  • I've used Auto send , and this is TIMS log



    08/26/07::18:15:29:Started DICOM Network
    08/26/07::18:15:29:Patient Name ZAHRA MEHRKISH : Patient ID: 81-HB
    08/26/07::18:15:29:Study CT - 08/26/2007 09:47:22
    08/26/07::18:15:29:Loaded DICOM file C:\Program Files\Foresight\TIMS\Studies\Firuzgar Hospital_81-HB_CT_20070826_094722\Firuzgar Hospital_81-HB_CT_20070826_094722_0001.DCM
    08/26/07::18:15:29:Study file 1 contains 1 image
    08/26/07::18:15:29:Connecting From TIMS 192.168.10.56 4006
    08/26/07::18:15:29:Connecting To host2 192.168.10.54 4006
    08/26/07::18:15:29:Connected to 192.168.10.54 Successfully
    08/26/07::18:15:34:Net Receive Associate Accept - Rejected: Transfer Syntax Not Supported
    08/26/07::18:15:34:Net Receive Associate Accept - Rejected: Transfer Syntax Not Supported
    08/26/07::18:15:34:Net Receive Associate Accept - Rejected: Transfer Syntax Not Supported
    08/26/07::18:15:34:Net Receive Associate Accept - Rejected: Transfer Syntax Not Supported
    08/26/07::18:15:34:No Acceptable Transfer Syntax Available to Send

  • The problem seems to be an unsupported transfer syntax. I compared the DICOM Conformance Statement of the TIMS and iQ-VIEW. TIMS supports JPEG Lossy transfer syntaxes but iQ-VIEW does not.
    Check in your configuration of your TIMS that you use JPEG Lossless (1.2.840.10008.1.2.4.70) or an uncompressed syntax.


    Regards, Yves