Query/Retrieve Problems with GUI on new Query node

  • Hi Marcel,


    008,0052 is "QueryRetrieveLevel" so sorry, that was my fault. I had

    a.QueryRetrieveLevel='PATIENT' set.


    I have changed this again to:

    a.QueryRetrieveLevel='STUDY'.


    So this is my output with STUDY level selected:


    Code
    27/04/2021 17:21:35 [RTPHYSICS] Server command sent using DGATE -- option
    27/04/2021 17:21:35 [RTPHYSICS] Server command sent using DGATE -- option
    27/04/2021 17:21:37 [RTPHYSICS] 0381930 XXXX^^Mr. 20210319 MRI PELVIS PROSTATE 1.3.51.0.1.1.10.1.236.46.13056692.3599568 U13056692
    27/04/2021 17:21:37 [RTPHYSICS] Start Move
    27/04/2021 17:21:37 [RTPHYSICS] ***No valid presentation contexts/transfer syntax found in 0 candidates
    27/04/2021 17:21:37 [RTPHYSICS] ***In 0 presentation contexts
    27/04/2021 17:21:37 [RTPHYSICS] ***#Possible transfer syntaxes: 1
    27/04/2021 17:21:37 [RTPHYSICS] ***dicommove: remote DICOM error
    27/04/2021 17:21:37 [RTPHYSICS] *** multiplex: connection terminated
    27/04/2021 17:21:37 [RTPHYSICS] End Move


    Using DCMTK software, STUDY level and the same UID I can successfully query and retrieve study.


    Let me know if you can think of any other steps I can try.


    Thanks for your help,


    John

  • Hi Marcel,


    If you are referring to the "Enable Explicit and JPEG Transfers" switch then yes that option was always checked.


    The images that were retrieved by DCMTK seemed to be a standard DICOM format?


    Thanks,


    John

  • Code
    ***No valid presentation contexts/transfer syntax found in 0 candidates

    This message means that the PACS tries to send something to conquest that it does not understand.


    This can be the transfer syntax or the object. Enable jpeg compression (enabling transfers in dgatesop.lst, can you check) allows conquest to accept more.


    Marcel

  • Hi Marcel,


    I think I have enabled jpeg compression. I have attached screengrab of config tab and also the dgatesop.lst file (changed to .txt file).


    I have also tried uncommenting out all parameters in dgatesop.lst file, restarted server and still no luck.


    Let me know if I need to enable jpeg compression anywhere else.


    Thanks,


    John

  • Hi Marcel,


    Another interesting thing I've just noticed is that when I retrieve images with DCMTK and then try and drag and drop (which usually works fine for any DICOM I've seen), it will not import images and also gives no error in logs.


    The images, MR images in this case will open fine with ImageJ. Would you like me to try and send you the anonymized text header?


    John

  • Apologies Marcel, with further testing it looks like it was failing the drag and drop import due to being run as admin (which I need to keep it running as service). I discovered this the hard way but spotted this here: Drag and Drop not working Conquest


    So PACS Retrieve function still not working but images ae importing through drag-drop no problem.


    I will email PACS again and just verify that images are sent as a dicom format? Would that be the best question to ask them?


    Apologies for all the questions, learning lots here so thanks!


    John

  • Hi,


    if i want a C-Move to work in the Conquest GUI i have to double click on the "Series number" to switch to "Study UID" and do a query first.

    If i do not do that a move wont work because Conquest probably does not ask for a study by it's studyuid in the C-Move.


    Regards,

    Robert

  • Hi Robert,


    Thanks very much for your suggestion, I didn't know that about the GUI change by double-clicking.


    I have tried that but to no avail. I have also manually used ZeroBraneStudio to query with the seriesUID seperatly so no good either.


    I'm still waiting on a rely from PAC's so hopefully that will throw up something obvious.


    John

  • Hi Robert,


    Thanks for your suggestion but unfortunately I have no access to PAC's admin myself so I'm at their mercy to send on applicable screenshots. They have said they will also try and get back to me re:any compression being applied.


    Re: AE,IP,port: I am confident they are all correct because I can use DCMTK software to query and retrieve successfully on that same machine.


    Also images that come across look to be standard dicom after that DCMTK retrieve.


    I'm open to any other suggestions though, hopefully I'm missing something obvious!


    John

  • Hi All,


    So just to update this morning, Agfa have some back and said:


    "For accession number - U13056692 we send formats:


    MR and SR


    MR - SOP is fully supported by EI


    SR - SOP class is fully supported by EI


    When I've checked the SOP classes for this accession I have noticed there is a rejection note here - SOP class '1.2.840.10008.5.1.4.1.1.88.59' - This could be the reason the external system is not accepting the images.


    Please can you ask John to check if that SOP class is fully supported in the external system?"


    Is this making any sense to you Marcel and is "SOP" class fully supported by ConQuest? Sorry if these are basic/obvious answers!


    Thanks,


    John

  • Hi,


    i can Q/R from AGFA Enterprise Imaging, check my settings:



    i do the Q/R from conquest like this:


    then i double click a studyuid returned from the query, and click copy to destination (conquest AE title).

    i'f i don't switch to the UID setting in conquest the move will not work.


    p.s. i think the rejection note is accepted by conquest it's called KeyObjectSelectionDocumentStorage in the dgatesop.lst file in conquest (at least it's in my file, don't remember if i added this manually a long time ago).


    regards,

    Robert