Hi,
I´ve been using conquest in the last months as a pacs server, but now I ´d like to use the worklist function.
So ´m interest, if anybody experienced the same problems.
Syngo uses two different ways to ask the RIS Server. Automatically every xx minutes and manual by asking e.g. a name field.
I tried the sample.hl7 file.
Using the automatic retrieve, Conquest returns a 0, i guess because the stationaet field is empty. I did not figure out, where to put the stationaet in the hl7 file. Should be somewhere in the obr line ?? Any idea. I modified the file for the correct date.
Using the manual retrieve, in this case asking for Z* in the name field, syngo returns an error message : fault in the worklist data. (version 1414. in version 1413 this request resulted in a syngo system hang).
In the conquest log file i can see, that conquest found the file. I have tested the sample.hl7 with two other worklistservers (iq-worklist and dicmwst) without problems (regarding the manual retrieve, the automatic retrieve didnt work as well).
So if you have any ideas how to solve the problem, please let me know.
Uli
I attached the conquest log file, first the automatic request, the second for the manual request.
12.04.2008 14:05:41 [CONQUESTSRV] UPACS THREAD 5: STARTED AT: Sat Apr 12 14:05:40 2008
12.04.2008 14:05:41 [CONQUESTSRV] A-ASSOCIATE-RQ Packet Dump
12.04.2008 14:05:41 [CONQUESTSRV] Calling Application Title : "HRI_MRC10503 "
12.04.2008 14:05:41 [CONQUESTSRV] Called Application Title : "CONQUESTSRV "
12.04.2008 14:05:41 [CONQUESTSRV] Application Context : "1.2.840.10008.3.1.1.1"
12.04.2008 14:05:41 [CONQUESTSRV] Number of Proposed Presentation Contexts: 1
12.04.2008 14:05:41 [CONQUESTSRV] Presentation Context 0 "1.2.840.10008.5.1.4.31"
12.04.2008 14:05:41 [CONQUESTSRV] Server Command := 0020
12.04.2008 14:05:41 [CONQUESTSRV] Message ID := 0003
12.04.2008 14:05:41 [CONQUESTSRV] (ModalityWorkListQuery) search level:
12.04.2008 14:05:41 [CONQUESTSRV] Query On Modality WorkList
12.04.2008 14:05:41 [CONQUESTSRV] Failed on VR Search: 0002 0010
12.04.2008 14:05:41 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:41 [CONQUESTSRV] Failed on VR Search: 0008 0005
12.04.2008 14:05:41 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:41 [CONQUESTSRV] Failed on VR Search: 0008 0080
12.04.2008 14:05:41 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:41 [CONQUESTSRV] Failed on VR Search: 0008 0081
12.04.2008 14:05:41 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:41 [CONQUESTSRV] Failed on VR Search: 0008 0090
12.04.2008 14:05:41 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:41 [CONQUESTSRV] Failed on VR Search: 0008 1080
12.04.2008 14:05:41 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:41 [CONQUESTSRV] Failed on VR Search: 0008 1110
12.04.2008 14:05:41 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:41 [CONQUESTSRV] Failed on VR Search: 0008 1120
12.04.2008 14:05:41 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:41 [CONQUESTSRV] Failed on VR Search: 0010 1000
12.04.2008 14:05:41 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:41 [CONQUESTSRV] Failed on VR Search: 0010 1001
12.04.2008 14:05:41 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:41 [CONQUESTSRV] Failed on VR Search: 0010 1020
12.04.2008 14:05:41 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:41 [CONQUESTSRV] Failed on VR Search: 0010 1030
12.04.2008 14:05:41 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:41 [CONQUESTSRV] Failed on VR Search: 0010 1040
12.04.2008 14:05:41 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:41 [CONQUESTSRV] Failed on VR Search: 0010 1080
12.04.2008 14:05:41 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:41 [CONQUESTSRV] Failed on VR Search: 0010 2160
12.04.2008 14:05:41 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:41 [CONQUESTSRV] Failed on VR Search: 0010 21a0
12.04.2008 14:05:41 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:41 [CONQUESTSRV] Failed on VR Search: 0010 21b0
12.04.2008 14:05:41 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:41 [CONQUESTSRV] Failed on VR Search: 0010 21c0
12.04.2008 14:05:41 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:41 [CONQUESTSRV] Failed on VR Search: 0010 21d0
12.04.2008 14:05:41 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:41 [CONQUESTSRV] Failed on VR Search: 0010 4000
12.04.2008 14:05:41 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:41 [CONQUESTSRV] Failed on VR Search: 0032 1033
12.04.2008 14:05:41 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:41 [CONQUESTSRV] Failed on VR Search: 0032 1064
12.04.2008 14:05:41 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:41 [CONQUESTSRV] Failed on VR Search: 0038 0010
12.04.2008 14:05:41 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:41 [CONQUESTSRV] Failed on VR Search: 0038 0011
12.04.2008 14:05:41 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:41 [CONQUESTSRV] Failed on VR Search: 0038 0050
12.04.2008 14:05:41 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:41 [CONQUESTSRV] Failed on VR Search: 0038 0300
12.04.2008 14:05:41 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:41 [CONQUESTSRV] Failed on VR Search: 0038 0500
12.04.2008 14:05:41 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:41 [CONQUESTSRV] Failed on VR Search: 0040 0004
12.04.2008 14:05:41 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:41 [CONQUESTSRV] Failed on VR Search: 0040 0005
12.04.2008 14:05:41 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:41 [CONQUESTSRV] Failed on VR Search: 0040 0008
12.04.2008 14:05:41 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:41 [CONQUESTSRV] Failed on VR Search: 0040 0020
12.04.2008 14:05:41 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:41 [CONQUESTSRV] Failed on VR Search: 0040 1002
12.04.2008 14:05:41 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:41 [CONQUESTSRV] Failed on VR Search: 0040 1004
12.04.2008 14:05:41 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:41 [CONQUESTSRV] Failed on VR Search: 0040 1005
12.04.2008 14:05:41 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:41 [CONQUESTSRV] Failed on VR Search: 0040 1008
12.04.2008 14:05:41 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:41 [CONQUESTSRV] Failed on VR Search: 0040 1009
12.04.2008 14:05:41 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:41 [CONQUESTSRV] Failed on VR Search: 0040 1010
12.04.2008 14:05:41 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:41 [CONQUESTSRV] Failed on VR Search: 0040 1400
12.04.2008 14:05:41 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:41 [CONQUESTSRV] Failed on VR Search: 0040 2001
12.04.2008 14:05:41 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:41 [CONQUESTSRV] Failed on VR Search: 0040 2004
12.04.2008 14:05:41 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:41 [CONQUESTSRV] Failed on VR Search: 0040 2005
12.04.2008 14:05:41 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:41 [CONQUESTSRV] Failed on VR Search: 0040 2008
12.04.2008 14:05:41 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:41 [CONQUESTSRV] Failed on VR Search: 0040 2009
12.04.2008 14:05:41 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:41 [CONQUESTSRV] Failed on VR Search: 0040 2010
12.04.2008 14:05:41 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:41 [CONQUESTSRV] Failed on VR Search: 0040 2016
12.04.2008 14:05:41 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:41 [CONQUESTSRV] Failed on VR Search: 0040 2017
12.04.2008 14:05:41 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:41 [CONQUESTSRV] Failed on VR Search: 0040 2400
12.04.2008 14:05:41 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:41 [CONQUESTSRV] Failed on VR Search: 0040 3001
12.04.2008 14:05:41 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:41 [CONQUESTSRV] Issue Query on Columns: DICOMWorkList.AccessionN, DICOMWorkList.PatientNam, DICOMWorkList.PatientID, DICOMWorkList.PatientBir, DICOMWorkList.PatientSex, DICOMWorkList.MedicalAle, DICOMWorkList.ContrastAl, DICOMWorkList.StudyInsta, DICOMWorkList.ReqPhysici, DICOMWorkList.ReqProcDes, DICOMWorkList.Modality, DICOMWorkList.ReqContras, DICOMWorkList.ScheduledA, DICOMWorkList.StartDate, DICOMWorkList.StartTime, DICOMWorkList.PerfPhysic, DICOMWorkList.SchedPSDes, DICOMWorkList.SchedPSID, DICOMWorkList.SchedStati, DICOMWorkList.SchedPSLoc, DICOMWorkList.PreMedicat, DICOMWorkList.SchedPSCom, DICOMWorkList.ReqProcID, DICOMWorkList.ReqProcPri
12.04.2008 14:05:41 [CONQUESTSRV] Values: DICOMWorkList.Modality = 'MR' and DICOMWorkList.StartDate = '20080412' and DICOMWorkList.StartTime >= '000000.000000' and DICOMWorkList.StartTime <= '235959.000000'
12.04.2008 14:05:41 [CONQUESTSRV] Tables: DICOMWorkList
12.04.2008 14:05:41 [CONQUESTSRV] Sorting (DICOMWorkList.PatientNam) DoSort := 1
12.04.2008 14:05:41 [CONQUESTSRV] Records = 0
12.04.2008 14:05:41 [CONQUESTSRV] C-Find (Modality Work List) located 0 records
12.04.2008 14:05:41 [CONQUESTSRV] UPACS THREAD 5: ENDED AT: Sat Apr 12 14:05:41 2008
12.04.2008 14:05:41 [CONQUESTSRV] UPACS THREAD 5: TOTAL RUNNING TIME: 1 SECONDS
12.04.2008 14:05:56 [CONQUESTSRV] UPACS THREAD 6: STARTED AT: Sat Apr 12 14:05:55 2008
12.04.2008 14:05:56 [CONQUESTSRV] A-ASSOCIATE-RQ Packet Dump
12.04.2008 14:05:56 [CONQUESTSRV] Calling Application Title : "HRI_MRC10503 "
12.04.2008 14:05:56 [CONQUESTSRV] Called Application Title : "CONQUESTSRV "
12.04.2008 14:05:56 [CONQUESTSRV] Application Context : "1.2.840.10008.3.1.1.1"
12.04.2008 14:05:56 [CONQUESTSRV] Number of Proposed Presentation Contexts: 1
12.04.2008 14:05:56 [CONQUESTSRV] Presentation Context 0 "1.2.840.10008.5.1.4.31"
12.04.2008 14:05:56 [CONQUESTSRV] Server Command := 0020
12.04.2008 14:05:56 [CONQUESTSRV] Message ID := 0004
12.04.2008 14:05:56 [CONQUESTSRV] (ModalityWorkListQuery) search level:
12.04.2008 14:05:56 [CONQUESTSRV] Query On Modality WorkList
12.04.2008 14:05:56 [CONQUESTSRV] Failed on VR Search: 0002 0010
12.04.2008 14:05:56 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:56 [CONQUESTSRV] Failed on VR Search: 0008 0005
12.04.2008 14:05:56 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:56 [CONQUESTSRV] Failed on VR Search: 0008 0080
12.04.2008 14:05:56 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:56 [CONQUESTSRV] Failed on VR Search: 0008 0081
12.04.2008 14:05:56 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:56 [CONQUESTSRV] Failed on VR Search: 0008 0090
12.04.2008 14:05:56 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:56 [CONQUESTSRV] Failed on VR Search: 0008 1080
12.04.2008 14:05:56 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:56 [CONQUESTSRV] Failed on VR Search: 0008 1110
12.04.2008 14:05:56 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:56 [CONQUESTSRV] Failed on VR Search: 0008 1120
12.04.2008 14:05:56 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:56 [CONQUESTSRV] Failed on VR Search: 0010 1000
12.04.2008 14:05:56 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:56 [CONQUESTSRV] Failed on VR Search: 0010 1001
12.04.2008 14:05:56 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:56 [CONQUESTSRV] Failed on VR Search: 0010 1020
12.04.2008 14:05:56 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:56 [CONQUESTSRV] Failed on VR Search: 0010 1030
12.04.2008 14:05:56 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:56 [CONQUESTSRV] Failed on VR Search: 0010 1040
12.04.2008 14:05:56 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:56 [CONQUESTSRV] Failed on VR Search: 0010 1080
12.04.2008 14:05:56 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:56 [CONQUESTSRV] Failed on VR Search: 0010 2160
12.04.2008 14:05:56 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:56 [CONQUESTSRV] Failed on VR Search: 0010 21a0
12.04.2008 14:05:56 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:56 [CONQUESTSRV] Failed on VR Search: 0010 21b0
12.04.2008 14:05:56 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:56 [CONQUESTSRV] Failed on VR Search: 0010 21c0
12.04.2008 14:05:56 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:56 [CONQUESTSRV] Failed on VR Search: 0010 21d0
12.04.2008 14:05:56 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:56 [CONQUESTSRV] Failed on VR Search: 0010 4000
12.04.2008 14:05:56 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:56 [CONQUESTSRV] Failed on VR Search: 0032 1033
12.04.2008 14:05:56 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:56 [CONQUESTSRV] Failed on VR Search: 0032 1064
12.04.2008 14:05:56 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:56 [CONQUESTSRV] Failed on VR Search: 0038 0010
12.04.2008 14:05:56 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:56 [CONQUESTSRV] Failed on VR Search: 0038 0011
12.04.2008 14:05:56 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:56 [CONQUESTSRV] Failed on VR Search: 0038 0050
12.04.2008 14:05:56 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:56 [CONQUESTSRV] Failed on VR Search: 0038 0300
12.04.2008 14:05:56 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:56 [CONQUESTSRV] Failed on VR Search: 0038 0500
12.04.2008 14:05:56 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:56 [CONQUESTSRV] Failed on VR Search: 0040 0004
12.04.2008 14:05:56 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:56 [CONQUESTSRV] Failed on VR Search: 0040 0005
12.04.2008 14:05:56 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:56 [CONQUESTSRV] Failed on VR Search: 0040 0008
12.04.2008 14:05:56 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:56 [CONQUESTSRV] Failed on VR Search: 0040 0020
12.04.2008 14:05:56 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:56 [CONQUESTSRV] Failed on VR Search: 0040 1002
12.04.2008 14:05:56 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:56 [CONQUESTSRV] Failed on VR Search: 0040 1004
12.04.2008 14:05:56 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:56 [CONQUESTSRV] Failed on VR Search: 0040 1005
12.04.2008 14:05:56 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:56 [CONQUESTSRV] Failed on VR Search: 0040 1008
12.04.2008 14:05:56 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:56 [CONQUESTSRV] Failed on VR Search: 0040 1009
12.04.2008 14:05:56 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:56 [CONQUESTSRV] Failed on VR Search: 0040 1010
12.04.2008 14:05:56 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:56 [CONQUESTSRV] Failed on VR Search: 0040 1400
12.04.2008 14:05:56 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:56 [CONQUESTSRV] Failed on VR Search: 0040 2001
12.04.2008 14:05:56 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:56 [CONQUESTSRV] Failed on VR Search: 0040 2004
12.04.2008 14:05:56 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:56 [CONQUESTSRV] Failed on VR Search: 0040 2005
12.04.2008 14:05:56 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:56 [CONQUESTSRV] Failed on VR Search: 0040 2008
12.04.2008 14:05:56 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:56 [CONQUESTSRV] Failed on VR Search: 0040 2009
12.04.2008 14:05:56 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:56 [CONQUESTSRV] Failed on VR Search: 0040 2010
12.04.2008 14:05:56 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:56 [CONQUESTSRV] Failed on VR Search: 0040 2016
12.04.2008 14:05:56 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:56 [CONQUESTSRV] Failed on VR Search: 0040 2017
12.04.2008 14:05:56 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:56 [CONQUESTSRV] Failed on VR Search: 0040 2400
12.04.2008 14:05:56 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:56 [CONQUESTSRV] Failed on VR Search: 0040 3001
12.04.2008 14:05:56 [CONQUESTSRV] (testing phase) - ignored
12.04.2008 14:05:56 [CONQUESTSRV] Issue Query on Columns: DICOMWorkList.AccessionN, DICOMWorkList.PatientNam, DICOMWorkList.PatientID, DICOMWorkList.PatientBir, DICOMWorkList.PatientSex, DICOMWorkList.MedicalAle, DICOMWorkList.ContrastAl, DICOMWorkList.StudyInsta, DICOMWorkList.ReqPhysici, DICOMWorkList.ReqProcDes, DICOMWorkList.Modality, DICOMWorkList.ReqContras, DICOMWorkList.ScheduledA, DICOMWorkList.StartDate, DICOMWorkList.StartTime, DICOMWorkList.PerfPhysic, DICOMWorkList.SchedPSDes, DICOMWorkList.SchedPSID, DICOMWorkList.SchedStati, DICOMWorkList.SchedPSLoc, DICOMWorkList.PreMedicat, DICOMWorkList.SchedPSCom, DICOMWorkList.ReqProcID, DICOMWorkList.ReqProcPri
12.04.2008 14:05:56 [CONQUESTSRV] Values: DICOMWorkList.PatientNam LIKE 'Z%'
12.04.2008 14:05:56 [CONQUESTSRV] Tables: DICOMWorkList
12.04.2008 14:05:56 [CONQUESTSRV] Sorting (DICOMWorkList.PatientNam) DoSort := 1
12.04.2008 14:05:56 [CONQUESTSRV] Coding sequence (0040,0100), level=1
12.04.2008 14:05:56 [CONQUESTSRV] Records = 1
12.04.2008 14:05:56 [CONQUESTSRV] C-Find (Modality Work List) located 1 records
12.04.2008 14:05:56 [CONQUESTSRV] UPACS THREAD 6: ENDED AT: Sat Apr 12 14:05:56 2008
12.04.2008 14:05:56 [CONQUESTSRV] UPACS THREAD 6: TOTAL RUNNING TIME: 1 SECONDS