Hi,
Device is the GE Sono Vivit T8
Worklist workst fine - if we transmit the data we get the error with the last CONQUEST version and newest version 1.4.19
04.04.2017 12:26:46 [CONQUESTSRV1] Connected by address: a00aa8c0
04.04.2017 12:26:46 [CONQUESTSRV1] Testing transfer: '1.2.840.10008.1.2.4.50' against list #0 = '1.2.840.10008.1.2'
04.04.2017 12:26:46 [CONQUESTSRV1] Testing transfer: '1.2.840.10008.1.2.4.50' against list #0 = '1.2.840.10008.1.2'
04.04.2017 12:26:46 [CONQUESTSRV1] Testing transfer: '1.2.840.10008.1.2.4.50' against list #0 = '1.2.840.10008.1.2'
04.04.2017 12:26:46 [CONQUESTSRV1] Testing transfer: '1.2.840.10008.1.2.4.50' against list #0 = '1.2.840.10008.1.2'
04.04.2017 12:26:46 [CONQUESTSRV1] Testing transfer: '1.2.840.10008.1.2.4.50' against list #0 = '1.2.840.10008.1.2'
04.04.2017 12:26:46 [CONQUESTSRV1] No valid presentation contexts found
04.04.2017 12:26:46 [CONQUESTSRV1] InterogateAAssociateRQ failed
04.04.2017 12:26:46 [CONQUESTSRV1] *** multiplex: connection terminated
The received Files are only 3-30k great and seems no image content taken.
Can we config the DICOM Server for correct handshake ? On the GE Vivit we used different DICOM option without success.
Greetings Harald