Query problem

  • Hello and greeting from Australia.


    We set up two Conquest server on one computer to receive patient data respectively from our Picker and Siemens Gamma camera, they worked quite well for the past three year till last week;


    When I tried to do a query from the Picker to the Conquest Picker database, it only returned back data of the 2006 and 2007, no data of the 2008; and if I from the Picker tried to query the Siemens database, everything was fine;


    Also if I query from our GE Xeleris to both the Conquest database, both results are normal;


    quite strange problem and anyone can help?



    thanks in advance.

  • Quote from marcelvanherk

    Hi,


    can you show the debug log of the correct and failed query?


    Marcel



    Thanks Marcel for your reply, also thanks very much for your wonderful job on Conquest.


    Sorry I was off for several days.

  • dicom.ini file for the wrong DICOM server:



    # This file contains configuration information for the DICOM server
    # Do not edit unless you know what you are doing


    [sscscp]
    MicroPACS = sscscp
    Edition = Personal


    # Network configuration: server name and TCP/IP port#
    MyACRNema = PICKER3_DB
    TCPPort = 107


    # Reference to other files: known dicom servers; database layout; sops
    ACRNemaMap = acrnema.map
    kFactorFile = dicom.sql
    SOPClassList = dgatesop.lst


    # Host(ignored), name, username and password for ODBC data source
    SQLHost = localhost
    SQLServer = S:\dbase\
    Username = conquest
    Password = conquest
    DoubleBackSlashToDB = 0


    # Configure database
    TruncateFieldNames = 10
    MaxFieldLength = 254
    FileNameSyntax = 3
    MaxFileNameLength = 255
    FixPhilips = 0
    FixKodak = 0
    KeepAlive = 60
    LargeFileSizeKB = 1024
    ZipTime = 05:
    UIDPrefix = 1.2.826.0.1.3680043.2.135.732666.68020625
    EnableReadAheadThread = 1
    PatientQuerySortOrder = 0
    StudyQuerySortOrder =
    SeriesQuerySortOrder =
    ImageQuerySortOrder =
    IndexDBF = 1
    PackDBF = 0
    LongQueryDBF = 1000
    TCPIPTimeOut = 300
    FailHoldOff = 60
    RetryDelay = 100
    QueueSize = 128
    WorkListMode = 0
    DebugLevel = 0


    # Configuration of compression for incoming images and archival
    DroppedFileCompression = un
    IncomingCompression = un
    ArchiveCompression = as


    # Names of the database tables
    PatientTableName = DICOMPatients
    StudyTableName = DICOMStudies
    SeriesTableName = DICOMSeries
    ImageTableName = DICOMImages
    DMarkTableName = DICOMAccessUpdates
    RegisteredMOPDeviceTable = RegisteredMOPIDs
    UIDToMOPIDTable = UIDToMOPID
    UIDToCDRIDTable = UIDToCDRID


    # Banner and host for debug information
    PACSName = PICKER3_DB
    OperatorConsole = 127.0.0.1


    # Configuration of disk(s) to store images
    MAGDeviceThreshhold = 0
    MAGDevices = 1
    MAGDevice0 = S:\
    NightlyCleanThreshhold = 0


    # Configuration of converter programs to export DICOM slices
    ExportConverters = 1
    ExportConverter0 = forward to JSWS2

  • dgatesop.lst file for the wrong server:



    #
    # DICOM Application / sop / transfer UID list.
    #
    # This list is used by the CheckedPDU_Service ( "filename" ) service
    # class. All incoming associations will be verified against this
    # file.
    #
    # Revision 2: disabled GEMRStorage and GECTStorage
    # Revision 3: extended with new sops and with JPEG transfer syntaxes
    # Revision 4: added Modality Worklist query
    #
    #None none RemoteAE
    #None none LocalAE
    #DICOM 1.2.840.10008.3.1.1.1 application
    Verification 1.2.840.10008.1.1 sop
    StoredPrintStorage 1.2.840.10008.5.1.1.27 sop
    HardcopyGrayscaleImageStorage 1.2.840.10008.5.1.1.29 sop
    HardcopyColorImageStorage 1.2.840.10008.5.1.1.30 sop
    CRStorage 1.2.840.10008.5.1.4.1.1.1 sop
    DXStorageForPresentation 1.2.840.10008.5.1.4.1.1.1.1 sop
    DXStorageForProcessing 1.2.840.10008.5.1.4.1.1.1.1.1 sop
    DMStorageForPresentation 1.2.840.10008.5.1.4.1.1.1.2 sop
    DMStorageForProcessing 1.2.840.10008.5.1.4.1.1.1.2.1 sop
    DOralStorageForPresentation 1.2.840.10008.5.1.4.1.1.1.3 sop
    DOralStorageForProcessing 1.2.840.10008.5.1.4.1.1.1.3.1 sop
    CTStorage 1.2.840.10008.5.1.4.1.1.2 sop
    RetiredUSMultiframeStorage 1.2.840.10008.5.1.4.1.1.3 sop
    USMultiframeStorage 1.2.840.10008.5.1.4.1.1.3.1 sop
    MRStorage 1.2.840.10008.5.1.4.1.1.4 sop
    MRImageStorageEnhanced 1.2.840.10008.5.1.4.1.1.4.1 sop
    MRStorageSpectroscopy 1.2.840.10008.5.1.4.1.1.4.2 sop
    RetiredNMStorage 1.2.840.10008.5.1.4.1.1.5 sop
    RetiredUSStorage 1.2.840.10008.5.1.4.1.1.6 sop
    USStorage 1.2.840.10008.5.1.4.1.1.6.1 sop
    SCStorage 1.2.840.10008.5.1.4.1.1.7 sop
    SCStorageSingleBitMF 1.2.840.10008.5.1.4.1.1.7.1 sop
    SCStorageGrayscaleByteMF 1.2.840.10008.5.1.4.1.1.7.2 sop
    SCStorageGrayscaleWordMF 1.2.840.10008.5.1.4.1.1.7.3 sop
    SCStorageTrueColorMF 1.2.840.10008.5.1.4.1.1.7.4 sop
    StandaloneOverlayStorage 1.2.840.10008.5.1.4.1.1.8 sop
    StandaloneCurveStorage 1.2.840.10008.5.1.4.1.1.9 sop
    #WFStorageTwelveLeadECG 1.2.840.10008.5.1.4.1.1.9.1.1 sop
    #WFStorageGeneralECG 1.2.840.10008.5.1.4.1.1.9.1.2 sop
    #WFStorageAmbulatoryECG 1.2.840.10008.5.1.4.1.1.9.1.3 sop
    #WFStorageHemodynamic 1.2.840.10008.5.1.4.1.1.9.2.1 sop
    #WFStorageCardiacElectrophysiology 1.2.840.10008.5.1.4.1.1.9.3.1 sop
    #WFStorageBasicVoiceAudio 1.2.840.10008.5.1.4.1.1.9.4.1 sop
    StandaloneModalityLUTStorage 1.2.840.10008.5.1.4.1.1.10 sop
    StandaloneVOILUTStorage 1.2.840.10008.5.1.4.1.1.11 sop
    GrayscaleSoftcopyPresentationStateStorage 1.2.840.10008.5.1.4.1.1.11.1 sop
    RetiredXASinglePlaneStorage 1.2.840.10008.5.1.4.1.1.12 sop
    XASinglePlaneStorage 1.2.840.10008.5.1.4.1.1.12.1 sop
    RFStorage 1.2.840.10008.5.1.4.1.1.12.2 sop
    XABiPlaneStorage 1.2.840.10008.5.1.4.1.1.12.3 sop
    NMStorage 1.2.840.10008.5.1.4.1.1.20 sop
    VLImageStorage 1.2.840.10008.5.1.4.1.1.77.1 sop
    VLMultiFrameImageStorage 1.2.840.10008.5.1.4.1.1.77.2 sop
    VLMicroscopicSlideStorage 1.2.840.10008.5.1.4.1.1.77.3 sop
    VLPhotographicStorage 1.2.840.10008.5.1.4.1.1.77.4 sop
    VLEndoscopicImageStorage 1.2.840.10008.5.1.4.1.1.77.1.1 sop
    VLMicroscopicImageStorage 1.2.840.10008.5.1.4.1.1.77.1.2 sop
    VLSlideCoordinatesMicroscopicImageStorage 1.2.840.10008.5.1.4.1.1.77.1.3 sop
    VLPhotographicImageStorage 1.2.840.10008.5.1.4.1.1.77.1.4 sop
    BasicTextSR 1.2.840.10008.5.1.4.1.1.88.11 sop
    EnhancedSR 1.2.840.10008.5.1.4.1.1.88.22 sop
    ComprehensiveSR 1.2.840.10008.5.1.4.1.1.88.33 sop
    MammographyCADSR 1.2.840.10008.5.1.4.1.1.88.50 sop
    KeyObjectSelectionDocument 1.2.840.10008.5.1.4.1.1.88.59 sop
    PETStorage 1.2.840.10008.5.1.4.1.1.128 sop
    StandalonePETCurveStorage 1.2.840.10008.5.1.4.1.1.129 sop
    RTImageStorage 1.2.840.10008.5.1.4.1.1.481.1 sop
    RTDoseStorage 1.2.840.10008.5.1.4.1.1.481.2 sop
    RTStructureStorage 1.2.840.10008.5.1.4.1.1.481.3 sop
    RTTreatmentRecordStorage 1.2.840.10008.5.1.4.1.1.481.4 sop
    RTPlanStorage 1.2.840.10008.5.1.4.1.1.481.5 sop
    RTBrachyTreatmentRecordStorage 1.2.840.10008.5.1.4.1.1.481.6 sop
    RTTreatmentSummaryRecordStorage 1.2.840.10008.5.1.4.1.1.481.7 sop
    #GEMRStorage 1.2.840.113619.4.2 sop
    #GECTStorage 1.2.840.113619.4.3 sop
    GE3DModelObjectStorage 1.2.840.113619.4.26 sop
    GERTPlanStorage 1.2.840.113619.5.249 sop
    GERTPlanStorage2 1.2.840.113619.4.5.249 sop
    GESaturnTDSObjectStorage 1.2.840.113619.5.253 sop
    Philips3DVolumeStorage 1.2.46.670589.5.0.1 sop
    Philips3DObjectStorage 1.2.46.670589.5.0.2 sop
    PhilipsSurfaceStorage 1.2.46.670589.5.0.3 sop
    PhilipsCompositeObjectStorage 1.2.46.670589.5.0.4 sop
    PhilipsMRCardioProfileStorage 1.2.46.670589.5.0.7 sop
    PhilipsMRCardioImageStorage 1.2.46.670589.5.0.8 sop
    PatientRootQuery 1.2.840.10008.5.1.4.1.2.1.1 sop
    PatientRootRetrieve 1.2.840.10008.5.1.4.1.2.1.2 sop
    StudyRootQuery 1.2.840.10008.5.1.4.1.2.2.1 sop
    StudyRootRetrieve 1.2.840.10008.5.1.4.1.2.2.2 sop
    PatientStudyOnlyQuery 1.2.840.10008.5.1.4.1.2.3.1 sop
    PatientStudyOnlyRetrieve 1.2.840.10008.5.1.4.1.2.3.2 sop
    FindModalityWorkList 1.2.840.10008.5.1.4.31 sop
    PatientRootRetrieveNKI 1.2.826.0.1.3680043.2.135.1066.5.1.4.1.2.1.2 sop
    StudyRootRetrieveNKI 1.2.826.0.1.3680043.2.135.1066.5.1.4.1.2.2.2 sop
    PatientStudyOnlyRetrieveNKI 1.2.826.0.1.3680043.2.135.1066.5.1.4.1.2.3.2 sop
    BasicGrayscalePrintManagementMeta 1.2.840.10008.5.1.1.9 sop
    BasicColorPrintManagementMeta 1.2.840.10008.5.1.1.18 sop
    BasicFilmSession 1.2.840.10008.5.1.1.1 sop
    BasicFilmBox 1.2.840.10008.5.1.1.2 sop
    BasicGrayscaleImageBox 1.2.840.10008.5.1.1.4 sop
    BasicColorImageBox 1.2.840.10008.5.1.1.4.1 sop
    BasicPrinter 1.2.840.10008.5.1.1.16 sop
    LittleEndianImplicit 1.2.840.10008.1.2 transfer
    #LittleEndianExplicit 1.2.840.10008.1.2.1 transfer
    #BigEndianExplicit 1.2.840.10008.1.2.2 transfer
    #JPEGBaseLine1 1.2.840.10008.1.2.4.50 transfer LittleEndianExplicit
    #JPEGExtended2and4 1.2.840.10008.1.2.4.51 transfer LittleEndianExplicit
    #JPEGExtended3and5 1.2.840.10008.1.2.4.52 transfer LittleEndianExplicit
    #JPEGSpectralNH6and8 1.2.840.10008.1.2.4.53 transfer LittleEndianExplicit
    #JPEGSpectralNH7and9 1.2.840.10008.1.2.4.54 transfer LittleEndianExplicit
    #JPEGFulllNH10and12 1.2.840.10008.1.2.4.55 transfer LittleEndianExplicit
    #JPEGFulllNH11and13 1.2.840.10008.1.2.4.56 transfer LittleEndianExplicit
    #JPEGLosslessNH14 1.2.840.10008.1.2.4.57 transfer LittleEndianExplicit
    #JPEGLosslessNH15 1.2.840.10008.1.2.4.58 transfer LittleEndianExplicit
    #JPEGExtended16and18 1.2.840.10008.1.2.4.59 transfer LittleEndianExplicit
    #JPEGExtended17and19 1.2.840.10008.1.2.4.60 transfer LittleEndianExplicit
    #JPEGSpectral20and22 1.2.840.10008.1.2.4.61 transfer LittleEndianExplicit
    #JPEGSpectral21and23 1.2.840.10008.1.2.4.62 transfer LittleEndianExplicit
    #JPEGFull24and26 1.2.840.10008.1.2.4.63 transfer LittleEndianExplicit
    #JPEGFull25and27 1.2.840.10008.1.2.4.64 transfer LittleEndianExplicit
    #JPEGLossless28 1.2.840.10008.1.2.4.65 transfer LittleEndianExplicit
    #JPEGLossless29 1.2.840.10008.1.2.4.66 transfer LittleEndianExplicit
    #JPEGLossless 1.2.840.10008.1.2.4.70 transfer LittleEndianExplicit
    #JPEGLS_Lossless 1.2.840.10008.1.2.4.80 transfer LittleEndianExplicit
    #JPEGLS_Lossy 1.2.840.10008.1.2.4.81 transfer LittleEndianExplicit
    #RLELossless 1.2.840.10008.1.2.5 transfer LittleEndianExplicit
    #LittleEndianExplicitDeflated 1.2.840.10008.1.2.1.99 transfer LittleEndianExplicit
    #JPEG2000LosslessOnly 1.2.840.10008.1.2.4.90 transfer LittleEndianExplicit
    #JPEG2000 1.2.840.10008.1.2.4.91 transfer LittleEndianExplicit

  • dicom.sql file for the wrong server:



    /*
    # DICOM Database layout
    # Example version for built-in DbaseIII driver (denormalized)
    #
    # (File DICOM.SQL)
    # ** DO NOT EDIT THIS FILE UNLESS YOU KNOW WHAT YOU ARE DOING **
    #
    # Version with modality moved to the series level and EchoNumber in image table
    # Revision 3: Patient birthday and sex, bolus agent, correct field lengths
    # Revision 4: Studymodality, Station and Department in study
    # Manufacturer, Model, BodyPart and Protocol in series
    # Acqdate/time, coil, acqnumber, slicelocation and pixel info in images
    # Notes for revision 4:
    # InstitutionalDepartmentName in study (should officially be in series, but eFilm expects it in study)
    # StationName is in study (should officially be in series, but more useful in study)
    # Revision 5: Added patientID in series and images for more efficient querying
    # Revision 6: Added frame of reference UID in series table
    # Revision 7: Added ImageType in image table, StudyModality to 64 chars, AcqDate to SQL_C_DATE
    # Revision 8: Denormalized study table (add patient ID, name, birthdate) to show consistency problems
    # Revision 9: Denormalized series and image table for full query support with built-in DBF driver
    # Revision 10: Fixed width of ReceivingCoil: to 16 chars
    # Revision 11: Added Manufacturer to denormalized image table for built-in DBF driver
    # Revision 12: Removed duplicate STUDYINST field in series for built-in DBF driver
    # Revision 13: Added ImageID to image database
    # Revision 14: Added WorkList database with HL7 tags
    #
    #
    # 5 databases need to be defined:
    #
    # *Patient*
    # *Study*
    # *Series*
    # *Image*
    # *WorkList*
    #
    #
    # The last defined element of Study is a link back to Patient
    # The last defined element of Series is a link back to Study
    # The last defined element of Image is a link back to Series
    #
    #
    # Format for DICOM databases :
    # { Group, Element, Column Name, Column Length, SQL-Type, DICOM-Type }
    # Format for Worklist database :
    # { Group, Element, Column Name, Column Length, SQL-Type, DICOM-Type, HL7 tag}
    # HL7 tags include SEQ.N, SEQ.N.M, SEQ.N.DATE, SEQ.N.TIME, *AN, *UI
    */


    *Patient*
    {
    { 0x0010, 0x0020, "PatientID", 64, SQL_C_CHAR, DT_STR },
    { 0x0010, 0x0010, "PatientName", 64, SQL_C_CHAR, DT_STR },
    { 0x0010, 0x0030, "PatientBirthDate", 8, SQL_C_DATE, DT_DATE },
    { 0x0010, 0x0040, "PatientSex", 16, SQL_C_CHAR, DT_STR }
    }


    *Study*
    {
    { 0x0020, 0x000d, "StudyInstanceUID", 64, SQL_C_CHAR, DT_UI },
    { 0x0008, 0x0020, "StudyDate", 8, SQL_C_DATE, DT_DATE },
    { 0x0008, 0x0030, "StudyTime", 16, SQL_C_CHAR, DT_TIME },
    { 0x0020, 0x0010, "StudyID", 16, SQL_C_CHAR, DT_STR },
    { 0x0008, 0x1030, "StudyDescription", 64, SQL_C_CHAR, DT_STR },
    { 0x0008, 0x0050, "AccessionNumber", 16, SQL_C_CHAR, DT_STR },
    { 0x0008, 0x0090, "ReferPhysician", 64, SQL_C_CHAR, DT_STR },
    { 0x0010, 0x1010, "PatientsAge", 16, SQL_C_CHAR, DT_STR },
    { 0x0010, 0x1030, "PatientsWeight", 16, SQL_C_CHAR, DT_STR },
    { 0x0008, 0x0061, "StudyModality", 64, SQL_C_CHAR, DT_STR },
    { 0x0008, 0x1010, "StationName", 16, SQL_C_CHAR, DT_STR },
    { 0x0008, 0x1040, "InstitutionalDepartmentName", 64, SQL_C_CHAR, DT_STR },


    { 0x0010, 0x0010, "PatientName", 64, SQL_C_CHAR, DT_STR },
    { 0x0010, 0x0030, "PatientBirthDate", 8, SQL_C_DATE, DT_DATE },
    { 0x0010, 0x0040, "PatientSex", 16, SQL_C_CHAR, DT_STR }


    { 0x0010, 0x0020, "PatientID", 64, SQL_C_CHAR, DT_STR }
    }


    *Series*
    {
    { 0x0020, 0x000e, "SeriesInstanceUID", 64, SQL_C_CHAR, DT_UI },
    { 0x0020, 0x0011, "SeriesNumber", 12, SQL_C_CHAR, DT_STR },
    { 0x0008, 0x0021, "SeriesDate", 8, SQL_C_DATE, DT_DATE },
    { 0x0008, 0x0031, "SeriesTime", 16, SQL_C_CHAR, DT_TIME },
    { 0x0008, 0x103e, "SeriesDescription", 64, SQL_C_CHAR, DT_STR },
    { 0x0008, 0x0060, "Modality", 16, SQL_C_CHAR, DT_STR },
    { 0x0018, 0x5100, "PatientPosition", 16, SQL_C_CHAR, DT_STR },
    { 0x0018, 0x0010, "ContrastBolusAgent", 64, SQL_C_CHAR, DT_STR },
    { 0x0008, 0x0070, "Manufacturer", 64, SQL_C_CHAR, DT_STR },
    { 0x0008, 0x1090, "ModelName", 64, SQL_C_CHAR, DT_STR },
    { 0x0018, 0x0015, "BodyPartExamined", 64, SQL_C_CHAR, DT_STR },
    { 0x0018, 0x1030, "ProtocolName", 64, SQL_C_CHAR, DT_STR },
    { 0x0020, 0x0052, "FrameOfReferenceUID", 64, SQL_C_CHAR, DT_UI },
    { 0x0010, 0x0020, "SeriesPat", 64, SQL_C_CHAR, DT_STR },


    { 0x0010, 0x0010, "PatientName", 64, SQL_C_CHAR, DT_STR },
    { 0x0010, 0x0030, "PatientBirthDate", 8, SQL_C_DATE, DT_DATE },
    { 0x0010, 0x0040, "PatientSex", 16, SQL_C_CHAR, DT_STR }


    { 0x0008, 0x0020, "StudyDate", 8, SQL_C_DATE, DT_DATE },
    { 0x0008, 0x0030, "StudyTime", 16, SQL_C_CHAR, DT_TIME },
    { 0x0020, 0x0010, "StudyID", 16, SQL_C_CHAR, DT_STR },
    { 0x0008, 0x1030, "StudyDescription", 64, SQL_C_CHAR, DT_STR },


    { 0x0020, 0x000d, "StudyInstanceUID", 64, SQL_C_CHAR, DT_UI }
    }


    *Image*
    {
    { 0x0008, 0x0018, "SOPInstanceUID", 64, SQL_C_CHAR, DT_UI },
    { 0x0008, 0x0016, "SOPClassUID", 64, SQL_C_CHAR, DT_UI },
    { 0x0020, 0x0013, "ImageNumber", 12, SQL_C_CHAR, DT_STR },
    { 0x0008, 0x0023, "ImageDate", 8, SQL_C_DATE, DT_DATE },
    { 0x0008, 0x0033, "ImageTime", 16, SQL_C_CHAR, DT_TIME },
    { 0x0018, 0x0086, "EchoNumber", 12, SQL_C_CHAR, DT_STR },
    { 0x0028, 0x0008, "NumberOfFrames", 12, SQL_C_CHAR, DT_STR },
    { 0x0008, 0x0022, "AcqDate", 8, SQL_C_DATE, DT_DATE },
    { 0x0008, 0x0032, "AcqTime", 16, SQL_C_CHAR, DT_TIME },
    { 0x0018, 0x1250, "ReceivingCoil", 16, SQL_C_CHAR, DT_STR },
    { 0x0020, 0x0012, "AcqNumber", 12, SQL_C_CHAR, DT_STR },
    { 0x0020, 0x1041, "SliceLocation", 16, SQL_C_CHAR, DT_STR },
    { 0x0028, 0x0002, "SamplesPerPixel", 5, SQL_C_CHAR, DT_UINT16 },
    { 0x0028, 0x0004, "PhotoMetricInterpretation", 16, SQL_C_CHAR, DT_STR },
    { 0x0028, 0x0010, "Rows", 5, SQL_C_CHAR, DT_UINT16 },
    { 0x0028, 0x0011, "Colums", 5, SQL_C_CHAR, DT_UINT16 },
    { 0x0028, 0x0101, "BitsStored", 5, SQL_C_CHAR, DT_UINT16 },
    { 0x0008, 0x0008, "ImageType", 128, SQL_C_CHAR, DT_STR },
    { 0x0054, 0x0400, "ImageID", 16, SQL_C_CHAR, DT_STR },
    { 0x0010, 0x0020, "ImagePat", 64, SQL_C_CHAR, DT_STR },


    { 0x0010, 0x0010, "PatientName", 64, SQL_C_CHAR, DT_STR },
    { 0x0010, 0x0030, "PatientBirthDate", 8, SQL_C_DATE, DT_DATE },
    { 0x0010, 0x0040, "PatientSex", 16, SQL_C_CHAR, DT_STR }


    { 0x0020, 0x000d, "StudyInstanceUID", 64, SQL_C_CHAR, DT_UI },
    { 0x0008, 0x0020, "StudyDate", 8, SQL_C_DATE, DT_DATE },
    { 0x0008, 0x0030, "StudyTime", 16, SQL_C_CHAR, DT_TIME },
    { 0x0020, 0x0010, "StudyID", 16, SQL_C_CHAR, DT_STR },
    { 0x0008, 0x1030, "StudyDescription", 64, SQL_C_CHAR, DT_STR },


    { 0x0020, 0x0011, "SeriesNumber", 12, SQL_C_CHAR, DT_STR },
    { 0x0008, 0x0021, "SeriesDate", 8, SQL_C_DATE, DT_DATE },
    { 0x0008, 0x0031, "SeriesTime", 16, SQL_C_CHAR, DT_TIME },
    { 0x0008, 0x103e, "SeriesDescription", 64, SQL_C_CHAR, DT_STR },
    { 0x0008, 0x0060, "Modality", 16, SQL_C_CHAR, DT_STR },
    { 0x0020, 0x0052, "FrameOfReferenceUID", 64, SQL_C_CHAR, DT_UI },
    { 0x0008, 0x0070, "Manufacturer", 64, SQL_C_CHAR, DT_STR },


    { 0x0020, 0x000e, "SeriesInstanceUID", 64, SQL_C_CHAR, DT_UI }
    }


    *WorkList*
    {
    { 0x0008, 0x0050, "AccessionNumber", 16, SQL_C_CHAR, DT_STR, "OBR.3" },
    { 0x0010, 0x0020, "PatientID", 64, SQL_C_CHAR, DT_STR, "PID.4" },
    { 0x0010, 0x0010, "PatientName", 64, SQL_C_CHAR, DT_STR, "PID.5" },
    { 0x0010, 0x0030, "PatientBirthDate", 8, SQL_C_DATE, DT_DATE, "PID.7" },
    { 0x0010, 0x0040, "PatientSex", 16, SQL_C_CHAR, DT_STR, "PID.8" },


    { 0x0010, 0x2000, "MedicalAlerts", 64, SQL_C_CHAR, DT_STR, "---" },
    { 0x0010, 0x2110, "ContrastAllergies", 64, SQL_C_CHAR, DT_STR, "---" },
    { 0x0020, 0x000d, "StudyInstanceUID", 64, SQL_C_CHAR, DT_UI, "---" },
    { 0x0032, 0x1032, "ReqPhysician", 64, SQL_C_CHAR, DT_STR, "OBR.16" },
    { 0x0032, 0x1060, "ReqProcDescription", 16, SQL_C_CHAR, DT_STR, "OBR.4.1" },


    { 0x0040, 0x0100, "--------", 0, SQL_C_CHAR, DT_STARTSEQUENCE, "---" },
    { 0x0008, 0x0060, "Modality", 16, SQL_C_CHAR, DT_STR, "OBR.21" },
    { 0x0032, 0x1070, "ReqContrastAgent", 64, SQL_C_CHAR, DT_STR, "---" },
    { 0x0040, 0x0001, "ScheduledAE", 16, SQL_C_CHAR, DT_STR, "---" },
    { 0x0040, 0x0002, "StartDate", 8, SQL_C_DATE, DT_DATE, "OBR.7.DATE" },
    { 0x0040, 0x0003, "StartTime", 16, SQL_C_CHAR, DT_TIME, "OBR.7.TIME" },
    { 0x0040, 0x0006, "PerfPhysician", 64, SQL_C_CHAR, DT_STR, "---" },
    { 0x0040, 0x0007, "SchedPSDescription", 64, SQL_C_CHAR, DT_STR, "---" },
    { 0x0040, 0x0009, "SchedPSID", 16, SQL_C_CHAR, DT_STR, "OBR.4" },
    { 0x0040, 0x0010, "SchedStationName", 16, SQL_C_CHAR, DT_STR, "OBR.24" },
    { 0x0040, 0x0011, "SchedPSLocation", 16, SQL_C_CHAR, DT_STR, "---" },
    { 0x0040, 0x0012, "PreMedication", 64, SQL_C_CHAR, DT_STR, "---" },
    { 0x0040, 0x0400, "SchedPSComments", 64, SQL_C_CHAR, DT_STR, "---" },
    { 0x0040, 0x0100, "---------", 0, SQL_C_CHAR, DT_ENDSEQUENCE, "---" },


    { 0x0040, 0x1001, "ReqProcID", 16, SQL_C_CHAR, DT_STR, "OBR.4.0" },
    { 0x0040, 0x1003, "ReqProcPriority", 16, SQL_C_CHAR, DT_STR, "OBR.27 }
    }

  • The problem is that everytime when I do a query, it only returns data of 2006 and 2007, lost the data of 2008 and, I can see the data if I browse the patient.


    it worked fine and proble just began last week.



    Thanks

  • Hi,


    I was referring to this statement:


    Quote

    Also if I query from our GE Xeleris to both the Conquest database, both results are normal;


    So one system's queries works and the other one fails. This is strange, and therefore I asked logs of a query originating from both systems to the failing conquest dicom server.


    When data in unailable this means that maybe the index is lost? For dbase, this index is -in memory on patient ID. What do you see in the logs when you restart the server?


    Marcel

  • Thaks Marcel,


    ya it is strange, simply say:


    we set up on one computer C two conquest server receiving data from A and B, on C we have A_DB and B_DB to store the data, it worked perfectly till last week.


    Now from A I can query both A_DB and B_DB, but from B I can query A_DB but when try to query B_DB, it only returns part of the data, not all ( only can get data of 2006 and 2007, no 2008 and 2009).


    We also have a Xeleris, from it querying A_DB and B_DB works fine.

  • Hi,


    sizes up to 50 million images have been used - so the size is not the problem. I still would like to see the debug log of queries from A and B:


    try


    kill the running dgate
    dgate -v &
    dgate --debuglevel:4


    now do a query from A and then a query from B and show me the query data.


    Marcel

  • Thanks Marcel, I will do that today and report you the result.


    Also we still use 1.4.12, and just use built-in dbaseIII driver, in the manual it said for small archives of up to 100.000 images, totally in the two folder we have more than that, is it a problem?


    Cheers

  • No,


    also dbaseIII can handle millions of images, although there is a 2 GB limit for the DICOMimages table in version 1.4.12. This limit is only reached at 1.5 million images. There was also patch for 1.4.12 for potential database corruption released (a modified ODBCI.CPP). Did you apply that patch? Note, modern versions don't have the limit or the bug.


    Or maybe you should just upgrade, and regenerate the database. That can be done in a separate folder pointing to the same data (i.e., side by side with a running 1.4.12), as long as you use a different port number for the server.


    Marcel