Hello Andreas
I have some problems to use IQ with TIMS . I have syntax error with that.
Regards
Behrang Karimi
HSP
Hello Andreas
I have some problems to use IQ with TIMS . I have syntax error with that.
Regards
Behrang Karimi
HSP
Can you give me further detail on this issue? What kind of communication are you trying and what error messages appear?
Regards,
Andreas
I've used Auto send , and this is TIMS log
08/26/07::18:15:29:Started DICOM Network
08/26/07::18:15:29:Patient Name ZAHRA MEHRKISH : Patient ID: 81-HB
08/26/07::18:15:29:Study CT - 08/26/2007 09:47:22
08/26/07::18:15:29:Loaded DICOM file C:\Program Files\Foresight\TIMS\Studies\Firuzgar Hospital_81-HB_CT_20070826_094722\Firuzgar Hospital_81-HB_CT_20070826_094722_0001.DCM
08/26/07::18:15:29:Study file 1 contains 1 image
08/26/07::18:15:29:Connecting From TIMS 192.168.10.56 4006
08/26/07::18:15:29:Connecting To host2 192.168.10.54 4006
08/26/07::18:15:29:Connected to 192.168.10.54 Successfully
08/26/07::18:15:34:Net Receive Associate Accept - Rejected: Transfer Syntax Not Supported
08/26/07::18:15:34:Net Receive Associate Accept - Rejected: Transfer Syntax Not Supported
08/26/07::18:15:34:Net Receive Associate Accept - Rejected: Transfer Syntax Not Supported
08/26/07::18:15:34:Net Receive Associate Accept - Rejected: Transfer Syntax Not Supported
08/26/07::18:15:34:No Acceptable Transfer Syntax Available to Send
The problem seems to be an unsupported transfer syntax. I compared the DICOM Conformance Statement of the TIMS and iQ-VIEW. TIMS supports JPEG Lossy transfer syntaxes but iQ-VIEW does not.
Check in your configuration of your TIMS that you use JPEG Lossless (1.2.840.10008.1.2.4.70) or an uncompressed syntax.
Regards, Yves