Hi! I have then next problem with Conquest 1.4.19b server: When CT device send study with multiple images at server (server compress on the fly using jped2000 loseless method) in the end of transfer in log server writes the next error OpenJ2:"Input data out of ranges". How decide this problem?
Conquest Server Error in log: "Input data out of ranges"
-
-
Can you post an excerpt of the relevant log?
Marcel
-
Quote from marcelvanherk
Can you post an excerpt of the relevant log?
Marcel
Yes, of course.
[CONQUESTSRV1] FreeStore Left 65659 on c:\
[CONQUESTSRV1] Written file: c:\dicomserver\data\13159\1.2.840.113704.1.111.6764.1521194381.7_0002_000298_1523824749012d.dcm
[CONQUESTSRV1] Server Command := 0001
[CONQUESTSRV1] Message ID := 012f
[CONQUESTSRV1] [recompress]: recompressed with mode = js (strip=0)
[CONQUESTSRV1] FreeStore Left 65659 on c:\
[CONQUESTSRV1] Written file: c:\dicomserver\data\13159\1.2.840.113704.1.111.6764.1521194381.7_0002_000299_1523824749012e.dcm
[CONQUESTSRV1] Server Command := 0001
[CONQUESTSRV1] Message ID := 0130
[CONQUESTSRV1] [recompress]: recompressed with mode = js (strip=0)
[CONQUESTSRV1] FreeStore Left 65659 on c:\
[CONQUESTSRV1] Written file: c:\dicomserver\data\13159\1.2.840.113704.1.111.6764.1521194381.7_0002_000300_1523824749012f.dcm
[CONQUESTSRV1] Server Command := 0001
[CONQUESTSRV1] Message ID := 0131
[CONQUESTSRV1] ***[CompressJPEGLS]: Error: input data out of range.
[CONQUESTSRV1] FreeStore Left 65658 on c:\
[CONQUESTSRV1] Written file: c:\dicomserver\data\13159\1.2.840.113704.7.1.1.5448.1521194516.304_80148_000001_15238247490130.dcm
[CONQUESTSRV1] UPACS THREAD 16: ENDED AT: Sun Apr 15 23:39:09 2018
[CONQUESTSRV1] UPACS THREAD 16: TOTAL RUNNING TIME: 11 SECONDS -
Aha,
I believe this happens when the DICOM header says the image is 12 bits, while more than 12 bits are stored in the imafe. In that case the image is not compressed. Can you check that this is the case and that the stored image is still correct (yet uncompressed).
This test was added in 1.4.19b because otherwise the pixel data would be corrupted.
Marcel
-
Quote from marcelvanherk
Aha,
I believe this happens when the DICOM header says the image is 12 bits, while more than 12 bits are stored in the imafe. In that case the image is not compressed. Can you check that this is the case and that the stored image is still correct (yet uncompressed).
This test was added in 1.4.19b because otherwise the pixel data would be corrupted.
Marcel
I found this image. it is image which contains dose info and the stored image is correct.But in hospital when Philips CT Brilliance 16 slice send CT study to conquest pacs server - server shutdown. In control panel Philips CT appear "Out of resources". At server pacs option IgnoreOutOfMemoryErrors = 1. I want to decide this problem, but i don't know how.
-
Hi,
the second question looks like a serious bug. Can you send the log where this happens, and if it is repeatable also a high level debug log?
Marcel
-
Quote from marcelvanherk
Hi,
the second question looks like a serious bug. Can you send the log where this happens, and if it is repeatable also a high level debug log?
Marcel
Good day, Marcel!
Thank you very much for your replies. In this log no anything info about shutdown conquest pacs service. But this happened every time after receiving some images of study from CT Philips.
LOG:
20180402 15:19:20 DGATE (1.4.19b, build Thu Oct 12 18:42:54 2017, bits 64) is running as threaded server
20180402 15:19:20 Database type: ODBC connection
02.04.2018 15:19:25 [GAUZBGB1PACS] Stopped zip and cleanup thread
02.04.2018 15:19:25 [GAUZBGB1PACS] User interface test: local server is running!
02.04.2018 15:19:25 [GAUZBGB1PACS] set debug level from GUI
02.04.2018 15:19:25 [GAUZBGB1PACS] set normal log from GUI
02.04.2018 15:19:29 [GAUZBGB1PACS] set debug level from GUI
02.04.2018 15:19:29 [GAUZBGB1PACS] set debug log from GUI
02.04.2018 15:19:34 [GAUZBGB1PACS]
02.04.2018 15:19:34 [GAUZBGB1PACS] UPACS THREAD 6: STARTED AT: Mon Apr 02 15:19:34 2018
02.04.2018 15:19:34 [GAUZBGB1PACS] A-ASSOCIATE-RQ Packet Dump
02.04.2018 15:19:34 [GAUZBGB1PACS] Calling Application Title : "MXV "
02.04.2018 15:19:34 [GAUZBGB1PACS] Called Application Title : "GAUZBGB1PACS "
02.04.2018 15:19:34 [GAUZBGB1PACS] Application Context : "1.2.840.10008.3.1.1.1", PDU length: 16384
02.04.2018 15:19:34 [GAUZBGB1PACS] Number of Proposed Presentation Contexts: 43
02.04.2018 15:19:34 [GAUZBGB1PACS] Presentation Context 0 "1.2.840.10008.5.1.4.1.1.1" 1
02.04.2018 15:19:34 [GAUZBGB1PACS] Presentation Context 1 "1.2.840.10008.5.1.4.1.1.2" 1
02.04.2018 15:19:34 [GAUZBGB1PACS] Presentation Context 2 "1.2.840.10008.5.1.4.1.1.3.1" 1
02.04.2018 15:19:34 [GAUZBGB1PACS] Presentation Context 3 "1.2.840.10008.5.1.4.1.1.4" 1
02.04.2018 15:19:34 [GAUZBGB1PACS] Presentation Context 4 "1.2.840.10008.5.1.4.1.1.6.1" 1
02.04.2018 15:19:34 [GAUZBGB1PACS] Presentation Context 5 "1.2.840.10008.5.1.4.1.1.7" 1
02.04.2018 15:19:34 [GAUZBGB1PACS] Presentation Context 6 "1.2.840.10008.5.1.4.1.1.8" 1
02.04.2018 15:19:34 [GAUZBGB1PACS] Presentation Context 7 "1.2.840.10008.5.1.4.1.1.9" 1
02.04.2018 15:19:34 [GAUZBGB1PACS] Presentation Context 8 "1.2.840.10008.5.1.4.1.1.9.1.2" 1
02.04.2018 15:19:34 [GAUZBGB1PACS] Presentation Context 9 "1.2.840.10008.5.1.4.1.1.11.1" 1
02.04.2018 15:19:34 [GAUZBGB1PACS] Presentation Context 10 "1.2.840.10008.5.1.4.1.1.10" 1
02.04.2018 15:19:34 [GAUZBGB1PACS] Presentation Context 11 "1.2.840.10008.5.1.4.1.1.11" 1
02.04.2018 15:19:34 [GAUZBGB1PACS] Presentation Context 12 "1.2.840.10008.5.1.4.1.1.12.1" 1
02.04.2018 15:19:34 [GAUZBGB1PACS] Presentation Context 13 "1.2.840.10008.5.1.4.1.1.12.2" 1
02.04.2018 15:19:34 [GAUZBGB1PACS] Presentation Context 14 "1.2.840.10008.5.1.4.1.1.20" 1
02.04.2018 15:19:34 [GAUZBGB1PACS] Presentation Context 15 "1.2.840.10008.5.1.4.1.1.77.1.1" 1
02.04.2018 15:19:34 [GAUZBGB1PACS] Presentation Context 16 "1.2.840.10008.5.1.4.1.1.77.1.2" 1
02.04.2018 15:19:34 [GAUZBGB1PACS] Presentation Context 17 "1.2.840.10008.5.1.4.1.1.77.1.3" 1
02.04.2018 15:19:34 [GAUZBGB1PACS] Presentation Context 18 "1.2.840.10008.5.1.4.1.1.77.1.4" 1
02.04.2018 15:19:34 [GAUZBGB1PACS] Presentation Context 19 "1.2.840.10008.5.1.4.1.1.128" 1
02.04.2018 15:19:34 [GAUZBGB1PACS] Presentation Context 20 "1.2.840.10008.5.1.4.1.1.129" 1
02.04.2018 15:19:34 [GAUZBGB1PACS] Presentation Context 21 "1.2.840.10008.5.1.4.1.1.481.1" 1
02.04.2018 15:19:34 [GAUZBGB1PACS] Presentation Context 22 "1.2.840.10008.5.1.4.1.1.481.2" 1
02.04.2018 15:19:34 [GAUZBGB1PACS] Presentation Context 23 "1.2.840.10008.5.1.4.1.1.481.3" 1
02.04.2018 15:19:34 [GAUZBGB1PACS] Presentation Context 24 "1.2.840.10008.5.1.4.1.1.481.4" 1
02.04.2018 15:19:34 [GAUZBGB1PACS] Presentation Context 25 "1.2.840.10008.5.1.4.1.1.481.5" 1
02.04.2018 15:19:34 [GAUZBGB1PACS] Presentation Context 26 "1.2.840.10008.5.1.4.1.1.481.6" 1
02.04.2018 15:19:34 [GAUZBGB1PACS] Presentation Context 27 "1.2.840.10008.5.1.4.1.1.481.7" 1
02.04.2018 15:19:34 [GAUZBGB1PACS] Presentation Context 28 "1.2.840.10008.5.1.4.1.1.1.1" 1
02.04.2018 15:19:34 [GAUZBGB1PACS] Presentation Context 29 "1.2.840.10008.5.1.4.1.1.1.1.1" 1
02.04.2018 15:19:34 [GAUZBGB1PACS] Presentation Context 30 "1.2.840.10008.5.1.4.1.1.1.2" 1
02.04.2018 15:19:34 [GAUZBGB1PACS] Presentation Context 31 "1.2.840.10008.5.1.4.1.1.1.1.2.1" 0
02.04.2018 15:19:34 [GAUZBGB1PACS] Presentation Context 32 "1.2.840.10008.5.1.4.1.1.1.3" 1
02.04.2018 15:19:35 [GAUZBGB1PACS] Presentation Context 33 "1.2.840.10008.5.1.4.1.1.1.1.3.1" 0
02.04.2018 15:19:35 [GAUZBGB1PACS] Presentation Context 34 "1.2.840.113704.7.0.1.9" 0
02.04.2018 15:19:35 [GAUZBGB1PACS] Presentation Context 35 "1.2.840.113704.7.0.5.1" 0
02.04.2018 15:19:35 [GAUZBGB1PACS] Presentation Context 36 "1.3.46.670589.5.0.1.1" 0
02.04.2018 15:19:35 [GAUZBGB1PACS] Presentation Context 37 "1.3.46.670589.5.0.2.1" 0
02.04.2018 15:19:35 [GAUZBGB1PACS] Presentation Context 38 "1.3.46.670589.5.0.3.1" 0
02.04.2018 15:19:35 [GAUZBGB1PACS] Presentation Context 39 "1.3.46.670589.5.0.10" 0
02.04.2018 15:19:35 [GAUZBGB1PACS] Presentation Context 40 "1.3.46.670589.5.0.9" 0
02.04.2018 15:19:35 [GAUZBGB1PACS] Presentation Context 41 "1.3.46.670589.2.4.1.1" 0
02.04.2018 15:19:35 [GAUZBGB1PACS] Presentation Context 42 "1.3.46.670589.5.0.12" 0
02.04.2018 15:19:35 [GAUZBGB1PACS] Server Command := 0001
02.04.2018 15:19:35 [GAUZBGB1PACS] Message ID := 0001
02.04.2018 15:19:35 [GAUZBGB1PACS] Move Originator Message ID := 0012
02.04.2018 15:19:35 [GAUZBGB1PACS] Move Originator AE := MXV
02.04.2018 15:19:35 [GAUZBGB1PACS] [recompress]: recompressed with mode = jk (strip=0)
02.04.2018 15:19:35 [GAUZBGB1PACS] Rewritten file: d:\distribution\data\431812\1.3.6.1.4.1.30071.8.176156457628161.5759843185196041\1.2.840.113704.1.111.4384.1522669604.7\1.2.840.113704.1.111.5796.1522669659.214.dcm
02.04.2018 15:19:35 [GAUZBGB1PACS] Server Command := 0001
02.04.2018 15:19:35 [GAUZBGB1PACS] Message ID := 0002
02.04.2018 15:19:35 [GAUZBGB1PACS] Move Originator Message ID := 0012
02.04.2018 15:19:35 [GAUZBGB1PACS] Move Originator AE := MXV
02.04.2018 15:19:35 [GAUZBGB1PACS] [recompress]: recompressed with mode = jk (strip=0)
02.04.2018 15:19:35 [GAUZBGB1PACS] Rewritten file: d:\distribution\data\431812\1.3.6.1.4.1.30071.8.176156457628161.5759843185196041\1.2.840.113704.1.111.4384.1522669604.7\1.2.840.113704.1.111.5796.1522669659.215.dcm
02.04.2018 15:19:35 [GAUZBGB1PACS] Server Command := 0001
02.04.2018 15:19:35 [GAUZBGB1PACS] Message ID := 0003
02.04.2018 15:19:35 [GAUZBGB1PACS] Move Originator Message ID := 0012
02.04.2018 15:19:35 [GAUZBGB1PACS] Move Originator AE := MXV
02.04.2018 15:19:35 [GAUZBGB1PACS] [recompress]: recompressed with mode = jk (strip=0)
02.04.2018 15:19:35 [GAUZBGB1PACS] Rewritten file: d:\distribution\data\431812\1.3.6.1.4.1.30071.8.176156457628161.5759843185196041\1.2.840.113704.1.111.4384.1522669604.7\1.2.840.113704.1.111.5796.1522669659.216.dcm
02.04.2018 15:19:35 [GAUZBGB1PACS] Server Command := 0001
02.04.2018 15:19:35 [GAUZBGB1PACS] Message ID := 0004
02.04.2018 15:19:35 [GAUZBGB1PACS] Move Originator Message ID := 0012
02.04.2018 15:19:35 [GAUZBGB1PACS] Move Originator AE := MXV
02.04.2018 15:19:35 [GAUZBGB1PACS] [recompress]: recompressed with mode = jk (strip=0)
02.04.2018 15:19:35 [GAUZBGB1PACS] Rewritten file: d:\distribution\data\431812\1.3.6.1.4.1.30071.8.176156457628161.5759843185196041\1.2.840.113704.1.111.4384.1522669604.7\1.2.840.113704.1.111.5796.1522669660.217.dcm
02.04.2018 15:19:35 [GAUZBGB1PACS] Server Command := 0001
02.04.2018 15:19:35 [GAUZBGB1PACS] Message ID := 0005
02.04.2018 15:19:35 [GAUZBGB1PACS] Move Originator Message ID := 0012
02.04.2018 15:19:35 [GAUZBGB1PACS] Move Originator AE := MXV
02.04.2018 15:19:35 [GAUZBGB1PACS] [recompress]: recompressed with mode = jk (strip=0)
02.04.2018 15:20:40 Server has service: D:\distribution\dgateserv.exe /process dgate64.exe -^serverstatus.log
20180402 15:20:41 Started zip and cleanup thread
20180402 15:20:41 Monitoring for files in: d:\distribution\data\incoming\
20180402 15:20:41 DGATE (1.4.19b, build Thu Oct 12 18:42:54 2017, bits 64) is running as threaded server
20180402 15:20:41 Database type: ODBC connection
20180402 15:20:42 User interface test: local server is running!
02.04.2018 15:20:47 [GAUZBGB1PACS] Stopped zip and cleanup thread
02.04.2018 15:20:47 [GAUZBGB1PACS] User interface test: local server is running!
02.04.2018 15:20:49 [GAUZBGB1PACS] set debug level from GUI
02.04.2018 15:20:49 [GAUZBGB1PACS] set debug log from GUI
02.04.2018 15:20:51 [GAUZBGB1PACS] set debug level from GUI
02.04.2018 15:20:51 [GAUZBGB1PACS] set normal log from GUI
02.04.2018 15:20:54 [GAUZBGB1PACS] set debug level from GUI
02.04.2018 15:20:54 [GAUZBGB1PACS] set debug log from GUI
02.04.2018 15:20:58 [GAUZBGB1PACS]
02.04.2018 15:20:58 [GAUZBGB1PACS] UPACS THREAD 9: STARTED AT: Mon Apr 02 15:20:57 2018
02.04.2018 15:20:58 [GAUZBGB1PACS] A-ASSOCIATE-RQ Packet Dump
02.04.2018 15:20:58 [GAUZBGB1PACS] Calling Application Title : "MXV "
02.04.2018 15:20:58 [GAUZBGB1PACS] Called Application Title : "GAUZBGB1PACS "
02.04.2018 15:20:58 [GAUZBGB1PACS] Application Context : "1.2.840.10008.3.1.1.1", PDU length: 16384
02.04.2018 15:20:58 [GAUZBGB1PACS] Number of Proposed Presentation Contexts: 43
02.04.2018 15:20:58 [GAUZBGB1PACS] Presentation Context 0 "1.2.840.10008.5.1.4.1.1.1" 1
02.04.2018 15:20:58 [GAUZBGB1PACS] Presentation Context 1 "1.2.840.10008.5.1.4.1.1.2" 1
02.04.2018 15:20:58 [GAUZBGB1PACS] Presentation Context 2 "1.2.840.10008.5.1.4.1.1.3.1" 1
02.04.2018 15:20:58 [GAUZBGB1PACS] Presentation Context 3 "1.2.840.10008.5.1.4.1.1.4" 1
02.04.2018 15:20:58 [GAUZBGB1PACS] Presentation Context 4 "1.2.840.10008.5.1.4.1.1.6.1" 1
02.04.2018 15:20:58 [GAUZBGB1PACS] Presentation Context 5 "1.2.840.10008.5.1.4.1.1.7" 1
02.04.2018 15:20:58 [GAUZBGB1PACS] Presentation Context 6 "1.2.840.10008.5.1.4.1.1.8" 1
02.04.2018 15:20:58 [GAUZBGB1PACS] Presentation Context 7 "1.2.840.10008.5.1.4.1.1.9" 1
02.04.2018 15:20:58 [GAUZBGB1PACS] Presentation Context 8 "1.2.840.10008.5.1.4.1.1.9.1.2" 1
02.04.2018 15:20:58 [GAUZBGB1PACS] Presentation Context 9 "1.2.840.10008.5.1.4.1.1.11.1" 1
02.04.2018 15:20:58 [GAUZBGB1PACS] Presentation Context 10 "1.2.840.10008.5.1.4.1.1.10" 1
02.04.2018 15:20:58 [GAUZBGB1PACS] Presentation Context 11 "1.2.840.10008.5.1.4.1.1.11" 1
02.04.2018 15:20:58 [GAUZBGB1PACS] Presentation Context 12 "1.2.840.10008.5.1.4.1.1.12.1" 1
02.04.2018 15:20:58 [GAUZBGB1PACS] Presentation Context 13 "1.2.840.10008.5.1.4.1.1.12.2" 1
02.04.2018 15:20:58 [GAUZBGB1PACS] Presentation Context 14 "1.2.840.10008.5.1.4.1.1.20" 1
02.04.2018 15:20:58 [GAUZBGB1PACS] Presentation Context 15 "1.2.840.10008.5.1.4.1.1.77.1.1" 1
02.04.2018 15:20:58 [GAUZBGB1PACS] Presentation Context 16 "1.2.840.10008.5.1.4.1.1.77.1.2" 1
02.04.2018 15:20:58 [GAUZBGB1PACS] Presentation Context 17 "1.2.840.10008.5.1.4.1.1.77.1.3" 1
02.04.2018 15:20:58 [GAUZBGB1PACS] Presentation Context 18 "1.2.840.10008.5.1.4.1.1.77.1.4" 1
02.04.2018 15:20:58 [GAUZBGB1PACS] Presentation Context 19 "1.2.840.10008.5.1.4.1.1.128" 1
02.04.2018 15:20:58 [GAUZBGB1PACS] Presentation Context 20 "1.2.840.10008.5.1.4.1.1.129" 1
02.04.2018 15:20:58 [GAUZBGB1PACS] Presentation Context 21 "1.2.840.10008.5.1.4.1.1.481.1" 1
02.04.2018 15:20:58 [GAUZBGB1PACS] Presentation Context 22 "1.2.840.10008.5.1.4.1.1.481.2" 1
02.04.2018 15:20:58 [GAUZBGB1PACS] Presentation Context 23 "1.2.840.10008.5.1.4.1.1.481.3" 1
02.04.2018 15:20:58 [GAUZBGB1PACS] Presentation Context 24 "1.2.840.10008.5.1.4.1.1.481.4" 1
02.04.2018 15:20:58 [GAUZBGB1PACS] Presentation Context 25 "1.2.840.10008.5.1.4.1.1.481.5" 1
02.04.2018 15:20:58 [GAUZBGB1PACS] Presentation Context 26 "1.2.840.10008.5.1.4.1.1.481.6" 1
02.04.2018 15:20:58 [GAUZBGB1PACS] Presentation Context 27 "1.2.840.10008.5.1.4.1.1.481.7" 1
02.04.2018 15:20:58 [GAUZBGB1PACS] Presentation Context 28 "1.2.840.10008.5.1.4.1.1.1.1" 1
02.04.2018 15:20:58 [GAUZBGB1PACS] Presentation Context 29 "1.2.840.10008.5.1.4.1.1.1.1.1" 1
02.04.2018 15:20:58 [GAUZBGB1PACS] Presentation Context 30 "1.2.840.10008.5.1.4.1.1.1.2" 1
02.04.2018 15:20:58 [GAUZBGB1PACS] Presentation Context 31 "1.2.840.10008.5.1.4.1.1.1.1.2.1" 0
02.04.2018 15:20:58 [GAUZBGB1PACS] Presentation Context 32 "1.2.840.10008.5.1.4.1.1.1.3" 1
02.04.2018 15:20:58 [GAUZBGB1PACS] Presentation Context 33 "1.2.840.10008.5.1.4.1.1.1.1.3.1" 0
02.04.2018 15:20:58 [GAUZBGB1PACS] Presentation Context 34 "1.2.840.113704.7.0.1.9" 0
02.04.2018 15:20:58 [GAUZBGB1PACS] Presentation Context 35 "1.2.840.113704.7.0.5.1" 0
02.04.2018 15:20:58 [GAUZBGB1PACS] Presentation Context 36 "1.3.46.670589.5.0.1.1" 0
02.04.2018 15:20:58 [GAUZBGB1PACS] Presentation Context 37 "1.3.46.670589.5.0.2.1" 0
02.04.2018 15:20:58 [GAUZBGB1PACS] Presentation Context 38 "1.3.46.670589.5.0.3.1" 0
02.04.2018 15:20:58 [GAUZBGB1PACS] Presentation Context 39 "1.3.46.670589.5.0.10" 0
02.04.2018 15:20:58 [GAUZBGB1PACS] Presentation Context 40 "1.3.46.670589.5.0.9" 0
02.04.2018 15:20:58 [GAUZBGB1PACS] Presentation Context 41 "1.3.46.670589.2.4.1.1" 0
02.04.2018 15:20:58 [GAUZBGB1PACS] Presentation Context 42 "1.3.46.670589.5.0.12" 0
02.04.2018 15:20:58 [GAUZBGB1PACS] Server Command := 0001
02.04.2018 15:20:58 [GAUZBGB1PACS] Message ID := 0001
02.04.2018 15:20:58 [GAUZBGB1PACS] Move Originator Message ID := 0013
02.04.2018 15:20:58 [GAUZBGB1PACS] Move Originator AE := MXV
02.04.2018 15:20:58 [GAUZBGB1PACS] Rewritten file: d:\distribution\data\431812\1.3.6.1.4.1.30071.8.176156457628161.5759843185196041\1.2.840.113704.1.111.4384.1522669604.7\1.2.840.113704.1.111.5796.1522669659.214.dcm
02.04.2018 15:20:58 [GAUZBGB1PACS] Server Command := 0001
02.04.2018 15:20:58 [GAUZBGB1PACS] Message ID := 0002
02.04.2018 15:20:58 [GAUZBGB1PACS] Move Originator Message ID := 0013
02.04.2018 15:20:58 [GAUZBGB1PACS] Move Originator AE := MXV
02.04.2018 15:20:58 [GAUZBGB1PACS] Rewritten file: d:\distribution\data\431812\1.3.6.1.4.1.30071.8.176156457628161.5759843185196041\1.2.840.113704.1.111.4384.1522669604.7\1.2.840.113704.1.111.5796.1522669659.215.dcm
02.04.2018 15:20:58 [GAUZBGB1PACS] Server Command := 0001
02.04.2018 15:20:58 [GAUZBGB1PACS] Message ID := 0003
02.04.2018 15:20:58 [GAUZBGB1PACS] Move Originator Message ID := 0013
02.04.2018 15:20:58 [GAUZBGB1PACS] Move Originator AE := MXV
02.04.2018 15:20:58 [GAUZBGB1PACS] Rewritten file: d:\distribution\data\431812\1.3.6.1.4.1.30071.8.176156457628161.5759843185196041\1.2.840.113704.1.111.4384.1522669604.7\1.2.840.113704.1.111.5796.1522669659.216.dcm
02.04.2018 15:20:58 [GAUZBGB1PACS] Server Command := 0001
02.04.2018 15:20:58 [GAUZBGB1PACS] Message ID := 0004
02.04.2018 15:20:58 [GAUZBGB1PACS] Move Originator Message ID := 0013
02.04.2018 15:20:58 [GAUZBGB1PACS] Move Originator AE := MXV
02.04.2018 15:20:58 [GAUZBGB1PACS] Rewritten file: d:\distribution\data\431812\1.3.6.1.4.1.30071.8.176156457628161.5759843185196041\1.2.840.113704.1.111.4384.1522669604.7\1.2.840.113704.1.111.5796.1522669660.217.dcm
02.04.2018 15:20:58 [GAUZBGB1PACS] Server Command := 0001
02.04.2018 15:20:58 [GAUZBGB1PACS] Message ID := 0005
02.04.2018 15:20:58 [GAUZBGB1PACS] Move Originator Message ID := 0013
02.04.2018 15:20:58 [GAUZBGB1PACS] Move Originator AE := MXV
02.04.2018 15:20:58 [GAUZBGB1PACS] Rewritten file: d:\distribution\data\431812\1.3.6.1.4.1.30071.8.176156457628161.5759843185196041\1.2.840.113704.1.111.4384.1522669604.7\1.2.840.113704.1.111.5796.1522669660.218.dcm
02.04.2018 15:20:58 [GAUZBGB1PACS] Server Command := 0001
02.04.2018 15:20:58 [GAUZBGB1PACS] Message ID := 0006
02.04.2018 15:20:58 [GAUZBGB1PACS] Move Originator Message ID := 0013
02.04.2018 15:20:58 [GAUZBGB1PACS] Move Originator AE := MXV
02.04.2018 15:20:58 [GAUZBGB1PACS] Rewritten file: d:\distribution\data\431812\1.3.6.1.4.1.30071.8.176156457628161.5759843185196041\1.2.840.113704.1.111.4384.1522669604.7\1.2.840.113704.1.111.5796.1522669660.219.dcm
02.04.2018 15:20:58 [GAUZBGB1PACS] Server Command := 0001
02.04.2018 15:20:58 [GAUZBGB1PACS] Message ID := 0007
02.04.2018 15:20:58 [GAUZBGB1PACS] Move Originator Message ID := 0013
02.04.2018 15:20:58 [GAUZBGB1PACS] Move Originator AE := MXV
02.04.2018 15:20:58 [GAUZBGB1PACS] Rewritten file: d:\distribution\data\431812\1.3.6.1.4.1.30071.8.176156457628161.5759843185196041\1.2.840.113704.1.111.4384.1522669604.7\1.2.840.113704.1.111.5796.1522669661.220.dcm
02.04.2018 15:20:58 [GAUZBGB1PACS] Server Command := 0001
02.04.2018 15:20:58 [GAUZBGB1PACS] Message ID := 0008
02.04.2018 15:20:58 [GAUZBGB1PACS] Move Originator Message ID := 0013
02.04.2018 15:20:58 [GAUZBGB1PACS] Move Originator AE := MXV
02.04.2018 15:20:58 [GAUZBGB1PACS] Rewritten file: d:\distribution\data\431812\1.3.6.1.4.1.30071.8.176156457628161.5759843185196041\1.2.840.113704.1.111.4384.1522669604.7\1.2.840.113704.1.111.5796.1522669661.221.dcm
02.04.2018 15:20:58 [GAUZBGB1PACS] Server Command := 0001
02.04.2018 15:20:58 [GAUZBGB1PACS] Message ID := 0009
02.04.2018 15:20:58 [GAUZBGB1PACS] Move Originator Message ID := 0013
02.04.2018 15:20:58 [GAUZBGB1PACS] Move Originator AE := MXV
02.04.2018 15:20:58 [GAUZBGB1PACS] Rewritten file: d:\distribution\data\431812\1.3.6.1.4.1.30071.8.176156457628161.5759843185196041\1.2.840.113704.1.111.4384.1522669604.7\1.2.840.113704.1.111.5796.1522669661.222.dcm
02.04.2018 15:20:58 [GAUZBGB1PACS] Server Command := 0001
02.04.2018 15:20:59 [GAUZBGB1PACS] Message ID := 000a
02.04.2018 15:20:59 [GAUZBGB1PACS] Move Originator Message ID := 0013
02.04.2018 15:20:59 [GAUZBGB1PACS] Move Originator AE := MXV
02.04.2018 15:20:59 [GAUZBGB1PACS] Rewritten file: d:\distribution\data\431812\1.3.6.1.4.1.30071.8.176156457628161.5759843185196041\1.2.840.113704.1.111.4384.1522669604.7\1.2.840.113704.1.111.5796.1522669662.223.dcm
02.04.2018 15:20:59 [GAUZBGB1PACS] Server Command := 0001
02.04.2018 15:20:59 [GAUZBGB1PACS] Message ID := 000b
02.04.2018 15:20:59 [GAUZBGB1PACS] Move Originator Message ID := 0013
02.04.2018 15:20:59 [GAUZBGB1PACS] Move Originator AE := MXV
02.04.2018 15:20:59 [GAUZBGB1PACS] Rewritten file: d:\distribution\data\431812\1.3.6.1.4.1.30071.8.176156457628161.5759843185196041\1.2.840.113704.1.111.4384.1522669604.7\1.2.840.113704.1.111.5796.1522669662.224.dcm
02.04.2018 15:20:59 [GAUZBGB1PACS] Server Command := 0001
02.04.2018 15:20:59 [GAUZBGB1PACS] Message ID := 000c
02.04.2018 15:20:59 [GAUZBGB1PACS] Move Originator Message ID := 0013
02.04.2018 15:20:59 [GAUZBGB1PACS] Move Originator AE := MXV
02.04.2018 15:20:59 [GAUZBGB1PACS] Rewritten file: d:\distribution\data\431812\1.3.6.1.4.1.30071.8.176156457628161.5759843185196041\1.2.840.113704.1.111.4384.1522669604.7\1.2.840.113704.1.111.5796.1522669662.225.dcm
02.04.2018 15:20:59 [GAUZBGB1PACS] Server Command := 0001
02.04.2018 15:20:59 [GAUZBGB1PACS] Message ID := 000d
02.04.2018 15:20:59 [GAUZBGB1PACS] Move Originator Message ID := 0013
02.04.2018 15:20:59 [GAUZBGB1PACS] Move Originator AE := MXV
02.04.2018 15:20:59 [GAUZBGB1PACS] Rewritten file: d:\distribution\data\431812\1.3.6.1.4.1.30071.8.176156457628161.5759843185196041\1.2.840.113704.1.111.4384.1522669604.7\1.2.840.113704.1.111.5796.1522669663.226.dcm
02.04.2018 15:20:59 [GAUZBGB1PACS] Server Command := 0001
02.04.2018 15:20:59 [GAUZBGB1PACS] Message ID := 000e
02.04.2018 15:20:59 [GAUZBGB1PACS] Move Originator Message ID := 0013
02.04.2018 15:20:59 [GAUZBGB1PACS] Move Originator AE := MXV
02.04.2018 15:20:59 [GAUZBGB1PACS] FreeStore Left 3593253 on d:\
02.04.2018 15:20:59 [GAUZBGB1PACS] Written file: d:\distribution\data\431812\1.3.6.1.4.1.30071.8.176156457628161.5759843185196041\1.2.840.113704.1.111.4384.1522669604.7\1.2.840.113704.1.111.5796.1522669663.227.dcm
02.04.2018 15:20:59 [GAUZBGB1PACS] Server Command := 0001
02.04.2018 15:20:59 [GAUZBGB1PACS] Message ID := 000f
02.04.2018 15:20:59 [GAUZBGB1PACS] Move Originator Message ID := 0013
02.04.2018 15:20:59 [GAUZBGB1PACS] Move Originator AE := MXV
02.04.2018 15:20:59 [GAUZBGB1PACS] FreeStore Left 3593253 on d:\
02.04.2018 15:20:59 [GAUZBGB1PACS] Written file: d:\distribution\data\431812\1.3.6.1.4.1.30071.8.176156457628161.5759843185196041\1.2.840.113704.1.111.4384.1522669604.7\1.2.840.113704.1.111.5796.1522669663.228.dcm
02.04.2018 15:20:59 [GAUZBGB1PACS] Server Command := 0001
02.04.2018 15:20:59 [GAUZBGB1PACS] Message ID := 0010
02.04.2018 15:20:59 [GAUZBGB1PACS] Move Originator Message ID := 0013
02.04.2018 15:20:59 [GAUZBGB1PACS] Move Originator AE := MXV
02.04.2018 15:20:59 [GAUZBGB1PACS] FreeStore Left 3593252 on d:\
02.04.2018 15:20:59 [GAUZBGB1PACS] Written file: d:\distribution\data\431812\1.3.6.1.4.1.30071.8.176156457628161.5759843185196041\1.2.840.113704.1.111.4384.1522669604.7\1.2.840.113704.1.111.5796.1522669664.229.dcm
02.04.2018 15:20:59 [GAUZBGB1PACS] Server Command := 0001
02.04.2018 15:20:59 [GAUZBGB1PACS] Message ID := 0011
02.04.2018 15:20:59 [GAUZBGB1PACS] Move Originator Message ID := 0013
02.04.2018 15:20:59 [GAUZBGB1PACS] Move Originator AE := MXV
02.04.2018 15:20:59 [GAUZBGB1PACS] FreeStore Left 3593252 on d:\
02.04.2018 15:20:59 [GAUZBGB1PACS] Written file: d:\distribution\data\431812\1.3.6.1.4.1.30071.8.176156457628161.5759843185196041\1.2.840.113704.1.111.4384.1522669604.7\1.2.840.113704.1.111.5796.1522669664.230.dcm
02.04.2018 15:20:59 [GAUZBGB1PACS] Server Command := 0001
02.04.2018 15:20:59 [GAUZBGB1PACS] Message ID := 0012
02.04.2018 15:20:59 [GAUZBGB1PACS] Move Originator Message ID := 0013
02.04.2018 15:20:59 [GAUZBGB1PACS] Move Originator AE := MXV
02.04.2018 15:20:59 [GAUZBGB1PACS] FreeStore Left 3593251 on d:\
02.04.2018 15:20:59 [GAUZBGB1PACS] Written file: d:\distribution\data\431812\1.3.6.1.4.1.30071.8.176156457628161.5759843185196041\1.2.840.113704.1.111.4384.1522669604.7\1.2.840.113704.1.111.5796.1522669664.231.dcm
02.04.2018 15:20:59 [GAUZBGB1PACS] Server Command := 0001
02.04.2018 15:20:59 [GAUZBGB1PACS] Message ID := 0013
02.04.2018 15:20:59 [GAUZBGB1PACS] Move Originator Message ID := 0013
02.04.2018 15:20:59 [GAUZBGB1PACS] Move Originator AE := MXV
02.04.2018 15:20:59 [GAUZBGB1PACS] FreeStore Left 3593251 on d:\
02.04.2018 15:20:59 [GAUZBGB1PACS] Written file: d:\distribution\data\431812\1.3.6.1.4.1.30071.8.176156457628161.5759843185196041\1.2.840.113704.1.111.4384.1522669604.7\1.2.840.113704.1.111.5796.1522669665.232.dcm
02.04.2018 15:20:59 [GAUZBGB1PACS] Server Command := 0001
02.04.2018 15:20:59 [GAUZBGB1PACS] Message ID := 0014
02.04.2018 15:20:59 [GAUZBGB1PACS] Move Originator Message ID := 0013
02.04.2018 15:20:59 [GAUZBGB1PACS] Move Originator AE := MXV
02.04.2018 15:20:59 [GAUZBGB1PACS] FreeStore Left 3593250 on d:\
02.04.2018 15:20:59 [GAUZBGB1PACS] Written file: d:\distribution\data\431812\1.3.6.1.4.1.30071.8.176156457628161.5759843185196041\1.2.840.113704.1.111.4384.1522669604.7\1.2.840.113704.1.111.5796.1522669665.233.dcm
02.04.2018 15:20:59 [GAUZBGB1PACS] Server Command := 0001
02.04.2018 15:20:59 [GAUZBGB1PACS] Message ID := 0015
02.04.2018 15:20:59 [GAUZBGB1PACS] Move Originator Message ID := 0013
02.04.2018 15:20:59 [GAUZBGB1PACS] Move Originator AE := MXV
02.04.2018 15:20:59 [GAUZBGB1PACS] FreeStore Left 3593250 on d:\
02.04.2018 15:20:59 [GAUZBGB1PACS] Written file: d:\distribution\data\431812\1.3.6.1.4.1.30071.8.176156457628161.5759843185196041\1.2.840.113704.1.111.4384.1522669604.7\1.2.840.113704.1.111.5796.1522669665.234.dcm
02.04.2018 15:20:59 [GAUZBGB1PACS] Server Command := 0001
02.04.2018 15:20:59 [GAUZBGB1PACS] Message ID := 0016
02.04.2018 15:20:59 [GAUZBGB1PACS] Move Originator Message ID := 0013
02.04.2018 15:20:59 [GAUZBGB1PACS] Move Originator AE := MXV
02.04.2018 15:20:59 [GAUZBGB1PACS] FreeStore Left 3593249 on d:\
02.04.2018 15:20:59 [GAUZBGB1PACS] Written file: d:\distribution\data\431812\1.3.6.1.4.1.30071.8.176156457628161.5759843185196041\1.2.840.113704.1.111.4384.1522669604.7\1.2.840.113704.1.111.5796.1522669666.235.dcm
02.04.2018 15:20:59 [GAUZBGB1PACS] Server Command := 0001
02.04.2018 15:20:59 [GAUZBGB1PACS] Message ID := 0017
02.04.2018 15:20:59 [GAUZBGB1PACS] Move Originator Message ID := 0013
02.04.2018 15:20:59 [GAUZBGB1PACS] Move Originator AE := MXV
02.04.2018 15:20:59 [GAUZBGB1PACS] FreeStore Left 3593248 on d:\
02.04.2018 15:20:59 [GAUZBGB1PACS] Written file: d:\distribution\data\431812\1.3.6.1.4.1.30071.8.176156457628161.5759843185196041\1.2.840.113704.1.111.4384.1522669604.7\1.2.840.113704.1.111.5796.1522669666.236.dcm
02.04.2018 15:20:59 [GAUZBGB1PACS] Server Command := 0001
02.04.2018 15:20:59 [GAUZBGB1PACS] Message ID := 0018
02.04.2018 15:20:59 [GAUZBGB1PACS] Move Originator Message ID := 0013
02.04.2018 15:20:59 [GAUZBGB1PACS] Move Originator AE := MXV
02.04.2018 15:20:59 [GAUZBGB1PACS] FreeStore Left 3593248 on d:\
02.04.2018 15:20:59 [GAUZBGB1PACS] Written file: d:\distribution\data\431812\1.3.6.1.4.1.30071.8.176156457628161.5759843185196041\1.2.840.113704.1.111.4384.1522669604.7\1.2.840.113704.1.111.5796.1522669666.237.dcm
02.04.2018 15:20:59 [GAUZBGB1PACS] Server Command := 0001
02.04.2018 15:20:59 [GAUZBGB1PACS] Message ID := 0019
02.04.2018 15:20:59 [GAUZBGB1PACS] Move Originator Message ID := 0013
02.04.2018 15:20:59 [GAUZBGB1PACS] Move Originator AE := MXV
02.04.2018 15:20:59 [GAUZBGB1PACS] FreeStore Left 3593247 on d:\
02.04.2018 15:20:59 [GAUZBGB1PACS] Written file: d:\distribution\data\431812\1.3.6.1.4.1.30071.8.176156457628161.5759843185196041\1.2.840.113704.1.111.4384.1522669604.7\1.2.840.113704.1.111.5796.1522669667.238.dcm
02.04.2018 15:20:59 [GAUZBGB1PACS] Server Command := 0001
02.04.2018 15:20:59 [GAUZBGB1PACS] Message ID := 001a
02.04.2018 15:21:00 [GAUZBGB1PACS] Move Originator Message ID := 0013
02.04.2018 15:21:00 [GAUZBGB1PACS] Move Originator AE := MXV
02.04.2018 15:21:00 [GAUZBGB1PACS] FreeStore Left 3593247 on d:\
02.04.2018 15:21:00 [GAUZBGB1PACS] Written file: d:\distribution\data\431812\1.3.6.1.4.1.30071.8.176156457628161.5759843185196041\1.2.840.113704.1.111.4384.1522669604.7\1.2.840.113704.1.111.5796.1522669667.239.dcm
02.04.2018 15:21:00 [GAUZBGB1PACS] Server Command := 0001
02.04.2018 15:21:00 [GAUZBGB1PACS] Message ID := 001b
02.04.2018 15:21:00 [GAUZBGB1PACS] Move Originator Message ID := 0013
02.04.2018 15:21:00 [GAUZBGB1PACS] Move Originator AE := MXV
02.04.2018 15:21:00 [GAUZBGB1PACS] FreeStore Left 3593246 on d:\
02.04.2018 15:21:00 [GAUZBGB1PACS] Written file: d:\distribution\data\431812\1.3.6.1.4.1.30071.8.176156457628161.5759843185196041\1.2.840.113704.1.111.4384.1522669604.7\1.2.840.113704.1.111.5796.1522669667.240.dcm
02.04.2018 15:21:00 [GAUZBGB1PACS] Server Command := 0001
02.04.2018 15:21:00 [GAUZBGB1PACS] Message ID := 001c
02.04.2018 15:21:00 [GAUZBGB1PACS] Move Originator Message ID := 0013
02.04.2018 15:21:00 [GAUZBGB1PACS] Move Originator AE := MXV
02.04.2018 15:21:00 [GAUZBGB1PACS] FreeStore Left 3593246 on d:\
02.04.2018 15:21:00 [GAUZBGB1PACS] Written file: d:\distribution\data\431812\1.3.6.1.4.1.30071.8.176156457628161.5759843185196041\1.2.840.113704.1.111.4384.1522669604.7\1.2.840.113704.1.111.5796.1522669668.241.dcm
02.04.2018 15:21:00 [GAUZBGB1PACS] Server Command := 0001
02.04.2018 15:21:00 [GAUZBGB1PACS] Message ID := 001d
02.04.2018 15:21:00 [GAUZBGB1PACS] Move Originator Message ID := 0013
02.04.2018 15:21:00 [GAUZBGB1PACS] Move Originator AE := MXV
02.04.2018 15:21:00 [GAUZBGB1PACS] FreeStore Left 3593245 on d:\
02.04.2018 15:21:00 [GAUZBGB1PACS] Written file: d:\distribution\data\431812\1.3.6.1.4.1.30071.8.176156457628161.5759843185196041\1.2.840.113704.1.111.4384.1522669604.7\1.2.840.113704.1.111.5796.1522669668.242.dcm
02.04.2018 15:21:00 [GAUZBGB1PACS] Server Command := 0001
02.04.2018 15:21:00 [GAUZBGB1PACS] Message ID := 001e
02.04.2018 15:21:00 [GAUZBGB1PACS] Move Originator Message ID := 0013
02.04.2018 15:21:00 [GAUZBGB1PACS] Move Originator AE := MXV
02.04.2018 15:21:00 [GAUZBGB1PACS] FreeStore Left 3593245 on d:\
02.04.2018 15:21:00 [GAUZBGB1PACS] Written file: d:\distribution\data\431812\1.3.6.1.4.1.30071.8.176156457628161.5759843185196041\1.2.840.113704.1.111.4384.1522669604.7\1.2.840.113704.1.111.5796.1522669668.243.dcm
02.04.2018 15:21:00 [GAUZBGB1PACS] Server Command := 0001
02.04.2018 15:21:00 [GAUZBGB1PACS] Message ID := 001f
02.04.2018 15:21:00 [GAUZBGB1PACS] Move Originator Message ID := 0013
02.04.2018 15:21:00 [GAUZBGB1PACS] Move Originator AE := MXV
02.04.2018 15:21:00 [GAUZBGB1PACS] FreeStore Left 3593244 on d:\
02.04.2018 15:21:00 [GAUZBGB1PACS] Written file: d:\distribution\data\431812\1.3.6.1.4.1.30071.8.176156457628161.5759843185196041\1.2.840.113704.1.111.4384.1522669604.7\1.2.840.113704.1.111.5796.1522669669.244.dcm
02.04.2018 15:21:00 [GAUZBGB1PACS] Server Command := 0001
02.04.2018 15:21:00 [GAUZBGB1PACS] Message ID := 0020
02.04.2018 15:21:00 [GAUZBGB1PACS] Move Originator Message ID := 0013
02.04.2018 15:21:00 [GAUZBGB1PACS] Move Originator AE := MXV
02.04.2018 15:21:00 [GAUZBGB1PACS] FreeStore Left 3593244 on d:\
02.04.2018 15:21:00 [GAUZBGB1PACS] Written file: d:\distribution\data\431812\1.3.6.1.4.1.30071.8.176156457628161.5759843185196041\1.2.840.113704.1.111.4384.1522669604.7\1.2.840.113704.1.111.5796.1522669669.245.dcm
02.04.2018 15:21:00 [GAUZBGB1PACS] Server Command := 0001
02.04.2018 15:21:00 [GAUZBGB1PACS] Message ID := 0021
02.04.2018 15:21:00 [GAUZBGB1PACS] Move Originator Message ID := 0013
02.04.2018 15:21:00 [GAUZBGB1PACS] Move Originator AE := MXV
02.04.2018 15:21:00 [GAUZBGB1PACS] FreeStore Left 3593243 on d:\
02.04.2018 15:21:00 [GAUZBGB1PACS] Written file: d:\distribution\data\431812\1.3.6.1.4.1.30071.8.176156457628161.5759843185196041\1.2.840.113704.1.111.4384.1522669604.7\1.2.840.113704.1.111.5796.1522669669.246.dcm
02.04.2018 15:21:00 [GAUZBGB1PACS] Server Command := 0001
02.04.2018 15:21:00 [GAUZBGB1PACS] Message ID := 0022
02.04.2018 15:21:00 [GAUZBGB1PACS] Move Originator Message ID := 0013
02.04.2018 15:21:00 [GAUZBGB1PACS] Move Originator AE := MXV
02.04.2018 15:21:00 [GAUZBGB1PACS] FreeStore Left 3593243 on d:\
02.04.2018 15:21:00 [GAUZBGB1PACS] Written file: d:\distribution\data\431812\1.3.6.1.4.1.30071.8.176156457628161.5759843185196041\1.2.840.113704.1.111.4384.1522669604.7\1.2.840.113704.1.111.5796.1522669670.247.dcm
02.04.2018 15:21:00 [GAUZBGB1PACS] Server Command := 0001
02.04.2018 15:21:00 [GAUZBGB1PACS] Message ID := 0023
02.04.2018 15:21:00 [GAUZBGB1PACS] Move Originator Message ID := 0013
02.04.2018 15:21:00 [GAUZBGB1PACS] Move Originator AE := MXV
02.04.2018 15:21:00 [GAUZBGB1PACS] FreeStore Left 3593242 on d:\
02.04.2018 15:21:00 [GAUZBGB1PACS] Written file: d:\distribution\data\431812\1.3.6.1.4.1.30071.8.176156457628161.5759843185196041\1.2.840.113704.1.111.4384.1522669604.7\1.2.840.113704.1.111.5796.1522669670.248.dcm
02.04.2018 15:21:00 [GAUZBGB1PACS] Server Command := 0001
02.04.2018 15:21:00 [GAUZBGB1PACS] Message ID := 0024
02.04.2018 15:21:00 [GAUZBGB1PACS] Move Originator Message ID := 0013
02.04.2018 15:21:00 [GAUZBGB1PACS] Move Originator AE := MXV
02.04.2018 15:21:00 [GAUZBGB1PACS] FreeStore Left 3593242 on d:\
02.04.2018 15:21:00 [GAUZBGB1PACS] Written file: d:\distribution\data\431812\1.3.6.1.4.1.30071.8.176156457628161.5759843185196041\1.2.840.113704.1.111.4384.1522669604.7\1.2.840.113704.1.111.5796.1522669670.249.dcm
02.04.2018 15:21:00 [GAUZBGB1PACS] Server Command := 0001
02.04.2018 15:21:00 [GAUZBGB1PACS] Message ID := 0025
02.04.2018 15:21:00 [GAUZBGB1PACS] Move Originator Message ID := 0013
02.04.2018 15:21:00 [GAUZBGB1PACS] Move Originator AE := MXV
02.04.2018 15:21:00 [GAUZBGB1PACS] FreeStore Left 3593241 on d:\
02.04.2018 15:21:00 [GAUZBGB1PACS] Written file: d:\distribution\data\431812\1.3.6.1.4.1.30071.8.176156457628161.5759843185196041\1.2.840.113704.1.111.4384.1522669604.7\1.2.840.113704.1.111.5796.1522669671.250.dcm
02.04.2018 15:21:01 [GAUZBGB1PACS] Server Command := 0001
02.04.2018 15:21:01 [GAUZBGB1PACS] Message ID := 0026
02.04.2018 15:21:01 [GAUZBGB1PACS] Move Originator Message ID := 0013
02.04.2018 15:21:01 [GAUZBGB1PACS] Move Originator AE := MXV
02.04.2018 15:21:01 [GAUZBGB1PACS] FreeStore Left 3593241 on d:\
02.04.2018 15:21:01 [GAUZBGB1PACS] Written file: d:\distribution\data\431812\1.3.6.1.4.1.30071.8.176156457628161.5759843185196041\1.2.840.113704.1.111.4384.1522669604.7\1.2.840.113704.1.111.5796.1522669671.251.dcm
02.04.2018 15:21:01 [GAUZBGB1PACS] Server Command := 0001
02.04.2018 15:21:01 [GAUZBGB1PACS] Message ID := 0027
02.04.2018 15:21:01 [GAUZBGB1PACS] Move Originator Message ID := 0013
02.04.2018 15:21:01 [GAUZBGB1PACS] Move Originator AE := MXV
02.04.2018 15:21:01 [GAUZBGB1PACS] FreeStore Left 3593240 on d:\
02.04.2018 15:21:01 [GAUZBGB1PACS] Written file: d:\distribution\data\431812\1.3.6.1.4.1.30071.8.176156457628161.5759843185196041\1.2.840.113704.1.111.4384.1522669604.7\1.2.840.113704.1.111.5796.1522669671.252.dcm
02.04.2018 15:21:01 [GAUZBGB1PACS] Server Command := 0001
02.04.2018 15:21:01 [GAUZBGB1PACS] Message ID := 0028
02.04.2018 15:21:01 [GAUZBGB1PACS] Move Originator Message ID := 0013
02.04.2018 15:21:01 [GAUZBGB1PACS] Move Originator AE := MXV
20180402 15:22:05 Started zip and cleanup thread
20180402 15:22:05 Monitoring for files in: d:\distribution\data\incoming\
20180402 15:22:05 DGATE (1.4.19b, build Thu Oct 12 18:42:54 2017, bits 64) is running as threaded server
20180402 15:22:05 Database type: ODBC connection
02.04.2018 15:22:11 [GAUZBGB1PACS] Stopped zip and cleanup thread
02.04.2018 15:22:11 [GAUZBGB1PACS] User interface test: local server is running!
02.04.2018 15:22:11 [GAUZBGB1PACS] set debug level from GUI
02.04.2018 15:22:11 [GAUZBGB1PACS] set normal log from GUI
02.04.2018 15:22:19 [GAUZBGB1PACS]
02.04.2018 15:22:19 [GAUZBGB1PACS] UPACS THREAD 4: STARTED AT: Mon Apr 02 15:22:19 2018
02.04.2018 15:22:19 [GAUZBGB1PACS] Calling Application Title : "MXV "
02.04.2018 15:22:19 [GAUZBGB1PACS] Called Application Title : "GAUZBGB1PACS "
02.04.2018 15:22:20 [GAUZBGB1PACS] Application Context : "1.2.840.10008.3.1.1.1", PDU length: 16384
02.04.2018 15:22:20 [GAUZBGB1PACS] Presentation Context 0 "1.2.840.10008.5.1.4.1.1.1" 1
02.04.2018 15:22:20 [GAUZBGB1PACS] Presentation Context 1 "1.2.840.10008.5.1.4.1.1.2" 1
02.04.2018 15:22:20 [GAUZBGB1PACS] Presentation Context 2 "1.2.840.10008.5.1.4.1.1.3.1" 1
02.04.2018 15:22:20 [GAUZBGB1PACS] Presentation Context 3 "1.2.840.10008.5.1.4.1.1.4" 1
02.04.2018 15:22:20 [GAUZBGB1PACS] Presentation Context 4 "1.2.840.10008.5.1.4.1.1.6.1" 1
02.04.2018 15:22:20 [GAUZBGB1PACS] Presentation Context 5 "1.2.840.10008.5.1.4.1.1.7" 1
02.04.2018 15:22:20 [GAUZBGB1PACS] Presentation Context 6 "1.2.840.10008.5.1.4.1.1.8" 1
02.04.2018 15:22:20 [GAUZBGB1PACS] Presentation Context 7 "1.2.840.10008.5.1.4.1.1.9" 1
02.04.2018 15:22:20 [GAUZBGB1PACS] Presentation Context 8 "1.2.840.10008.5.1.4.1.1.9.1.2" 1
02.04.2018 15:22:20 [GAUZBGB1PACS] Presentation Context 9 "1.2.840.10008.5.1.4.1.1.11.1" 1
02.04.2018 15:22:20 [GAUZBGB1PACS] Presentation Context 10 "1.2.840.10008.5.1.4.1.1.10" 1
02.04.2018 15:22:20 [GAUZBGB1PACS] Presentation Context 11 "1.2.840.10008.5.1.4.1.1.11" 1
02.04.2018 15:22:20 [GAUZBGB1PACS] Presentation Context 12 "1.2.840.10008.5.1.4.1.1.12.1" 1
02.04.2018 15:22:20 [GAUZBGB1PACS] Presentation Context 13 "1.2.840.10008.5.1.4.1.1.12.2" 1
02.04.2018 15:22:20 [GAUZBGB1PACS] Presentation Context 14 "1.2.840.10008.5.1.4.1.1.20" 1
02.04.2018 15:22:20 [GAUZBGB1PACS] Presentation Context 15 "1.2.840.10008.5.1.4.1.1.77.1.1" 1
02.04.2018 15:22:20 [GAUZBGB1PACS] Presentation Context 16 "1.2.840.10008.5.1.4.1.1.77.1.2" 1
02.04.2018 15:22:20 [GAUZBGB1PACS] Presentation Context 17 "1.2.840.10008.5.1.4.1.1.77.1.3" 1
02.04.2018 15:22:20 [GAUZBGB1PACS] Presentation Context 18 "1.2.840.10008.5.1.4.1.1.77.1.4" 1
02.04.2018 15:22:20 [GAUZBGB1PACS] Presentation Context 19 "1.2.840.10008.5.1.4.1.1.128" 1
02.04.2018 15:22:20 [GAUZBGB1PACS] Presentation Context 20 "1.2.840.10008.5.1.4.1.1.129" 1
02.04.2018 15:22:20 [GAUZBGB1PACS] Presentation Context 21 "1.2.840.10008.5.1.4.1.1.481.1" 1
02.04.2018 15:22:20 [GAUZBGB1PACS] Presentation Context 22 "1.2.840.10008.5.1.4.1.1.481.2" 1
02.04.2018 15:22:20 [GAUZBGB1PACS] Presentation Context 23 "1.2.840.10008.5.1.4.1.1.481.3" 1
02.04.2018 15:22:20 [GAUZBGB1PACS] Presentation Context 24 "1.2.840.10008.5.1.4.1.1.481.4" 1
02.04.2018 15:22:20 [GAUZBGB1PACS] Presentation Context 25 "1.2.840.10008.5.1.4.1.1.481.5" 1
02.04.2018 15:22:20 [GAUZBGB1PACS] Presentation Context 26 "1.2.840.10008.5.1.4.1.1.481.6" 1
02.04.2018 15:22:20 [GAUZBGB1PACS] Presentation Context 27 "1.2.840.10008.5.1.4.1.1.481.7" 1
02.04.2018 15:22:20 [GAUZBGB1PACS] Presentation Context 28 "1.2.840.10008.5.1.4.1.1.1.1" 1
02.04.2018 15:22:20 [GAUZBGB1PACS] Presentation Context 29 "1.2.840.10008.5.1.4.1.1.1.1.1" 1
02.04.2018 15:22:20 [GAUZBGB1PACS] Presentation Context 30 "1.2.840.10008.5.1.4.1.1.1.2" 1
02.04.2018 15:22:20 [GAUZBGB1PACS] Presentation Context 31 "1.2.840.10008.5.1.4.1.1.1.1.2.1" 0
02.04.2018 15:22:20 [GAUZBGB1PACS] Presentation Context 32 "1.2.840.10008.5.1.4.1.1.1.3" 1
02.04.2018 15:22:20 [GAUZBGB1PACS] Presentation Context 33 "1.2.840.10008.5.1.4.1.1.1.1.3.1" 0
02.04.2018 15:22:20 [GAUZBGB1PACS] Presentation Context 34 "1.2.840.113704.7.0.1.9" 0
02.04.2018 15:22:20 [GAUZBGB1PACS] Presentation Context 35 "1.2.840.113704.7.0.5.1" 0
02.04.2018 15:22:20 [GAUZBGB1PACS] Presentation Context 36 "1.3.46.670589.5.0.1.1" 0
02.04.2018 15:22:20 [GAUZBGB1PACS] Presentation Context 37 "1.3.46.670589.5.0.2.1" 0
02.04.2018 15:22:20 [GAUZBGB1PACS] Presentation Context 38 "1.3.46.670589.5.0.3.1" 0
02.04.2018 15:22:20 [GAUZBGB1PACS] Presentation Context 39 "1.3.46.670589.5.0.10" 0
02.04.2018 15:22:20 [GAUZBGB1PACS] Presentation Context 40 "1.3.46.670589.5.0.9" 0
02.04.2018 15:22:20 [GAUZBGB1PACS] Presentation Context 41 "1.3.46.670589.2.4.1.1" 0
02.04.2018 15:22:20 [GAUZBGB1PACS] Presentation Context 42 "1.3.46.670589.5.0.12" 0
02.04.2018 15:22:20 [GAUZBGB1PACS] Rewritten file: d:\distribution\data\431812\1.3.6.1.4.1.30071.8.176156457628161.5759843185196041\1.2.840.113704.1.111.4384.1522669604.7\1.2.840.113704.1.111.5796.1522669659.214.dcm
20180402 15:23:53 Started zip and cleanup thread
20180402 15:23:53 Monitoring for files in: d:\distribution\data\incoming\
20180402 15:23:53 DGATE (1.4.19b, build Thu Oct 12 18:42:54 2017, bits 64) is running as threaded server
20180402 15:23:53 Database type: ODBC connection
02.04.2018 15:23:57 [GAUZBGB1PACS] Stopped zip and cleanup thread
02.04.2018 15:23:57 [GAUZBGB1PACS] User interface test: local server is running!
02.04.2018 15:24:14 [GAUZBGB1PACS]
02.04.2018 15:24:14 [GAUZBGB1PACS] UPACS THREAD 2: STARTED AT: Mon Apr 02 15:24:14 2018
02.04.2018 15:24:14 [GAUZBGB1PACS] Calling Application Title : "MXV "
02.04.2018 15:24:14 [GAUZBGB1PACS] Called Application Title : "GAUZBGB1PACS "
02.04.2018 15:24:14 [GAUZBGB1PACS] Application Context : "1.2.840.10008.3.1.1.1", PDU length: 16384
02.04.2018 15:24:14 [GAUZBGB1PACS] Presentation Context 0 "1.2.840.10008.5.1.4.1.1.1" 1
02.04.2018 15:24:14 [GAUZBGB1PACS] Presentation Context 1 "1.2.840.10008.5.1.4.1.1.2" 1
02.04.2018 15:24:14 [GAUZBGB1PACS] Presentation Context 2 "1.2.840.10008.5.1.4.1.1.3.1" 1
02.04.2018 15:24:14 [GAUZBGB1PACS] Presentation Context 3 "1.2.840.10008.5.1.4.1.1.4" 1
02.04.2018 15:24:14 [GAUZBGB1PACS] Presentation Context 4 "1.2.840.10008.5.1.4.1.1.6.1" 1
02.04.2018 15:24:14 [GAUZBGB1PACS] Presentation Context 5 "1.2.840.10008.5.1.4.1.1.7" 1
02.04.2018 15:24:14 [GAUZBGB1PACS] Presentation Context 6 "1.2.840.10008.5.1.4.1.1.8" 1
02.04.2018 15:24:14 [GAUZBGB1PACS] Presentation Context 7 "1.2.840.10008.5.1.4.1.1.9" 1
02.04.2018 15:24:14 [GAUZBGB1PACS] Presentation Context 8 "1.2.840.10008.5.1.4.1.1.9.1.2" 1
02.04.2018 15:24:14 [GAUZBGB1PACS] Presentation Context 9 "1.2.840.10008.5.1.4.1.1.11.1" 1
02.04.2018 15:24:14 [GAUZBGB1PACS] Presentation Context 10 "1.2.840.10008.5.1.4.1.1.10" 1
02.04.2018 15:24:14 [GAUZBGB1PACS] Presentation Context 11 "1.2.840.10008.5.1.4.1.1.11" 1
02.04.2018 15:24:14 [GAUZBGB1PACS] Presentation Context 12 "1.2.840.10008.5.1.4.1.1.12.1" 1
02.04.2018 15:24:14 [GAUZBGB1PACS] Presentation Context 13 "1.2.840.10008.5.1.4.1.1.12.2" 1
02.04.2018 15:24:14 [GAUZBGB1PACS] Presentation Context 14 "1.2.840.10008.5.1.4.1.1.20" 1
02.04.2018 15:24:14 [GAUZBGB1PACS] Presentation Context 15 "1.2.840.10008.5.1.4.1.1.77.1.1" 1
02.04.2018 15:24:14 [GAUZBGB1PACS] Presentation Context 16 "1.2.840.10008.5.1.4.1.1.77.1.2" 1
02.04.2018 15:24:14 [GAUZBGB1PACS] Presentation Context 17 "1.2.840.10008.5.1.4.1.1.77.1.3" 1
02.04.2018 15:24:14 [GAUZBGB1PACS] Presentation Context 18 "1.2.840.10008.5.1.4.1.1.77.1.4" 1
02.04.2018 15:24:14 [GAUZBGB1PACS] Presentation Context 19 "1.2.840.10008.5.1.4.1.1.128" 1
02.04.2018 15:24:14 [GAUZBGB1PACS] Presentation Context 20 "1.2.840.10008.5.1.4.1.1.129" 1
02.04.2018 15:24:14 [GAUZBGB1PACS] Presentation Context 21 "1.2.840.10008.5.1.4.1.1.481.1" 1
02.04.2018 15:24:14 [GAUZBGB1PACS] Presentation Context 22 "1.2.840.10008.5.1.4.1.1.481.2" 1
02.04.2018 15:24:14 [GAUZBGB1PACS] Presentation Context 23 "1.2.840.10008.5.1.4.1.1.481.3" 1
02.04.2018 15:24:14 [GAUZBGB1PACS] Presentation Context 24 "1.2.840.10008.5.1.4.1.1.481.4" 1
02.04.2018 15:24:14 [GAUZBGB1PACS] Presentation Context 25 "1.2.840.10008.5.1.4.1.1.481.5" 1
02.04.2018 15:24:14 [GAUZBGB1PACS] Presentation Context 26 "1.2.840.10008.5.1.4.1.1.481.6" 1
02.04.2018 15:24:14 [GAUZBGB1PACS] Presentation Context 27 "1.2.840.10008.5.1.4.1.1.481.7" 1
02.04.2018 15:24:14 [GAUZBGB1PACS] Presentation Context 28 "1.2.840.10008.5.1.4.1.1.1.1" 1
02.04.2018 15:24:14 [GAUZBGB1PACS] Presentation Context 29 "1.2.840.10008.5.1.4.1.1.1.1.1" 1
02.04.2018 15:24:14 [GAUZBGB1PACS] Presentation Context 30 "1.2.840.10008.5.1.4.1.1.1.2" 1
02.04.2018 15:24:14 [GAUZBGB1PACS] Presentation Context 31 "1.2.840.10008.5.1.4.1.1.1.1.2.1" 0
02.04.2018 15:24:14 [GAUZBGB1PACS] Presentation Context 32 "1.2.840.10008.5.1.4.1.1.1.3" 1
02.04.2018 15:24:14 [GAUZBGB1PACS] Presentation Context 33 "1.2.840.10008.5.1.4.1.1.1.1.3.1" 0
02.04.2018 15:24:14 [GAUZBGB1PACS] Presentation Context 34 "1.2.840.113704.7.0.1.9" 0
02.04.2018 15:24:14 [GAUZBGB1PACS] Presentation Context 35 "1.2.840.113704.7.0.5.1" 0
02.04.2018 15:24:14 [GAUZBGB1PACS] Presentation Context 36 "1.3.46.670589.5.0.1.1" 0
02.04.2018 15:24:14 [GAUZBGB1PACS] Presentation Context 37 "1.3.46.670589.5.0.2.1" 0
02.04.2018 15:24:14 [GAUZBGB1PACS] Presentation Context 38 "1.3.46.670589.5.0.3.1" 0
02.04.2018 15:24:14 [GAUZBGB1PACS] Presentation Context 39 "1.3.46.670589.5.0.10" 0
02.04.2018 15:24:14 [GAUZBGB1PACS] Presentation Context 40 "1.3.46.670589.5.0.9" 0
02.04.2018 15:24:14 [GAUZBGB1PACS] Presentation Context 41 "1.3.46.670589.2.4.1.1" 0
02.04.2018 15:24:14 [GAUZBGB1PACS] Presentation Context 42 "1.3.46.670589.5.0.12" 0
02.04.2018 15:24:14 [GAUZBGB1PACS] Rewritten file: d:\distribution\data\431812\1.3.6.1.4.1.30071.8.176156457628161.5759843185196041\1.2.840.113704.1.111.4384.1522669553.3\1.2.840.113704.1.111.5796.1522669591.1.dcm
02.04.2018 15:24:14 [GAUZBGB1PACS] Rewritten file: d:\distribution\data\431812\1.3.6.1.4.1.30071.8.176156457628161.5759843185196041\1.2.840.113704.1.111.4384.1522669553.3\1.2.840.113704.7.1.1.1092.1522669616.1.dcm
02.04.2018 15:24:14 [GAUZBGB1PACS] Rewritten file: d:\distribution\data\431812\1.3.6.1.4.1.30071.8.176156457628161.5759843185196041\1.2.840.113704.1.111.4384.1522669553.3\1.2.840.113704.7.1.1.5248.1522669592.2.dcm
02.04.2018 15:24:14 [GAUZBGB1PACS] Rewritten file: d:\distribution\data\431812\1.3.6.1.4.1.30071.8.176156457628161.5759843185196041\1.2.840.113704.1.111.4384.1522669553.3\1.2.840.113704.7.1.1.5248.1522669813.482.dcm
02.04.2018 15:24:14 [GAUZBGB1PACS] Rewritten file: d:\distribution\data\431812\1.3.6.1.4.1.30071.8.176156457628161.5759843185196041\1.2.840.113704.1.111.4384.1522669604.7\1.2.840.113704.1.111.5796.1522669659.214.dcm
20180402 15:28:14 Started zip and cleanup thread
20180402 15:28:14 Monitoring for files in: d:\distribution\data\incoming\
20180402 15:28:14 DGATE (1.4.19b, build Thu Oct 12 18:42:54 2017, bits 64) is running as threaded server
20180402 15:28:14 Database type: ODBC connection
02.04.2018 15:28:18 [GAUZBGB1PACS] Stopped zip and cleanup thread
02.04.2018 15:28:18 [GAUZBGB1PACS] User interface test: local server is running!
02.04.2018 15:28:20 [GAUZBGB1PACS] set debug level from GUI
02.04.2018 15:28:20 [GAUZBGB1PACS] set debug log from GUI
02.04.2018 15:28:48 [GAUZBGB1PACS] -
Hi,
can you setup a temp server with another database (sqlite) to see if that is the issue? Just make a new folder and run it instead of the other. If you have access to visual studio I would like to see what it reports when dgate64.exe crashes (attach it to dgate64.exe).
Marcel
-
Quote from marcelvanherk
Hi,
can you setup a temp server with another database (sqlite) to see if that is the issue? Just make a new folder and run it instead of the other. If you have access to visual studio I would like to see what it reports when dgate64.exe crashes (attach it to dgate64.exe).
Marcel
Hi!
In the near future I will do that you said and write result. So, one question: Does conquest pacs contain storage commitment service? in settings of CT philips i set option that conquest pacs support storage commitment service. May be problem in that? -
Hi,
conquest does not do storage commitment and it should not be enabled in dgatesop.lst.
regards,
Marcel