DataSetType 0
[CONQUESTSRV1] 0000,1000 50 UI AffectedSOPInstanceU "1.2.276.0.26.1.1.1.2.2011.50.55490.573239.12451840"
[CONQUESTSRV1] Importconverter0.0 executes: save to Z:\IMG_ECO\P000041\TT000116\temp.dcm
[CONQUESTSRV1] Importconverter1.0 executes: process with dgate.exe "--convert_to_jpg:Z:\IMG_ECO\P000041\TT000116\temp.dcm,1024,Z:\IMG_ECO\P000041\TT000116\2.jpg,//0" Z:\IMG_ECO\P000041\TT000116\printer_files\1.2.826.0.1.3680043.2.135.730956.43877812.6.1295449592.109.0.dcm
[CONQUESTSRV1] Connected by address: 0100007f
[CONQUESTSRV1] Testing transfer: '1.2.840.10008.1.2' against list #0 = '1.2.840.10008.1.2'
[CONQUESTSRV1] 0000,0002 18 UI AffectedSOPClassUID "1.2.840.10008.1.1"
[CONQUESTSRV1] 0000,0100 2 US CommandField 48
[CONQUESTSRV1] 0000,0110 2 US MessageID 1
[CONQUESTSRV1] 0000,0800 2 US DataSetType 257
[CONQUESTSRV1] 9999,0300 42 UN "Server command sent using DGATE -- option"
[CONQUESTSRV1] 9999,0400 94 UN "convert_to_jpg:Z:\\IMG_ECO\\P000041\\TT000116\\temp.dcm,1024,Z:\\IMG_ECO\\P000041\\TT000116\\2.jpg,//0"
[CONQUESTSRV1] Server command sent using DGATE -- option
NO JPG WAS CREATED!!!
Does the following command work from the command line?
dgate.exe "--convert_to_jpg:Z:\IMG_ECO\P000041\TT000116\temp.dcm,1024,Z:\IMG_ECO\P000041\TT000116\2.jpg,//0"
And what is the "Z:\IMG_ECO\P000041\TT000116\printer_files\1.2.826.0.1.3680043.2.135.730956.43877812.6.1295449592.109.0.dcm" doing there as second argument?
I assume this is a 1.4.16 version. Did you try the much simpler:
ImportConverter1 = save jpg to Z:\IMG_ECO\P000041\TT000116\temp.dcm,1024,Z:\IMG_ECO\P000041\TT000116\2.jpg
Marcel
Dear Marcel, I'm getting mad with this thing:
this is my dicom.ini and after you'll find the server response....
# This file contains configuration information for the DICOM server
# Do not edit unless you know what you are doing
[sscscp]
MicroPACS = sscscp
Edition = Personal
# Network configuration: server name and TCP/IP port#
MyACRNema = VOLUSON
TCPPort = 104
# Reference to other files: known dicom servers; database layout; sops
ACRNemaMap = acrnema.map
kFactorFile = dicom.sql
SOPClassList = dgatesop.lst
# Host(ignored), name, username and password for ODBC data source
SQLHost =
SQLServer =
Username =
Password =
DoubleBackSlashToDB = 0
# Configure database
TruncateFieldNames = 10
MaxFieldLength = 254
MaxFileNameLength = 255
FixPhilips = 0
FixKodak = 0
KeepAlive = 0
LargeFileSizeKB = 1024
ZipTime = 05:
UIDPrefix = 1.2.826.0.1.3680043.2.135.730956.43877812
EnableReadAheadThread = 1
PatientQuerySortOrder =
StudyQuerySortOrder =
SeriesQuerySortOrder =
ImageQuerySortOrder =
IndexDBF = 1
PackDBF = 0
LongQueryDBF = 1000
TCPIPTimeOut = 300
FailHoldOff = 60
RetryDelay = 100
QueueSize = 128
WorkListMode = 0
DebugLevel = 0
Prefetcher = 0
LRUSort =
AllowTruncate =
DecompressNon16BitsJpeg = 1
UseBuiltInDecompressor = 1
FileNameSyntax = 0
# Configuration of compression for incoming images and archival
DroppedFileCompression = un
IncomingCompression = un
ArchiveCompression = as
# Names of the database tables
PatientTableName = DICOMPatients
StudyTableName = DICOMStudies
SeriesTableName = DICOMSeries
ImageTableName = DICOMImages
DMarkTableName = DICOMAccessUpdates
RegisteredMOPDeviceTable = RegisteredMOPIDs
UIDToMOPIDTable = UIDToMOPID
UIDToCDRIDTable = UIDToCDRID
# Banner and host for debug information
PACSName = CONQUESTSRV1
OperatorConsole = 127.0.0.1
# Configuration of disk(s) to store images
MAGDeviceThreshhold = 0
MAGDevices = 1
MAGDevice0 = Z:\IMG_ECO\P000041\EG000198\
NightlyCleanThreshhold = 0
ImportConverters = 2
ImportConverter0 = save to Z:\IMG_ECO\P000041\EG000198\temp.dcm
ImportConverter1 = save jpg to Z:\IMG_ECO\P000041\EG000198\temp.dcm,1024,Z:\IMG_ECO\P000041\EG000198\TEMP.jpg
Server Response:
[CONQUESTSRV1] User interface test: local server is running!
[CONQUESTSRV1]
[CONQUESTSRV1] UPACS THREAD 1: STARTED AT: Tue Jan 25 18:34:12 2011
[CONQUESTSRV1] Calling Application Title : "Voluson "
[CONQUESTSRV1] Called Application Title : "VOLUSON "
[CONQUESTSRV1] Application Context : "1.2.840.10008.3.1.1.1", PDU length: 28672
[CONQUESTSRV1] Presentation Context 0 "1.2.840.10008.5.1.4.1.1.6.1" 1
[CONQUESTSRV1] Presentation Context 1 "1.2.840.10008.5.1.4.1.1.3.1" 1
[CONQUESTSRV1] Presentation Context 2 "1.2.840.10008.1.1" 1
[CONQUESTSRV1] Presentation Context 3 "1.2.840.10008.5.1.4.1.1.6" 1
[CONQUESTSRV1] Presentation Context 4 "1.2.840.10008.5.1.4.1.1.7" 1
[CONQUESTSRV1] Importconverter0.0 executes: save to Z:\IMG_ECO\P000041\EG000198\temp.dcm
[CONQUESTSRV1] ***Error writing file: Z:\IMG_ECO\P000041\EG000198\spirito^_D03105-1\1_20_1295976852.v2
[CONQUESTSRV1] UPACS THREAD 1: ENDED AT: Tue Jan 25 18:34:12 2011
[CONQUESTSRV1] UPACS THREAD 1: TOTAL RUNNING TIME: 0 SECONDS
[CONQUESTSRV1]
[CONQUESTSRV1] UPACS THREAD 2: STARTED AT: Tue Jan 25 18:34:34 2011
[CONQUESTSRV1] Calling Application Title : "Voluson "
[CONQUESTSRV1] Called Application Title : "VOLUSON "
[CONQUESTSRV1] Application Context : "1.2.840.10008.3.1.1.1", PDU length: 28672
[CONQUESTSRV1] Presentation Context 0 "1.2.840.10008.5.1.4.1.1.6.1" 1
[CONQUESTSRV1] Presentation Context 1 "1.2.840.10008.5.1.4.1.1.3.1" 1
[CONQUESTSRV1] Presentation Context 2 "1.2.840.10008.1.1" 1
[CONQUESTSRV1] Presentation Context 3 "1.2.840.10008.5.1.4.1.1.6" 1
[CONQUESTSRV1] Presentation Context 4 "1.2.840.10008.5.1.4.1.1.7" 1
[CONQUESTSRV1] Importconverter0.0 executes: save to Z:\IMG_ECO\P000041\EG000198\temp.dcm
[CONQUESTSRV1] ***Error writing file: Z:\IMG_ECO\P000041\EG000198\spirito^_D03105-1\1_18_1295976875.v2
[CONQUESTSRV1] UPACS THREAD 2: ENDED AT: Tue Jan 25 18:34:35 2011
[CONQUESTSRV1] UPACS THREAD 2: TOTAL RUNNING TIME: 1 SECONDS
no jpg is created!!!!
if I leave the import converters
in dicom.ini:
ImportConverters = 2
ImportConverter0 = save to Z:\IMG_ECO\P000041\EG000200\temp.dcm
ImportConverter1 = process with dgate.exe "--convert_to_jpg:Z:\IMG_ECO\P000041\EG000200\temp.dcm,1024,Z:\IMG_ECO\P000041\EG000200\%V0020,0013.jpg,//0";destroy
this is the resut:
[CONQUESTSRV1] User interface test: local server is running!
[CONQUESTSRV1]
[CONQUESTSRV1] UPACS THREAD 0: STARTED AT: Tue Jan 25 18:37:20 2011
[CONQUESTSRV1] Calling Application Title : "Voluson "
[CONQUESTSRV1] Called Application Title : "VOLUSON "
[CONQUESTSRV1] Application Context : "1.2.840.10008.3.1.1.1", PDU length: 28672
[CONQUESTSRV1] Presentation Context 0 "1.2.840.10008.5.1.4.1.1.6.1" 1
[CONQUESTSRV1] Presentation Context 1 "1.2.840.10008.5.1.4.1.1.3.1" 1
[CONQUESTSRV1] Presentation Context 2 "1.2.840.10008.1.1" 1
[CONQUESTSRV1] Presentation Context 3 "1.2.840.10008.5.1.4.1.1.6" 1
[CONQUESTSRV1] Presentation Context 4 "1.2.840.10008.5.1.4.1.1.7" 1
[CONQUESTSRV1] Importconverter0.0 executes: save to Z:\IMG_ECO\P000041\EG000200\temp.dcm
[CONQUESTSRV1] Importconverter1.0 executes: process with dgate.exe "--convert_to_jpg:Z:\IMG_ECO\P000041\EG000200\temp.dcm,1024,Z:\IMG_ECO\P000041\EG000200\21.jpg,//0" Z:\IMG_ECO\P000041\EG000200\printer_files\1.2.826.0.1.3680043.2.135.730956.43877812.6.1295977040.546.0.dcm
[CONQUESTSRV1] Server command sent using DGATE -- option
what the hell is:
[CONQUESTSRV1] Importconverter1.0 executes: process with dgate.exe "--convert_to_jpg:Z:\IMG_ECO\P000041\EG000200\temp.dcm,1024,Z:\IMG_ECO\P000041\EG000200\21.jpg,//0" Z:\IMG_ECO\P000041\EG000200\printer_files\1.2.826.0.1.3680043.2.135.730956.43877812.6.1295977040.546.0.dcm
it corresponds to destoy??? and the dgate crashes!!!
Please answer my question first. Does this work from the command line:
dgate.exe "--convert_to_jpg:Z:\IMG_ECO\P000041\TT000116\temp.dcm,1024,Z:\IMG_ECO\P000041\TT000116\2.jpg,//0"
This line was a mistake on my side:
ImportConverter1 = save jpg to Z:\IMG_ECO\P000041\TT000116\temp.dcm,1024,Z:\IMG_ECO\P000041\TT000116\2.jpg
should have been the simple:
ImportConverter1 = save jpg to Z:\IMG_ECO\P000041\TT000116\2.jpg
can dgate actually write to Z: at all? Is sounds like a drive mapping that is not available to a service (use unc paths).
The command "process with" writes the dicom object as temp file and starts the command passing the temp file name. Use "system" instead or leave the process with out. All these command are documented in the manual.
Marcel