I'm rather not experienced with Conquest, but situation looks reletively clear: the CR console (workstation) of brand new Vita CR is connected to the same private network segment as Conquest serwer (1 Gb/s). Although images of smaller dimensions are passing to server without problems, images of 35x35 cm and larger have to be sended 2 - 5 times to be finaly stored. The situation looks "stable". Looking forward for suggestions...
Store problem with Carestream Vita CR
-
-
Hi,
How big are the final images on disk and what are the compression settings? This looks like a timeout issue.
Are there any error messages? What version are you using?
Marcel
-
More than 10 MB. In most of cases 11 to 14 MB.
-
Aha: no error messages in serverstatus.log.
Version 1.4.17 -
Hm,
How long does the transfer of the big images take?
1) try enable jpeg support and store the images as UJ (uncompressed or jpeg).
2) Maybe also try to update to 1.4.17c (see forum) to be on the safe side.
Marcel
-
TYpical errornous sequence looks (in tke log) as follows:
[MILICZ1] 20140217 09:59:13 UPACS THREAD 13: STARTED AT: Mon Feb 17 09:59:10 2014
[MILICZ1] 20140217 09:59:14 Calling Application Title : "CSH_IMAGE "
[MILICZ1] 20140217 09:59:14 Called Application Title : "MILICZ1 "
[MILICZ1] 20140217 09:59:14 Application Context : "1.2.840.10008.3.1.1.1", PDU length: 32768
[MILICZ1] 20140217 09:59:14 Presentation Context 0 "1.2.840.10008.1.1" 1
[MILICZ1] 20140217 09:59:14 Presentation Context 1 "1.2.840.10008.5.1.4.1.1.1" 1
[MILICZ1] 20140217 09:59:14 Presentation Context 2 "1.2.840.10008.5.1.4.1.1.1" 1
[MILICZ1] 20140217 09:59:21 [recompress]: recompressed with mode = jk (strip=0)
[MILICZ1] 20140217 09:59:21 Rewritten file: D:\Conquest\Data\PACJ000043\1.2.840.113564.52.192.168.3.11.3652.20140214164256295_0001_000001_13923926390007.dcm
[MILICZ1] 20140217 09:59:21 UPACS THREAD 13: ENDED AT: Mon Feb 17 09:59:21 2014
[MILICZ1] 20140217 09:59:21 UPACS THREAD 13: TOTAL RUNNING TIME: 11 SECONDSOf course after 11 seconds transmission is broken.
-
I will try with Your suggestions tomorrow. Thanks so far.
-
Hi,
JK is jpeg2000 compression which is quite slow. If the VITA has a short timeout it may hang up while processing is still ongoing.
By the looks of it VITA takes 4 s to send the image, while it takes 7 s to compress and store. Compression UN and J1 will be much faster.
Marcel
-
After upgrade to 1.4.17c and 2 days of observation must say that now it goes well. Even techicians are satisfied.
So I will look for logs for couple of days and then give You final answer.
And after that I'll ask You another question... but about a bit different problem.