    Hi,


    I've been setting correctly but does not work well in Carestream CRVita ImageSuite. Though setting PACS works well. For comparation with CT99. This is log :


    Thank's



    [CONQUESTSRV2]
    [CONQUESTSRV2] UPACS THREAD 39203: STARTED AT: Thu Nov 20 12:54:24 2014
    [CONQUESTSRV2] A-ASSOCIATE-RQ Packet Dump
    [CONQUESTSRV2] Calling Application Title : "ImageSuite "
    [CONQUESTSRV2] Called Application Title : "conquestsrv2 "
    [CONQUESTSRV2] Application Context : "1.2.840.10008.3.1.1.1", PDU length: 65536
    [CONQUESTSRV2] Number of Proposed Presentation Contexts: 1
    [CONQUESTSRV2] Presentation Context 0 "1.2.840.10008.5.1.4.31" 1
    [CONQUESTSRV2] Server Command := 0020
    [CONQUESTSRV2] Message ID := 0000
    [CONQUESTSRV2] 0000,0002 22 UI AffectedSOPClassUID "1.2.840.10008.5.1.4.31"
    [CONQUESTSRV2] 0000,0100 2 US CommandField 32
    [CONQUESTSRV2] 0000,0110 2 US MessageID 0
    [CONQUESTSRV2] 0000,0700 2 US Priority 0
    [CONQUESTSRV2] 0000,0800 2 US DataSetType 258
    [CONQUESTSRV2] 0002,0010 17 UI TransferSyntaxUID "1.2.840.10008.1.2"
    [CONQUESTSRV2] (ModalityWorkListQuery) search level:
    [CONQUESTSRV2] 0002,0010 17 UI TransferSyntaxUID "1.2.840.10008.1.2"
    [CONQUESTSRV2] 0008,0050 0 SH AccessionNumber (empty)
    [CONQUESTSRV2] 0008,0090 0 PN ReferringPhysicianNa (empty)
    [CONQUESTSRV2] 0010,0010 0 PN PatientName (empty)
    [CONQUESTSRV2] 0010,0020 0 LO PatientID (empty)
    [CONQUESTSRV2] 0010,0030 0 DA PatientBirthDate (empty)
    [CONQUESTSRV2] 0010,0040 0 CS PatientSex (empty)
    [CONQUESTSRV2] 0010,1000 0 LO OtherPatientIDs (empty)
    [CONQUESTSRV2] 0020,000d 0 UI StudyInstanceUID (empty)
    [CONQUESTSRV2] 0032,1060 0 LO RequestedProcedureDe (empty)
    [CONQUESTSRV2] 0032,1064 0 SQ RequestedProcedureCo Sequence Item
    [CONQUESTSRV2] >0008,0100 0 SH CodeValue (empty)
    [CONQUESTSRV2] >0008,0104 0 LO CodeMeaning (empty)
    [CONQUESTSRV2] 0038,0008 0 CS VisitStatusID (empty)
    [CONQUESTSRV2] 0038,0300 0 LO CurrentPatientLocati (empty)
    [CONQUESTSRV2] 0040,0100 0 SQ ScheduledProcedureSt Sequence Item
    [CONQUESTSRV2] >0008,0060 0 CS Modality (empty)
    [CONQUESTSRV2] >0032,1070 0 LO RequestedContrastAge (empty)
    [CONQUESTSRV2] >0040,0002 18 DA ScheduledProcedureSt "20141120-20141120 "
    [CONQUESTSRV2] >0040,0003 14 TM ScheduledProcedureSt "115338-125338 "
    [CONQUESTSRV2] >0040,0004 0 DA ScheduledProcedureSt (empty)
    [CONQUESTSRV2] >0040,0008 0 SQ ScheduledProtocolCod Sequence Item
    [CONQUESTSRV2] >>0008,0100 0 SH CodeValue (empty)
    [CONQUESTSRV2] >0040,0009 0 SH ScheduledProcedureSt (empty)
    [CONQUESTSRV2] 0040,1001 0 SH RequestedProcedureID (empty)
    [CONQUESTSRV2] 0040,1005 0 LO RequestedProcedureLo (empty)
    [CONQUESTSRV2] Query On Modality WorkList
    [CONQUESTSRV2] Queried item 0008 0090 is not in the database
    [CONQUESTSRV2] Queried item 0010 1000 is not in the database
    [CONQUESTSRV2] Queried item 0008 0100 is not in the database
    [CONQUESTSRV2] Queried item 0008 0104 is not in the database
    [CONQUESTSRV2] Queried item 0038 0008 is not in the database
    [CONQUESTSRV2] Queried item 0038 0300 is not in the database
    [CONQUESTSRV2] Queried item 0040 0004 is not in the database
    [CONQUESTSRV2] Queried item 0008 0100 is not in the database
    [CONQUESTSRV2] Queried item 0040 1005 is not in the database
    [CONQUESTSRV2] Issue Query on Columns: DICOMWorkList.AccessionN, DICOMWorkList.PatientNam, DICOMWorkList.PatientID, DICOMWorkList.PatientBir, DICOMWorkList.PatientSex, DICOMWorkList.StudyInsta, DICOMWorkList.ReqProcDes, DICOMWorkList.Modality, DICOMWorkList.ReqContras, DICOMWorkList.StartDate, DICOMWorkList.StartTime, DICOMWorkList.SchedPSID, DICOMWorkList.ReqProcID
    [CONQUESTSRV2] Values: DICOMWorkList.StartDate >= '20141120' and DICOMWorkList.StartDate <= '20141120' and DICOMWorkList.StartTime >= '115338' and DICOMWorkList.StartTime <= '125338'
    [CONQUESTSRV2] Tables: DICOMWorkList
    [CONQUESTSRV2] Sorting (DICOMWorkList.PatientNam) DoSort := 1
    [CONQUESTSRV2] Query Distinct Tables: DICOMWorkList
    [CONQUESTSRV2] Columns : DICOMWorkList.AccessionN, DICOMWorkList.PatientNam, DICOMWorkList.PatientID, DICOMWorkList.PatientBir, DICOMWorkList.PatientSex, DICOMWorkList.StudyInsta, DICOMWorkList.ReqProcDes, DICOMWorkList.Modality, DICOMWorkList.ReqContras, DICOMWorkList.StartDate, DICOMWorkList.StartTime, DICOMWorkList.SchedPSID, DICOMWorkList.ReqProcID
    [CONQUESTSRV2] Where : DICOMWorkList.StartDate >= '20141120' and DICOMWorkList.StartDate <= '20141120' and DICOMWorkList.StartTime >= '115338' and DICOMWorkList.StartTime <= '125338'
    [CONQUESTSRV2] Order : DICOMWorkList.PatientNam
    [CONQUESTSRV2] Records = 0
    [CONQUESTSRV2] C-Find (Modality Work List) located 0 records
    [CONQUESTSRV2] UPACS THREAD 39203: ENDED AT: Thu Nov 20 12:54:25 2014
    [CONQUESTSRV2] UPACS THREAD 39203: TOTAL RUNNING TIME: 1 SECONDS



    [CONQUESTSRV2] Connected by address: 0f02a8c0
    [CONQUESTSRV2] Testing transfer: '1.2.840.10008.1.2' against list #0 = '1.2.840.10008.1.2'
    [CONQUESTSRV2]
    [CONQUESTSRV2] UPACS THREAD 39236: STARTED AT: Thu Nov 20 13:04:16 2014
    [CONQUESTSRV2] A-ASSOCIATE-RQ Packet Dump
    [CONQUESTSRV2] Calling Application Title : "ct99 "
    [CONQUESTSRV2] Called Application Title : "conquestsrv2 "
    [CONQUESTSRV2] Application Context : "1.2.840.10008.3.1.1.1", PDU length: 65464
    [CONQUESTSRV2] Number of Proposed Presentation Contexts: 1
    [CONQUESTSRV2] Presentation Context 0 "1.2.840.10008.5.1.4.31" 1
    [CONQUESTSRV2] Server Command := 0020
    [CONQUESTSRV2] Message ID := 0000
    [CONQUESTSRV2] 0000,0002 22 UI AffectedSOPClassUID "1.2.840.10008.5.1.4.31"
    [CONQUESTSRV2] 0000,0100 2 US CommandField 32
    [CONQUESTSRV2] 0000,0110 2 US MessageID 0
    [CONQUESTSRV2] 0000,0700 2 US Priority 0
    [CONQUESTSRV2] 0000,0800 2 US DataSetType 258
    [CONQUESTSRV2] 0002,0010 17 UI TransferSyntaxUID "1.2.840.10008.1.2"
    [CONQUESTSRV2] (ModalityWorkListQuery) search level:
    [CONQUESTSRV2] 0002,0010 17 UI TransferSyntaxUID "1.2.840.10008.1.2"
    [CONQUESTSRV2] 0008,0005 10 CS SpecificCharacterSet "ISO_IR 100"
    [CONQUESTSRV2] 0008,0050 0 SH AccessionNumber (empty)
    [CONQUESTSRV2] 0008,0080 0 LO InstitutionName (empty)
    [CONQUESTSRV2] 0008,0090 0 PN ReferringPhysicianNa (empty)
    [CONQUESTSRV2] 0008,1110 0 SQ ReferencedStudySeque Sequence Item
    [CONQUESTSRV2] >0008,1150 0 UI ReferencedSOPClassUI (empty)
    [CONQUESTSRV2] >0008,1155 0 UI ReferencedSOPInstanc (empty)
    [CONQUESTSRV2] 0008,1120 0 SQ ReferencedPatientSeq Sequence Item
    [CONQUESTSRV2] >0008,1150 0 UI ReferencedSOPClassUI (empty)
    [CONQUESTSRV2] >0008,1155 0 UI ReferencedSOPInstanc (empty)
    [CONQUESTSRV2] 0010,0010 0 PN PatientName (empty)
    [CONQUESTSRV2] 0010,0020 0 LO PatientID (empty)
    [CONQUESTSRV2] 0010,0030 0 DA PatientBirthDate (empty)
    [CONQUESTSRV2] 0010,0040 0 CS PatientSex (empty)
    [CONQUESTSRV2] 0010,1000 0 LO OtherPatientIDs (empty)
    [CONQUESTSRV2] 0010,1020 0 DS PatientSize (empty)
    [CONQUESTSRV2] 0010,1030 0 DS PatientWeight (empty)
    [CONQUESTSRV2] 0010,1040 0 LO PatientAddress (empty)
    [CONQUESTSRV2] 0010,2000 0 LO MedicalAlerts (empty)
    [CONQUESTSRV2] 0010,2110 0 LO Allergies (empty)
    [CONQUESTSRV2] 0010,2154 0 SH PatientTelephoneNumb (empty)
    [CONQUESTSRV2] 0010,2160 0 SH EthnicGroup (empty)
    [CONQUESTSRV2] 0010,21b0 0 LT AdditionalPatientHis (empty)
    [CONQUESTSRV2] 0010,21c0 0 US PregnancyStatus (empty)
    [CONQUESTSRV2] 0010,4000 0 LT PatientComments (empty)
    [CONQUESTSRV2] 0020,000d 0 UI StudyInstanceUID (empty)
    [CONQUESTSRV2] 0032,1032 0 PN RequestingPhysician (empty)
    [CONQUESTSRV2] 0032,1033 0 LO RequestingService (empty)
    [CONQUESTSRV2] 0032,1060 0 LO RequestedProcedureDe (empty)
    [CONQUESTSRV2] 0032,1064 0 SQ RequestedProcedureCo Sequence Item
    [CONQUESTSRV2] >0008,0100 0 SH CodeValue (empty)
    [CONQUESTSRV2] >0008,0102 0 SH CodingSchemeDesignat (empty)
    [CONQUESTSRV2] >0008,0104 0 LO CodeMeaning (empty)
    [CONQUESTSRV2] 0038,0010 0 LO AdmissionID (empty)
    [CONQUESTSRV2] 0038,0050 0 LO SpecialNeeds (empty)
    [CONQUESTSRV2] 0038,0300 0 LO CurrentPatientLocati (empty)
    [CONQUESTSRV2] 0038,0500 0 LO PatientState (empty)
    [CONQUESTSRV2] 0040,0100 0 SQ ScheduledProcedureSt Sequence Item
    [CONQUESTSRV2] >0008,0060 2 CS Modality "CT"
    [CONQUESTSRV2] >0032,1070 0 LO RequestedContrastAge (empty)
    [CONQUESTSRV2] >0040,0001 0 AE ScheduledStationAETi (empty)
    [CONQUESTSRV2] >0040,0002 18 DA ScheduledProcedureSt "20141120-20141120 "
    [CONQUESTSRV2] >0040,0003 0 TM ScheduledProcedureSt (empty)
    [CONQUESTSRV2] >0040,0006 0 PN ScheduledPerformingP (empty)
    [CONQUESTSRV2] >0040,0007 0 LO ScheduledProcedureSt (empty)
    [CONQUESTSRV2] >0040,0008 0 SQ ScheduledProtocolCod Sequence Item
    [CONQUESTSRV2] >>0008,0100 0 SH CodeValue (empty)
    [CONQUESTSRV2] >>0008,0102 0 SH CodingSchemeDesignat (empty)
    [CONQUESTSRV2] >>0008,0104 0 LO CodeMeaning (empty)
    [CONQUESTSRV2] >0040,0009 0 SH ScheduledProcedureSt (empty)
    [CONQUESTSRV2] >0040,0010 0 SH ScheduledStationName (empty)
    [CONQUESTSRV2] >0040,0011 0 SH ScheduledProcedureSt (empty)
    [CONQUESTSRV2] >0040,0012 0 LO PreMedication (empty)
    [CONQUESTSRV2] 0040,1001 0 SH RequestedProcedureID (empty)
    [CONQUESTSRV2] 0040,1003 0 SH RequestedProcedurePr (empty)
    [CONQUESTSRV2] 0040,1004 0 LO PatientTransportArra (empty)
    [CONQUESTSRV2] 0040,1005 0 LO RequestedProcedureLo (empty)
    [CONQUESTSRV2] 0040,1008 0 LO ConfidentialityCode (empty)
    [CONQUESTSRV2] 0040,1010 0 PN NamesOfIntendedRecip (empty)
    [CONQUESTSRV2] 0040,1400 0 LT RequestedProcedureCo (empty)
    [CONQUESTSRV2] 0040,2400 0 LT ImagingServiceReques (empty)
    [CONQUESTSRV2] 0040,3001 0 LO ConfidentialityConst (empty)
    [CONQUESTSRV2] Query On Modality WorkList
    [CONQUESTSRV2] Queried item 0008 0005 is not in the database
    [CONQUESTSRV2] Queried item 0008 0080 is not in the database
    [CONQUESTSRV2] Queried item 0008 0090 is not in the database
    [CONQUESTSRV2] Queried item 0008 1150 is not in the database
    [CONQUESTSRV2] Queried item 0008 1155 is not in the database
    [CONQUESTSRV2] Queried item 0008 1150 is not in the database
    [CONQUESTSRV2] Queried item 0008 1155 is not in the database
    [CONQUESTSRV2] Queried item 0010 1000 is not in the database
    [CONQUESTSRV2] Queried item 0010 1020 is not in the database
    [CONQUESTSRV2] Queried item 0010 1030 is not in the database
    [CONQUESTSRV2] Queried item 0010 1040 is not in the database
    [CONQUESTSRV2] Queried item 0010 2154 is not in the database
    [CONQUESTSRV2] Queried item 0010 2160 is not in the database
    [CONQUESTSRV2] Queried item 0010 21b0 is not in the database
    [CONQUESTSRV2] Queried item 0010 21c0 is not in the database
    [CONQUESTSRV2] Queried item 0010 4000 is not in the database
    [CONQUESTSRV2] Queried item 0032 1033 is not in the database
    [CONQUESTSRV2] Queried item 0008 0100 is not in the database
    [CONQUESTSRV2] Queried item 0008 0102 is not in the database
    [CONQUESTSRV2] Queried item 0008 0104 is not in the database
    [CONQUESTSRV2] Queried item 0038 0010 is not in the database
    [CONQUESTSRV2] Queried item 0038 0050 is not in the database
    [CONQUESTSRV2] Queried item 0038 0300 is not in the database
    [CONQUESTSRV2] Queried item 0038 0500 is not in the database
    [CONQUESTSRV2] Queried item 0008 0100 is not in the database
    [CONQUESTSRV2] Queried item 0008 0102 is not in the database
    [CONQUESTSRV2] Queried item 0008 0104 is not in the database
    [CONQUESTSRV2] Queried item 0040 1004 is not in the database
    [CONQUESTSRV2] Queried item 0040 1005 is not in the database
    [CONQUESTSRV2] Queried item 0040 1008 is not in the database
    [CONQUESTSRV2] Queried item 0040 1010 is not in the database
    [CONQUESTSRV2] Queried item 0040 1400 is not in the database
    [CONQUESTSRV2] Queried item 0040 2400 is not in the database
    [CONQUESTSRV2] Queried item 0040 3001 is not in the database
    [CONQUESTSRV2] Issue Query on Columns: DICOMWorkList.AccessionN, DICOMWorkList.PatientNam, DICOMWorkList.PatientID, DICOMWorkList.PatientBir, DICOMWorkList.PatientSex, DICOMWorkList.MedicalAle, DICOMWorkList.ContrastAl, DICOMWorkList.StudyInsta, DICOMWorkList.ReqPhysici, DICOMWorkList.ReqProcDes, DICOMWorkList.Modality, DICOMWorkList.ReqContras, DICOMWorkList.ScheduledA, DICOMWorkList.StartDate, DICOMWorkList.StartTime, DICOMWorkList.PerfPhysic, DICOMWorkList.SchedPSDes, DICOMWorkList.SchedPSID, DICOMWorkList.SchedStati, DICOMWorkList.SchedPSLoc, DICOMWorkList.PreMedicat, DICOMWorkList.ReqProcID, DICOMWorkList.ReqProcPri
    [CONQUESTSRV2] Values: DICOMWorkList.Modality = 'CT' and DICOMWorkList.StartDate >= '20141120' and DICOMWorkList.StartDate <= '20141120'
    [CONQUESTSRV2] Tables: DICOMWorkList
    [CONQUESTSRV2] Sorting (DICOMWorkList.PatientNam) DoSort := 1
    [CONQUESTSRV2] Query Distinct Tables: DICOMWorkList
    [CONQUESTSRV2] Columns : DICOMWorkList.AccessionN, DICOMWorkList.PatientNam, DICOMWorkList.PatientID, DICOMWorkList.PatientBir, DICOMWorkList.PatientSex, DICOMWorkList.MedicalAle, DICOMWorkList.ContrastAl, DICOMWorkList.StudyInsta, DICOMWorkList.ReqPhysici, DICOMWorkList.ReqProcDes, DICOMWorkList.Modality, DICOMWorkList.ReqContras, DICOMWorkList.ScheduledA, DICOMWorkList.StartDate, DICOMWorkList.StartTime, DICOMWorkList.PerfPhysic, DICOMWorkList.SchedPSDes, DICOMWorkList.SchedPSID, DICOMWorkList.SchedStati, DICOMWorkList.SchedPSLoc, DICOMWorkList.PreMedicat, DICOMWorkList.ReqProcID, DICOMWorkList.ReqProcPri
    [CONQUESTSRV2] Where : DICOMWorkList.Modality = 'CT' and DICOMWorkList.StartDate >= '20141120' and DICOMWorkList.StartDate <= '20141120'
    [CONQUESTSRV2] Order : DICOMWorkList.PatientNam
    [CONQUESTSRV2] 0008,0050 10 SH AccessionNumber "1411108037"
    [CONQUESTSRV2] Coding sequence (0008,1110), level=1
    [CONQUESTSRV2] (sequence is empty)
    [CONQUESTSRV2] Coding sequence (0008,1120), level=1
    [CONQUESTSRV2] (sequence is empty)
    [CONQUESTSRV2] 0010,0010 6 PN PatientName "XXXXXX"
    [CONQUESTSRV2] 0010,0020 16 LO PatientID "1411108037C0095 "
    [CONQUESTSRV2] 0010,0030 8 DA PatientBirthDate "19630209"
    [CONQUESTSRV2] 0010,0040 2 CS PatientSex "M "
    [CONQUESTSRV2] 0010,2000 0 LO MedicalAlerts (empty)
    [CONQUESTSRV2] 0010,2110 0 LO Allergies (empty)
    [CONQUESTSRV2] 0020,000d 0 UI StudyInstanceUID (empty)
    [CONQUESTSRV2] First record of cleaned response:
    [CONQUESTSRV2] UPACS THREAD 39236: ENDED AT: Thu Nov 20 13:04:16 2014
    [CONQUESTSRV2] UPACS THREAD 39236: TOTAL RUNNING TIME: 0 SECONDS

    Hi,



    StudyDate QueryRetrieveLevel PatientName PatientID
    ---------------------------------------------------------------------------------------------------------------------------------------------------------------
    ---------------------------------------------------------------------------------------------------------------------------------------------------------------
    Total of 0 item(s) found


    not found but there in system AW46


    if DICOM system : CONQUEST2


    StudyDate QueryRetrieveLevel PatientName PatientID
    ---------------------------------------------------------------------------------------------------------------------------------------------------------------
    [20140612] [STUDY ] [RUJA] [1406118017C0069 ]
    ---------------------------------------------------------------------------------------------------------------------------------------------------------------
    Total of 1 item(s) found



    my system :
    1. GE MRI BRIVO MR355
    2. GE OPTIMA 660
    3. Conquest1417d

    hi,


    1. how commandline --patientfinder and studyfinder ? ex: srv=AWCT,
    search=*maman*
    2. sample commandline all Database options ?
    3. not responding if use Query/Move CONQUEST DICOM server
    QUERY LEVEL : PATIENT AND DICOM SYSTEM : AWCT
    but if use DICOM SYSTEM : CONQUEST good work.


    ------------ working --------------
    *** DICOM query error (patient ID = 1402049008C0011 ) : Association lost
    ---------------------------------------------------------------------------------------------------------------------------------------------------------------
    Total of -1 item(s) found





    mrudiansyah

    Hi,


    [sscscp]
    MicroPACS = sscscp


    Tempdir = c:\temp


    # database layout (copy dicom.sql to the web server script directory or point to the one in your dicom server directory)


    kFactorFile = c:\dicomserver\dicom.sql
    TruncateFieldNames = 10


    # gives access to the SQL server of the DICOM server
    # use of independent database is also allowed (depends on scripts used)


    SQLServer = c:\dicomserver\data\dbase\conquest.db3
    SQlite = 1


    # gives access to all DICOM servers known in acrnema.map


    ACRNemaMap = C:\dicomserver\acrnema.map
    Dictionary = C:\dicomserver\dgate.dic


    # default IP address and port of DICOM server (may be non-local, web pages empty if wrong)


    WebServerFor = 127.0.0.1
    TCPPort = 5678


    # AE title: only used if web client originates queries or moves


    MyACRNema = CONQUESTSRV1


    # path to script engine: ocx will not download images if wrong - shows as black square with controls
    # for wamp: dgate.exe runs if it is put in C:\wamp\Apache2\cgi-bin


    WebScriptAddress = http://127.0.0.1/cgi-bin/dgate.exe


    # web or local location of ActiveFormProj1.ocx for download (include trailing / or \)
    # the activeX control will not download if wrong or security too high - shows as white square with red x
    # note: it only needs to be registered by the client, not the server!
    # for wamp: the ocx canNOT be in C:\wamp\Apache2\cgi-bin, I put it in c:\wamp\www (above cgi-bin)
    # - the default value is derived from WebScriptAddress


    #WebCodeBase = http://127.0.0.1/



    # if set to 1 (default), the web user cannot edit databases and (in future) other things
    # webpush enables push of data to other servers


    WebReadonly = 0
    WebPush = 1


    # this is an optional virtual directory used to http: all images from mag0
    # this entry is experimental and unused except for viewer=seriesviewer2
    # in this mode (only) ocx will not download images if wrong - shows as black square with
    # controls - the default value is derived from WebScriptAddress


    #WebMAG0Address = http://127.0.0.1/mag0


    # excerpt from C:\wamp\Apache2\conf\httpd.conf required for WebMAG0Address (un-# there)
    # or use the wamp traybar menu to create the alias


    #Alias /mag0/ "c:/dicomserver/data/"
    #
    #<Directory "c:/dicomserver/data">
    # Options Indexes MultiViews
    # AllowOverride None
    # Order allow,deny
    # Allow from all
    #</Directory>


    # these settings control size of slice and series viewers, max size of transmitted dicom images
    # (0=original), compression for images sent to the activex (may be un,n1..4,j1,j2), the size of
    # the icons in the image list, the image type used for icons and slice display, and the dgate
    # mode containing the viewer (may be seriesviewer, seriesviewer2, noviewer, serversideviewer,
    # or aiviewer - java code of the latter not included with 1.4.14).
    # note: all items require at least one space left and right of the '=' !


    [webdefaults]
    size = 560
    dsize = 0
    compress = un
    iconsize = 48
    graphic = gif
    viewer = serversideviewer
    studyviewer = dwv


    # enter address (up to not including the ?) of the WADO server for each DICOM AE listed
    # the sample (for AE TESTWADOSRV) comes from the DICOM standard and is not valid
    # the default is the local conquest server (which could use virtualservers as WADO bridge)
    # These WADO entries are used in 'list .... on ....'


    [wadoservers]
    TESTWADOSRV = http://www.hospital-stmarco/radiology/wado.php


    # Here are scripted series viewers: ImageJA, dwv, flexviewer (=aiviewer), weasis
    # And one scripted study viewer: studyweasis


    [imagejaviewer]
    source = viewers\imagejaviewer.cq


    [dwv]
    source = viewers\dwv.lua


    [flexviewer]
    source = viewers\flexviewer.lua


    [weasis]
    source = viewers\LaunchWeasis.cq
    header = Content-Type: application/x-java-jnlp-file\


    [studyweasis]
    source = viewers\LaunchWeasisStudy.cq
    header = Content-Type: application/x-java-jnlp-file\


    [weasisseriesxml]
    source = viewers\weasisseriesxml.lua


    [weasisstudyxml]
    source = viewers\weasisstudyxml.lua


    # These entries enable the OpenClinica - Conquest interface


    # default
    [DefaultPage]
    source = *.lua


    [OpenClinica]


    Source=(local)
    TargetServer=user@server.domain:
    password=xxxxx


    # Below here are all samples with no clear functionality


    # Here are some sample web pages to show how scripting works
    # this creates web page http://xxxxx/cgi-bin/dgate.exe?mode=sample


    [sample]
    variable = sample 1
    source = samples\sample.cq


    # this creates web page http://xxxxx/cgi-bin/dgate.exe?mode=sample2
    # This sample posts a file
    # Note: in the header parameter newline is written as \


    [sample2]
    variable = sample 2
    header= Content-type: text/html\Cache-Control: no-cache\
    line0 = <HEAD><TITLE>Conquest DICOM server - %variable%</TITLE></HEAD>
    line1 = <BODY BGCOLOR='CFDFCF'>
    line2 = <H2>Conquest DICOM server - %query_string%</H2>
    line3 = <FORM ACTION="dgate.exe" METHOD=POST ENCTYPE="multipart/form-data">
    line4 = <INPUT NAME=mode TYPE=HIDDEN VALUE=soaprequest>
    line5 = <INPUT NAME=port TYPE=HIDDEN VALUE=5678>
    line6 = <INPUT NAME=address TYPE=HIDDEN VALUE=127.0.0.1>
    line7 = Upload file to enter into server (dcm/v2/HL7/zip/7z/gz/tar): <INPUT NAME=filetoupload SIZE=40 TYPE=file VALUE=>
    line8 = <INPUT TYPE=SUBMIT VALUE=Go>
    line9 = </FORM>
    line10 = </BODY>


    # this creates a web page scripted in lua http://xxxxx/cgi-bin/dgate.exe?mode=sample3


    #[sample3]
    #source = samples\sample3.lua


    # these entries create an experimental SOAP interface scripted in lua


    [wsdl]
    source = soap\wsdl.xml
    header = Content-type: text/xml\


    [soaprequest]
    source = soap\soaprequest.lua


    # these entries create an experimental JSON interface scripted in lua


    [sample4]
    source = json\json_sample.html
    header = Content-type: text/html\Cache-Control: no-cache\


    [jsonrequest]
    source = json\jsonrequest.lua


    # these entries create a simple eCRF interface
    # ..?mode=studyfinder&dest=CONQUESTSRV1&key=afc0501:baseline&query==2040XXXX


    # creates SQL database and enters information
    [markstudy]
    source = ecrf\markstudyseries.lua
    caption= Select for submission


    [markseries]
    source = ecrf\markstudyseries.lua
    caption= Select for submission


    # show SQL database and selects processing
    [shoppingcart]
    source = ecrf\shoppingcart.lua
    caption= Process selected data


    # process information
    [ecrfprocess]
    source = ecrf\ecrfprocess.lua


    # new query form
    [queryforecrf]
    source = ecrf\queryforecrf.lua
















    mrudiansyah

    hi Marcell


    1. How to use DVW?
    2. I've done the same with the tutorial. But there is an error
    RequestError : An error accured while retrieving the file: (http)0.


    debug CONQUESTSRV1
    [CONQUESTSRV1]
    [CONQUESTSRV1] UPACS THREAD 7359: STARTED AT: Fri Apr 04 00:46:08 2014
    [CONQUESTSRV1] Calling Application Title : "CONQUESTSRV1 "
    [CONQUESTSRV1] Called Application Title : "CONQUESTSRV1 "
    [CONQUESTSRV1] Application Context : "1.2.840.10008.3.1.1.1", PDU length: 16384
    [CONQUESTSRV1] Presentation Context 0 "1.2.840.10008.5.1.4.1.2.2.1" 1
    [CONQUESTSRV1] (StudyRootQuery) search level: SERIES
    [CONQUESTSRV1] C-Find (StudyRoot) located 162 records
    [CONQUESTSRV1] UPACS THREAD 7359: ENDED AT: Fri Apr 04 00:46:08 2014
    [CONQUESTSRV1] UPACS THREAD 7359: TOTAL RUNNING TIME: 0 SECONDS
    [CONQUESTSRV1]
    [CONQUESTSRV1] UPACS THREAD 7361: STARTED AT: Fri Apr 04 00:46:08 2014
    [CONQUESTSRV1] Calling Application Title : "CONQUESTSRV1 "
    [CONQUESTSRV1] Called Application Title : "CONQUESTSRV1 "
    [CONQUESTSRV1] Application Context : "1.2.840.10008.3.1.1.1", PDU length: 16384
    [CONQUESTSRV1] Presentation Context 0 "1.2.840.10008.5.1.4.1.2.2.1" 1
    [CONQUESTSRV1] (StudyRootQuery) search level: IMAGE
    [CONQUESTSRV1] C-Find (StudyRoot) located 3270 records
    [CONQUESTSRV1] UPACS THREAD 7361: ENDED AT: Fri Apr 04 00:46:08 2014
    [CONQUESTSRV1] UPACS THREAD 7361: TOTAL RUNNING TIME: 0 SECONDS


    I Use Chrome in Samsung Galaxy Tab 10.1"


    3. if Mozilla in PC no image. But view DVW. Debug CONQUESTSRV1

    [CONQUESTSRV1]
    [CONQUESTSRV1] UPACS THREAD 4088: STARTED AT: Fri Apr 04 00:43:53 2014
    [CONQUESTSRV1] Calling Application Title : "CONQUESTSRV1 "
    [CONQUESTSRV1] Called Application Title : "CONQUESTSRV1 "
    [CONQUESTSRV1] Application Context : "1.2.840.10008.3.1.1.1", PDU length: 16384
    [CONQUESTSRV1] Presentation Context 0 "1.2.840.10008.5.1.4.1.2.2.1" 1
    [CONQUESTSRV1] (StudyRootQuery) search level: SERIES
    [CONQUESTSRV1] C-Find (StudyRoot) located 162 records
    [CONQUESTSRV1] UPACS THREAD 4088: ENDED AT: Fri Apr 04 00:43:53 2014
    [CONQUESTSRV1] UPACS THREAD 4088: TOTAL RUNNING TIME: 0 SECONDS
    [CONQUESTSRV1]
    [CONQUESTSRV1] UPACS THREAD 4090: STARTED AT: Fri Apr 04 00:43:53 2014
    [CONQUESTSRV1] Calling Application Title : "CONQUESTSRV1 "
    [CONQUESTSRV1] Called Application Title : "CONQUESTSRV1 "
    [CONQUESTSRV1] Application Context : "1.2.840.10008.3.1.1.1", PDU length: 16384
    [CONQUESTSRV1] Presentation Context 0 "1.2.840.10008.5.1.4.1.2.2.1" 1
    [CONQUESTSRV1] (StudyRootQuery) search level: IMAGE
    [CONQUESTSRV1] C-Find (StudyRoot) located 3270 records
    [CONQUESTSRV1] UPACS THREAD 4090: ENDED AT: Fri Apr 04 00:43:54 2014
    [CONQUESTSRV1] UPACS THREAD 4090: TOTAL RUNNING TIME: 1 SECONDS



    mrudiansyah

    hi,


    dgate -pport -qip --movepatient:CONQUESTSRV2,CONQUESTSRV1,R1009
    Move failed from CONQUESTSRV2 to CONQUESTSRV1


    No success. by using Hamachi. Computer server in office and computer in home as if local network. but upload and download DICOM slow....
    is there a better way bisides use hamachi.


    thanks


    google translate........


    picture :



    conquestsrv1 (no ip public) ----------> ip public ----- port forward to ----------- > server in office (ip local)

    hi Marcell,


    why I could not take pictures DICOM from GUI or command. Example :


    c:\dicomserver>dgate --movepatient:CONQUESTSRV2,CONQUESTSRV1,R1009
    Move failed from CONQUESTSRV2 to CONQUESTSRV1
    c:\dicomserver>


    but
    c:\dicomserver>dgate --movepatient:CONQUESTSRV1,CONQUESTSRV2,R1009
    c:\dicomserver>
    success


    status :
    CONQUESTSRV1 is server in home and no ip public
    CONQUESTSRV2 is server in office, use ip public (ip:x.x.x.x) router mikrotik forward to server CONQUESTSRV2 (ip:y.y.y.y)


    query CONQUESTSRV2 success.


    but --movepatient no success



    Thanks

    Hi,


    1. I have edited :
    1. file dgatesop.lst :


    ............
    ............
    ............
    #None none RemoteAE
    #None none LocalAE
    #DICOM 1.2.840.10008.3.1.1.1 application
    Brivo MR355 1.2.840.113619.6.275 sop
    Optima CT660 1.2.840.113619.6.289 sop
    Verification 1.2.840.10008.1.1 sop
    StoredPrintStorage 1.2.840.10008.5.1.1.27 sop
    ..............
    ..............
    ..............


    and dicom.sql


    { 0x0020, 0x000d, "StudyInstanceUID", 64, SQL_C_CHAR, DT_UI, "---" },
    { 0x0032, 0x1032, "ReqPhysician", 64, SQL_C_CHAR, DT_STR, "OBR.16" },
    { 0x0008, 0x0090, "ReferringPhysicianName",64, SQL_C_CHAR, DT_STR, "OBR.17" },
    { 0x0008, 0x1060, "NameOfPhysicianReadingStudy",64, SQL_C_CHAR, DT_STR, "OBR.18" },
    { 0x0032, 0x1060, "ReqProcDescription", 64, SQL_C_CHAR, DT_STR, "OBR.4.1" },


    but, 0x0008, 0x0090, "ReferringPhysicianName" and 0x0008, 0x1060, "NameOfPhysicianReadingStudy" can not send to modality worklist.


    2. Problem Query : I plug in the computer 3 PACS. K-PACS, Lead Tools, and Conquest. K-PACS and Lead Tools Query successful, but could not Conquest ?



    Terima kasih (English : thank you) he....he......


    Rudiansyah

    there is another difficulty. the tab Query / Move. if the query has its own success. but if the other modality elalu error:


    ------------ working --------------
    *** DICOM query error (patient ID = ) : Unable to connect
    ---------------------------------------------------------------------------------------------------------------------------------------------------------------
    Total of -1 item(s) found



    and there is no message in the Debug Server status.
    Thank's

    [SINMIR]
    [SINMIR] UPACS THREAD 37: STARTED AT: Mon Jul 15 10:56:27 2013
    [SINMIR] A-ASSOCIATE-RQ Packet Dump
    [SINMIR] Calling Application Title : "RSKM4000942 "
    [SINMIR] Called Application Title : "SINMIR "
    [SINMIR] Application Context : "1.2.840.10008.3.1.1.1", PDU length: 131072
    [SINMIR] Number of Proposed Presentation Contexts: 1
    [SINMIR] Presentation Context 0 "1.2.840.10008.5.1.4.31" 1
    [SINMIR] Server Command := 0020
    [SINMIR] Message ID := 0000
    [SINMIR] 0000,0002 22 UI AffectedSOPClassUID "1.2.840.10008.5.1.4.31"
    [SINMIR] 0000,0100 2 US CommandField 32
    [SINMIR] 0000,0110 2 US MessageID 0
    [SINMIR] 0000,0700 2 US Priority 0
    [SINMIR] 0000,0800 2 US DataSetType 258
    [SINMIR] 0002,0010 17 UI TransferSyntaxUID "1.2.840.10008.1.2"
    [SINMIR] (ModalityWorkListQuery) search level:
    [SINMIR] Query On Modality WorkList
    [SINMIR] Queried item 0008 0005 is not in the database
    [SINMIR] Queried item 0008 0090 is not in the database
    [SINMIR] Queried item 0008 1080 is not in the database
    [SINMIR] Queried item 0008 1150 is not in the database
    [SINMIR] Queried item 0008 1155 is not in the database
    [SINMIR] Queried item 0010 0032 is not in the database
    [SINMIR] Queried item 0010 1000 is not in the database
    [SINMIR] Queried item 0010 1001 is not in the database
    [SINMIR] Queried item 0010 1010 is not in the database
    [SINMIR] Queried item 0010 1020 is not in the database
    [SINMIR] Queried item 0010 1030 is not in the database
    [SINMIR] Queried item 0010 2160 is not in the database
    [SINMIR] Queried item 0010 2180 is not in the database
    [SINMIR] Queried item 0010 21b0 is not in the database
    [SINMIR] Queried item 0010 21c0 is not in the database
    [SINMIR] Queried item 0010 4000 is not in the database
    [SINMIR] Queried item 0032 1033 is not in the database
    [SINMIR] Queried item 0008 0100 is not in the database
    [SINMIR] Queried item 0008 0102 is not in the database
    [SINMIR] Queried item 0008 0103 is not in the database
    [SINMIR] Queried item 0008 0104 is not in the database
    [SINMIR] Queried item 0038 0008 is not in the database
    [SINMIR] Queried item 0038 0400 is not in the database
    [SINMIR] Queried item 0008 0100 is not in the database
    [SINMIR] Queried item 0008 0102 is not in the database
    [SINMIR] Queried item 0008 0103 is not in the database
    [SINMIR] Queried item 0008 0104 is not in the database
    [SINMIR] Issue Query on Columns: DICOMWorkList.AccessionN, DICOMWorkList.PatientNam, DICOMWorkList.PatientID, DICOMWorkList.PatientBir, DICOMWorkList.PatientSex, DICOMWorkList.MedicalAle, DICOMWorkList.StudyInsta, DICOMWorkList.ReqProcDes, DICOMWorkList.Modality, DICOMWorkList.ReqContras, DICOMWorkList.ScheduledA, DICOMWorkList.StartDate, DICOMWorkList.StartTime, DICOMWorkList.SchedPSDes, DICOMWorkList.SchedPSID, DICOMWorkList.SchedStati, DICOMWorkList.ReqProcID, DICOMWorkList.ReqProcPri
    [SINMIR] Values: DICOMWorkList.PatientNam LIKE '%' and DICOMWorkList.StartDate = '20130715'
    [SINMIR] Tables: DICOMWorkList
    [SINMIR] Sorting (DICOMWorkList.PatientNam) DoSort := 1
    [SINMIR] Records = 0
    [SINMIR] C-Find (Modality Work List) located 0 records
    [SINMIR] UPACS THREAD 37: ENDED AT: Mon Jul 15 10:56:27 2013
    [SINMIR] UPACS THREAD 37: TOTAL RUNNING TIME: 0 SECONDS

    log :
    [SINMIR] UPACS THREAD 42: STARTED AT: Sun Jul 14 13:40:38 2013
    [SINMIR] Calling Application Title : "RSKM4000942 "
    [SINMIR] Called Application Title : "SINMIR "
    [SINMIR] Application Context : "1.2.840.10008.3.1.1.1", PDU length: 131072
    [SINMIR] Presentation Context 0 "1.2.840.10008.5.1.4.31" 1
    [SINMIR] (ModalityWorkListQuery) search level:
    [SINMIR] C-Find (Modality Work List) located 0 records
    [SINMIR] UPACS THREAD 42: ENDED AT: Sun Jul 14 13:40:39 2013
    [SINMIR] UPACS THREAD 42: TOTAL RUNNING TIME: 1 SECONDS
    [SINMIR]
    [SINMIR] UPACS THREAD 50: STARTED AT: Sun Jul 14 13:42:46 2013
    [SINMIR] Calling Application Title : "GEMRGEMR "
    [SINMIR] Called Application Title : "SINMIR "
    [SINMIR] Application Context : "1.2.840.10008.3.1.1.1", PDU length: 16352
    [SINMIR] Presentation Context 0 "1.2.840.10008.5.1.4.31" 1
    [SINMIR] (ModalityWorkListQuery) search level:
    [SINMIR] C-Find (Modality Work List) located 0 records
    [SINMIR] UPACS THREAD 50: ENDED AT: Sun Jul 14 13:42:47 2013
    [SINMIR] UPACS THREAD 50: TOTAL RUNNING TIME: 1 SECONDS
    [SINMIR]
    [SINMIR] UPACS THREAD 51: STARTED AT: Sun Jul 14 13:44:39 2013
    [SINMIR] Calling Application Title : "ct99 "
    [SINMIR] Called Application Title : "SINMIR "
    [SINMIR] Application Context : "1.2.840.10008.3.1.1.1", PDU length: 65464
    [SINMIR] Presentation Context 0 "1.2.840.10008.5.1.4.31" 1
    [SINMIR] (ModalityWorkListQuery) search level:
    [SINMIR] C-Find (Modality Work List) located 1 records
    [SINMIR] UPACS THREAD 51: ENDED AT: Sun Jul 14 13:44:39 2013
    [SINMIR] UPACS THREAD 51: TOTAL RUNNING TIME: 0 SECONDS


    I push the data directly 3 record into MYSQL table dicomworklist Conquest. Are distinguished only in each modality. "CT", "CR", "MR". But still the only entry in the GE Optima 660 CT. Thank you for replying to this question.

    Hi,