Hi Marcel,
many thanks for your help. Seems working now with this converter. Have to make further tests.
Is there also a way to delete 0010,1002 for already stored data, when conquest sends data to a workstation after a retrieve command ?
Uli
Hi Marcel,
many thanks for your help. Seems working now with this converter. Have to make further tests.
Is there also a way to delete 0010,1002 for already stored data, when conquest sends data to a workstation after a retrieve command ?
Uli
Hi to all experts in Hl7 and worklists.
in the past- though having only absoluty rudimentary knowledge of Hl7, i managed to use a sligtly modified sample.hl7 file to work with siemens syngo worklist request. Siemens changed something in its software in the last months, but worklist still worked fine for us. But due to these changes, image data contains now a dicom tag 0010,1002 with item 0010,0020 (otherpatientidsSequence), but this dicom field seems to be incomplete , there should be an additional idsissuer item (0010,0021). This results in errors when importing those dicom images in e.g. dcm4chee pacs databases. Import is possible, after editing meta data and deleting this dicom tag. As fare as i could figure out, this tag is filled in because theres is no
issuer of Patient ID (0010,0021) equivalent to HL7 v2 PID-3 component 4 in the my conquest worklist data. By trial and error i tested numerous combinations of entries in pid:3 without success.
Now my question: Can anyone help me ( a simple example hl7 file would be nice) how to correctly build the hl7 pid entries, so that i get rid of this missing 0010,0021 field , or - even better - the complete dicom tag 0010,1002
Thanks for any help in advance
Uli
Hi everyone
first time this happened : importing external patient data from CD, conquest mixed the data with an existing database entry having the exact same Patientid.
Different patients could still be identified by their names when viewing them in a dicom viewer.
I thought, conquest would change the patiend id by itself to prevent mixing when there is no complete match with patient id, name or date of birth. Is there a way to prevent Conquest to mix patients data ?
Thanks
Uli
Hi Marcel
another question, not worth to make a new topic.
Is there a way to prevent conquest to delete a database entry, when the corresponding image is not found ?
In case that one of the NFS-towers is disconnected by accident or for maintance reasons, conquest cannot find the requested image data and delets the database entry. This makes some work to restore the image in the database.
Regarding the mysql connection error, i think this to be a corrupted mysql database file. I noticed, that dgate occupies continuously more and more space in the task manager (more than 100000k) until crash. After restoring a one week old backup of the mysql files everything returns to normal, as far as i can see yet.
Thanks for your help.
Uli
Hi Marcel
yes, Maxuserport is set to 65534 and TCPtimedwaitdelay to 30.
System runs on windows 2003 server.
yes, synchronisation error is part of pacstroublelog. Seems to be no serious, but i worry about performance when tons of those errors occur.
Uli
Hi
i have a problem with mysql database. Running stable for a long time, i do have lots of error messages in the last days. Main problem is
20131010 09:55:14 ***Cannot connect to MySQL on port 3306
20131010 09:55:14 ***MySQL Error message: Too many connections
repeating hundreds of times. Kill and restart does resolve only for a few minutes.
I tried to check for mysql database errors (mysql administrator), but seems to be o.k.
Another annoying error message is e.g. the following
20131010 09:54:08 Inconsistent SeriesTime in DICOMSeries: PatientID = '297626' SeriesInst = '1.2.840.113619.2.5.2147483647.24172316.131010094101.603' AND SeriesPat = '297626', Old='094746', New='094747'
conquest version 1.4.17 alpha
Thanks for any help
Uli Barkow
Hi Marcel,
it is not the conquest version. I tried with a fresh 1417beta (inbuild db), and jpeg2000 worked fine. So something is wrong with my old installation. I tried to figure out the differences in dicom.ini and copied those entries, which were new compared to the old ini (without success).
All updates in the last years were made by copying only exe and dll over the old installation, leaving dicom.sql and dgateopt.lst untouched.
Any idea before i start to risk reading back all my images ?
Thanks
Uli
Hi Marcel,
had some trouble to get 1414 work. Downgrade with copying exe and dll over copy of 1414a seems to make problems. So i could not really test it this weekend. I will try later.
But something strange i could notice. It does not seem to be a problem with jpeg compression at all, as all modalities other than MR do work with 1417a using jpeg2000.
Uli
Hi Marcel
i will test this as soon as possible.
Thanks for your reply.
Uli
Hi
anybody out there using also Conquest as server and osirix as client ?
Everything works fine as long as images are sent uncompressed. Trying to use compression ( j2 up to jl) does not work
and transfer is rejected with : DIMSE Failed to receive message. Works fine when Q/R from Osirix to Osirix, so Osirix team could not help me.
Log file says: Accepted compression jk , but then : transfer recejected.
Additional information: Conquest 1.4.17 beta. Osirix 5.6. Connection is made via vpn-tunnel.
Any idea ?
Uli Barkow
Hi,
i am faced with the follwing issue since using the conquest worklist (1.4.16):
conquest is configured to generate a studyincanceuid, otherwise Siemens MRI does not accept worklist. So far, so good, everything works fine in our internal system.
Since we have changed partially to the conquest worklist, one of our cooperation partners cannot import those files from cd in their system.
In parallel we do use another worklist , which also generates the studyinstanceuid, those files are imported without problems. So at the moment the only obvious
difference between the cds which can be imported and those which cannot, seems to be the different format of the studyinstanceuid.
e.g. conquest studyinstanceuid 1.2.826.0.1.3680043.2.135.733152.72083546.6.1349952586.125.16
mwlworklist server: studyinstanceuid: uuu.vvv.www.xxxx
Unfortunatly i could not yet get an error logfile or any more specific informations about the system, which failes to import our files (still trying to get those informations).
I´d like to know if there is any way to configure conquest to use another format for the studyincanceuid.
Thanks for any help
Uli
Problem solved :
in my configuration conquest sample.hl7 does work with syngo, when procedurestartdate/time in hl7 file is in position obr7 and conquest gives a study uid (configuration of worklist in dicom.sql file) .
Uli
Hello Conquest team,
i got stuck in trying to solve my Siemens syngo worklist problem (even though there a already some topics in this forum). As Siemens cannot assist in solving the issue, i hope someone can help me.
Worklist retrieve results in an error message, records were found, but rejected.
with debug level 4 i attach the conquest log file
1. the import of a slightly modified sample.hl7 file (the date and time of procedure start date/ time were copied one OBR Segment to the right (OBR 7), otherwise the request for the actual date results in zero records) shows the hl7 tags.
2. the log data of the Siemens request, which is far beyond my knowledge of whats going on there exactly.
Thanks for any help
Uli
20.09.2012 22:14:46 ------------ Adding image files to server -----------
20.09.2012 22:14:46 [CONQUESTSRV] Connected by address: 0100007f
20.09.2012 22:14:46 [CONQUESTSRV] Testing transfer: '1.2.840.10008.1.2' against list #0 = '1.2.840.10008.1.2'
20.09.2012 22:14:46 [CONQUESTSRV] 0000,0002 18 UI AffectedSOPClassUID "1.2.840.10008.1.1"
20.09.2012 22:14:46 [CONQUESTSRV] 0000,0100 2 US CommandField 48
20.09.2012 22:14:46 [CONQUESTSRV] 0000,0110 2 US MessageID 7
20.09.2012 22:14:46 [CONQUESTSRV] 0000,0800 2 US DataSetType 257
20.09.2012 22:14:46 [CONQUESTSRV] 0002,0010 17 UI TransferSyntaxUID "1.2.840.10008.1.2"
20.09.2012 22:14:46 [CONQUESTSRV] 9999,0400 64 UN "addimagefile:C:\\dicomserver1414beta\\hl7incoming\\temp\\sample.hl7"
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: MSH.0, contents: MSH
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: MSH.1, contents: ^~\&
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: MSH.1.0, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: MSH.1.1, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: MSH.1.2, contents: \&
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: MSH.1.END, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: MSH.2, contents: conquest
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: MSH.3, contents: conquest
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: MSH.4, contents: REC_APP
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: MSH.5, contents: REC_FAC
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: MSH.6, contents: 201209201613
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: MSH.6.DATE, contents: 20120920
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: MSH.6.TIME, contents: 1613
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: MSH.7, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: MSH.8, contents: ORM^001
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: MSH.8.0, contents: ORM
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: MSH.8.1, contents: 001
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: MSH.8.END, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: MSH.9, contents: 123456
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: MSH.10, contents: P
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: MSH.11, contents: 2.0.0
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: MSH.12, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: MSH.13, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: MSH.14, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: MSH.15, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: EOS, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PID.0, contents: PID
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PID.1, contents: 1670
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PID.2, contents: 1234567897
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PID.3, contents: 1
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PID.4, contents: 20504567
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PID.5, contents: Zijp^Lambert
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PID.5.0, contents: Zijp
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PID.5.1, contents: Lambert
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PID.5.END, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PID.6, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PID.7, contents: 19591206
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PID.7.DATE, contents: 19591206
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PID.7.TIME, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PID.8, contents: M
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PID.9, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PID.10, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PID.11, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PID.12, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PID.13, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PID.14, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PID.15, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PID.16, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PID.17, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PID.18, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PID.19, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PID.20, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PID.21, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PID.22, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PID.23, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PID.24, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PID.25, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PID.26, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: EOS, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PV1.0, contents: PV1
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PV1.1, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PV1.2, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PV1.3, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PV1.4, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PV1.5, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PV1.6, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PV1.7, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PV1.8, contents: 15^Marcel van Herk
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PV1.8.0, contents: 15
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PV1.8.1, contents: Marcel van Herk
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PV1.8.END, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PV1.9, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PV1.10, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PV1.11, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PV1.12, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PV1.13, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PV1.14, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PV1.15, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PV1.16, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PV1.17, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PV1.18, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PV1.19, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PV1.20, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PV1.21, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PV1.22, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PV1.23, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PV1.24, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PV1.25, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PV1.26, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PV1.27, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PV1.28, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PV1.29, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PV1.30, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PV1.31, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PV1.32, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PV1.33, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PV1.34, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PV1.35, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PV1.36, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PV1.37, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PV1.38, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PV1.39, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PV1.40, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PV1.41, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PV1.42, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PV1.43, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PV1.44, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PV1.44.DATE, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PV1.44.TIME, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PV1.45, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PV1.45.DATE, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PV1.45.TIME, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PV1.46, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PV1.47, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PV1.48, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: PV1.49, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: EOS, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: ORC.0, contents: ORC
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: ORC.1, contents: NW
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: ORC.2, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: ORC.3, contents: 2005-9999
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: ORC.4, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: ORC.5, contents: SC
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: ORC.6, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: ORC.7, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: ORC.8, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: ORC.9, contents: 201209191613529409
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: ORC.9.DATE, contents: 20120919
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: ORC.9.TIME, contents: 1613529409
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: ORC.10, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: ORC.11, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: ORC.12, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: ORC.13, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: ORC.14, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: ORC.15, contents: 201209191613529409
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: ORC.15.DATE, contents: 20120919
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: ORC.15.TIME, contents: 1613529409
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: ORC.16, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: ORC.17, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: ORC.18, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: EOS, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: OBR.0, contents: OBR
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: OBR.1, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: OBR.2, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: OBR.3, contents: 2005-9999
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: OBR.4, contents: 1234^BRAIN
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: OBR.4.0, contents: 1234
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: OBR.4.1, contents: BRAIN
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: OBR.4.END, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: OBR.5, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: OBR.6, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: OBR.6.DATE, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: OBR.6.TIME, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: OBR.7, contents: 20120920000001
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: OBR.7.DATE, contents: 20120920
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: OBR.7.TIME, contents: 000001
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: OBR.8, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: OBR.8.DATE, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: OBR.8.TIME, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: OBR.9, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: OBR.10, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: OBR.11, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: OBR.12, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: OBR.13, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: OBR.14, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: OBR.15, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: OBR.16, contents:
20.09.2012 22:14:46 [CONQUESTSRV] HL7 item: OBR.17, contents:
20.09.2012 22:14:46 -----------------------------------------------------
20.09.2012 22:14:51 [CONQUESTSRV] Connected by address: 0100007f
20.09.2012 22:14:51 [CONQUESTSRV] Testing transfer: '1.2.840.10008.1.2' against list #0 = '1.2.840.10008.1.2'
20.09.2012 22:14:51 [CONQUESTSRV] 0000,0002 18 UI AffectedSOPClassUID "1.2.840.10008.1.1"
20.09.2012 22:14:51 [CONQUESTSRV] 0000,0100 2 US CommandField 48
20.09.2012 22:14:51 [CONQUESTSRV] 0000,0110 2 US MessageID 7
20.09.2012 22:14:51 [CONQUESTSRV] 0000,0800 2 US DataSetType 257
20.09.2012 22:14:51 [CONQUESTSRV] 0002,0010 17 UI TransferSyntaxUID "1.2.840.10008.1.2"
20.09.2012 22:14:51 [CONQUESTSRV] 9999,0400 6 UN "silent"
20.09.2012 22:15:08 [CONQUESTSRV] Connected by address: e601a8c0
20.09.2012 22:15:08 [CONQUESTSRV] Testing transfer: '1.2.840.10008.1.2.1' against list #0 = '1.2.840.10008.1.2'
20.09.2012 22:15:08 [CONQUESTSRV] Testing transfer: '1.2.840.10008.1.2.1' against list #1 = '1.2.840.10008.1.2.1'
20.09.2012 22:15:09 [CONQUESTSRV]
20.09.2012 22:15:09 [CONQUESTSRV] UPACS THREAD 51689: STARTED AT: Thu Sep 20 22:15:08 2012
20.09.2012 22:15:09 [CONQUESTSRV] A-ASSOCIATE-RQ Packet Dump
20.09.2012 22:15:09 [CONQUESTSRV] Calling Application Title : "AWP45274 "
20.09.2012 22:15:09 [CONQUESTSRV] Called Application Title : "CONQUESTSRV "
20.09.2012 22:15:09 [CONQUESTSRV] Application Context : "1.2.840.10008.3.1.1.1", PDU length: 262144
20.09.2012 22:15:09 [CONQUESTSRV] Number of Proposed Presentation Contexts: 1
20.09.2012 22:15:09 [CONQUESTSRV] Presentation Context 0 "1.2.840.10008.5.1.4.31" 1
20.09.2012 22:15:09 [CONQUESTSRV] Server Command := 0020
20.09.2012 22:15:09 [CONQUESTSRV] Message ID := 0030
20.09.2012 22:15:09 [CONQUESTSRV] 0000,0002 22 UI AffectedSOPClassUID "1.2.840.10008.5.1.4.31"
20.09.2012 22:15:09 [CONQUESTSRV] 0000,0100 2 US CommandField 32
20.09.2012 22:15:09 [CONQUESTSRV] 0000,0110 2 US MessageID 48
20.09.2012 22:15:09 [CONQUESTSRV] 0000,0700 2 US Priority 0
20.09.2012 22:15:09 [CONQUESTSRV] 0000,0800 2 US DataSetType 0
20.09.2012 22:15:09 [CONQUESTSRV] 0002,0010 19 UI TransferSyntaxUID "1.2.840.10008.1.2.1"
20.09.2012 22:15:09 [CONQUESTSRV] (ModalityWorkListQuery) search level:
20.09.2012 22:15:09 [CONQUESTSRV] 0002,0010 19 UI TransferSyntaxUID "1.2.840.10008.1.2.1"
20.09.2012 22:15:09 [CONQUESTSRV] 0008,0005 0 CS SpecificCharacterSet (empty)
20.09.2012 22:15:09 [CONQUESTSRV] 0008,0050 0 SH AccessionNumber (empty)
20.09.2012 22:15:09 [CONQUESTSRV] 0008,0080 0 LO InstitutionName (empty)
20.09.2012 22:15:09 [CONQUESTSRV] 0008,0081 0 ST InstitutionAddress (empty)
20.09.2012 22:15:09 [CONQUESTSRV] 0008,0082 0 SQ InstitutionCodeSeque Sequence Item
20.09.2012 22:15:09 [CONQUESTSRV] 0008,0090 0 PN ReferringPhysicianNa (empty)
20.09.2012 22:15:09 [CONQUESTSRV] 0008,1080 0 LO AdmittingDiagnosisDe (empty)
20.09.2012 22:15:09 [CONQUESTSRV] 0008,1110 0 SQ ReferencedStudySeque Sequence Item
20.09.2012 22:15:09 [CONQUESTSRV] 0008,1120 0 SQ ReferencedPatientSeq Sequence Item
20.09.2012 22:15:09 [CONQUESTSRV] 0008,1125 0 SQ ReferencedVisitSeque Sequence Item
20.09.2012 22:15:09 [CONQUESTSRV] 0010,0010 0 PN PatientName (empty)
20.09.2012 22:15:09 [CONQUESTSRV] 0010,0020 0 LO PatientID (empty)
20.09.2012 22:15:09 [CONQUESTSRV] 0010,0021 0 LO IssuerOfPatientID (empty)
20.09.2012 22:15:09 [CONQUESTSRV] 0010,0030 0 DA PatientBirthDate (empty)
20.09.2012 22:15:09 [CONQUESTSRV] 0010,0032 0 TM PatientBirthTime (empty)
20.09.2012 22:15:09 [CONQUESTSRV] 0010,0040 0 CS PatientSex (empty)
20.09.2012 22:15:09 [CONQUESTSRV] 0010,0050 0 SQ PatientInsurancePlan Sequence Item
20.09.2012 22:15:09 [CONQUESTSRV] 0010,1000 0 LO OtherPatientID (empty)
20.09.2012 22:15:09 [CONQUESTSRV] 0010,1001 0 PN OtherPatientName (empty)
20.09.2012 22:15:09 [CONQUESTSRV] 0010,1002 0 UN (empty)
20.09.2012 22:15:09 [CONQUESTSRV] >0010,0020 0 LO PatientID (empty)
20.09.2012 22:15:09 [CONQUESTSRV] >0010,0021 0 LO IssuerOfPatientID (empty)
20.09.2012 22:15:09 [CONQUESTSRV] >0010,0022 0 UN (empty)
20.09.2012 22:15:09 [CONQUESTSRV] 0010,1005 0 PN PatientBirthName (empty)
20.09.2012 22:15:09 [CONQUESTSRV] 0010,1010 0 AS PatientAge (empty)
20.09.2012 22:15:09 [CONQUESTSRV] 0010,1020 0 DS PatientSize (empty)
20.09.2012 22:15:09 [CONQUESTSRV] 0010,1030 0 DS PatientWeight (empty)
20.09.2012 22:15:09 [CONQUESTSRV] 0010,1040 0 LO PatientAddress (empty)
20.09.2012 22:15:09 [CONQUESTSRV] 0010,1060 0 PN PatientMotherBirthNa (empty)
20.09.2012 22:15:09 [CONQUESTSRV] 0010,1080 0 LO MilitaryRank (empty)
20.09.2012 22:15:09 [CONQUESTSRV] 0010,1081 0 LO BranchOfService (empty)
20.09.2012 22:15:09 [CONQUESTSRV] 0010,1090 0 LO MedicalRecordLocator (empty)
20.09.2012 22:15:09 [CONQUESTSRV] 0010,2000 0 LO MedicalAlerts (empty)
20.09.2012 22:15:09 [CONQUESTSRV] 0010,2110 0 LO ContrastAllergies (empty)
20.09.2012 22:15:09 [CONQUESTSRV] 0010,2150 0 LO CountryOfResidence (empty)
20.09.2012 22:15:09 [CONQUESTSRV] 0010,2152 0 LO RegionOfResidence (empty)
20.09.2012 22:15:09 [CONQUESTSRV] 0010,2154 0 SH PatientTelephoneNumb (empty)
20.09.2012 22:15:09 [CONQUESTSRV] 0010,2160 0 SH EthnicGroup (empty)
20.09.2012 22:15:09 [CONQUESTSRV] 0010,2180 0 SH Occupation (empty)
20.09.2012 22:15:09 [CONQUESTSRV] 0010,21a0 0 CS SmokingStatus (empty)
20.09.2012 22:15:09 [CONQUESTSRV] 0010,21b0 0 LT AdditionalPatientHis (empty)
20.09.2012 22:15:09 [CONQUESTSRV] 0010,21c0 0 US PregnancyStatus (empty)
20.09.2012 22:15:09 [CONQUESTSRV] 0010,21d0 0 DA LastMenstrualDate (empty)
20.09.2012 22:15:09 [CONQUESTSRV] 0010,21f0 0 LO PatientReligiousPref (empty)
20.09.2012 22:15:09 [CONQUESTSRV] 0010,4000 0 LT PatientComments (empty)
20.09.2012 22:15:09 [CONQUESTSRV] 0020,000d 0 UI StudyInstanceUID (empty)
20.09.2012 22:15:09 [CONQUESTSRV] 0032,1032 0 PN RequestingPhysician (empty)
20.09.2012 22:15:09 [CONQUESTSRV] 0032,1033 0 LO RequestingService (empty)
20.09.2012 22:15:09 [CONQUESTSRV] 0032,1060 0 LO RequestedProcedureDe (empty)
20.09.2012 22:15:09 [CONQUESTSRV] 0032,1064 0 SQ RequestedProcedureCo Sequence Item
20.09.2012 22:15:09 [CONQUESTSRV] 0038,0004 0 SQ ReferencedPatientAli Sequence Item
20.09.2012 22:15:09 [CONQUESTSRV] 0038,0008 0 CS VisitStatusID (empty)
20.09.2012 22:15:09 [CONQUESTSRV] 0038,0010 0 LO AdmissionID (empty)
20.09.2012 22:15:09 [CONQUESTSRV] 0038,0011 0 LO IssuerOfAdmissionID (empty)
20.09.2012 22:15:09 [CONQUESTSRV] 0038,0050 0 LO SpecialNeeds (empty)
20.09.2012 22:15:09 [CONQUESTSRV] 0038,0300 0 LO CurrentPatientLocati (empty)
20.09.2012 22:15:09 [CONQUESTSRV] 0038,0400 0 LO PatientInstitutionRe (empty)
20.09.2012 22:15:09 [CONQUESTSRV] 0038,0500 0 LO PatientState (empty)
20.09.2012 22:15:09 [CONQUESTSRV] 0038,4000 0 LT VisitComments (empty)
20.09.2012 22:15:09 [CONQUESTSRV] 0040,0100 0 SQ ScheduledProcedureSt Sequence Item
20.09.2012 22:15:09 [CONQUESTSRV] >0008,0060 0 CS Modality (empty)
20.09.2012 22:15:09 [CONQUESTSRV] >0032,1070 0 LO RequestedContrastAge (empty)
20.09.2012 22:15:09 [CONQUESTSRV] >0040,0001 0 AE ScheduledStationAETi (empty)
20.09.2012 22:15:09 [CONQUESTSRV] >0040,0002 18 DA ScheduledProcedureSt "20120919-20120921 "
20.09.2012 22:15:09 [CONQUESTSRV] >0040,0003 14 TM ScheduledProcedureSt "000000-235959 "
20.09.2012 22:15:09 [CONQUESTSRV] >0040,0004 0 DA ScheduledProcedureSt (empty)
20.09.2012 22:15:09 [CONQUESTSRV] >0040,0005 0 TM ScheduledProcedureSt (empty)
20.09.2012 22:15:09 [CONQUESTSRV] >0040,0006 0 PN ScheduledPerformingP (empty)
20.09.2012 22:15:09 [CONQUESTSRV] >0040,0007 0 LO ScheduledProcedureSt (empty)
20.09.2012 22:15:09 [CONQUESTSRV] >0040,0008 0 SQ ScheduledProcedureSt Sequence Item
20.09.2012 22:15:09 [CONQUESTSRV] >0040,0009 0 SH ScheduledProcedureSt (empty)
20.09.2012 22:15:09 [CONQUESTSRV] >0040,0010 0 SH ScheduledStationName (empty)
20.09.2012 22:15:09 [CONQUESTSRV] >0040,0011 0 SH ScheduledProcedureSt (empty)
20.09.2012 22:15:09 [CONQUESTSRV] >0040,0012 0 LO ScheduledPreOrderOfM (empty)
20.09.2012 22:15:09 [CONQUESTSRV] >0040,0020 0 UN (empty)
20.09.2012 22:15:09 [CONQUESTSRV] >0040,0400 0 LT CommentsOnScheduledP (empty)
20.09.2012 22:15:09 [CONQUESTSRV] 0040,1001 0 SH RequestedProcedureID (empty)
20.09.2012 22:15:09 [CONQUESTSRV] 0040,1002 0 LO ReasonForRequestedPr (empty)
20.09.2012 22:15:09 [CONQUESTSRV] 0040,1003 0 SH RequestedProcedurePr (empty)
20.09.2012 22:15:09 [CONQUESTSRV] 0040,1004 0 LO PatientTransportArra (empty)
20.09.2012 22:15:09 [CONQUESTSRV] 0040,1005 0 LO RequestedProcedureLo (empty)
20.09.2012 22:15:09 [CONQUESTSRV] 0040,1008 0 LO ConfidentialityCode (empty)
20.09.2012 22:15:09 [CONQUESTSRV] 0040,1009 0 UN (empty)
20.09.2012 22:15:09 [CONQUESTSRV] 0040,1010 0 UN (empty)
20.09.2012 22:15:09 [CONQUESTSRV] 0040,1400 0 LT RequestedProcedureCo (empty)
20.09.2012 22:15:09 [CONQUESTSRV] 0040,2004 0 DA IssueDateOfImagingSe (empty)
20.09.2012 22:15:09 [CONQUESTSRV] 0040,2005 0 TM IssueTimeOfImagingSe (empty)
20.09.2012 22:15:09 [CONQUESTSRV] 0040,2008 0 PN OrderEnteredBy (empty)
20.09.2012 22:15:09 [CONQUESTSRV] 0040,2009 0 SH OrderEntererLocation (empty)
20.09.2012 22:15:09 [CONQUESTSRV] 0040,2010 0 SH OrderCallbackPhoneNu (empty)
20.09.2012 22:15:09 [CONQUESTSRV] 0040,2016 0 UN (empty)
20.09.2012 22:15:09 [CONQUESTSRV] 0040,2017 0 UN (empty)
20.09.2012 22:15:09 [CONQUESTSRV] 0040,2400 0 LT ImagingServiceReques (empty)
20.09.2012 22:15:09 [CONQUESTSRV] 0040,3001 0 LT ConfidentialityConst (empty)
20.09.2012 22:15:09 [CONQUESTSRV] Query On Modality WorkList
20.09.2012 22:15:09 [CONQUESTSRV] Queried item 0008 0005 is not in the database
20.09.2012 22:15:09 [CONQUESTSRV] Queried item 0008 0080 is not in the database
20.09.2012 22:15:09 [CONQUESTSRV] Queried item 0008 0081 is not in the database
20.09.2012 22:15:09 [CONQUESTSRV] Queried item 0008 0082 is not in the database
20.09.2012 22:15:09 [CONQUESTSRV] Queried item 0008 0090 is not in the database
20.09.2012 22:15:09 [CONQUESTSRV] Queried item 0008 1080 is not in the database
20.09.2012 22:15:09 [CONQUESTSRV] Queried item 0008 1110 is not in the database
20.09.2012 22:15:09 [CONQUESTSRV] Queried item 0008 1120 is not in the database
20.09.2012 22:15:09 [CONQUESTSRV] Queried item 0008 1125 is not in the database
20.09.2012 22:15:09 [CONQUESTSRV] Queried item 0010 0021 is not in the database
20.09.2012 22:15:09 [CONQUESTSRV] Queried item 0010 0032 is not in the database
20.09.2012 22:15:09 [CONQUESTSRV] Queried item 0010 0050 is not in the database
20.09.2012 22:15:09 [CONQUESTSRV] Queried item 0010 1000 is not in the database
20.09.2012 22:15:09 [CONQUESTSRV] Queried item 0010 1001 is not in the database
20.09.2012 22:15:09 [CONQUESTSRV] Queried item 0010 0021 is not in the database
20.09.2012 22:15:09 [CONQUESTSRV] Queried item 0010 0022 is not in the database
20.09.2012 22:15:09 [CONQUESTSRV] 0008,0050 10 SH AccessionNumber "2005-9999 "
20.09.2012 22:15:09 [CONQUESTSRV] UPACS THREAD 51689: ENDED AT: Thu Sep 20 22:15:09 2012
20.09.2012 22:15:09 [CONQUESTSRV] UPACS THREAD 51689: TOTAL RUNNING TIME: 1 SECONDS
20.09.2012 22:15:51 [CONQUESTSRV] Connected by address: 0100007f
20.09.2012 22:15:51 [CONQUESTSRV] Testing transfer: '1.2.840.10008.1.2' against list #0 = '1.2.840.10008.1.2'
20.09.2012 22:15:51 [CONQUESTSRV] 0000,0002 18 UI AffectedSOPClassUID "1.2.840.10008.1.1"
20.09.2012 22:15:51 [CONQUESTSRV] 0000,0100 2 US CommandField 48
20.09.2012 22:15:51 [CONQUESTSRV] 0000,0110 2 US MessageID 7
20.09.2012 22:15:51 [CONQUESTSRV] 0000,0800 2 US DataSetType 257
20.09.2012 22:15:51 [CONQUESTSRV] 0002,0010 17 UI TransferSyntaxUID "1.2.840.10008.1.2"
20.09.2012 22:15:51 [CONQUESTSRV] 9999,0400 6 UN "silent"
20.09.2012 22:16:51 [CONQUESTSRV] Connected by address: 0100007f
20.09.2012 22:16:51 [CONQUESTSRV] Testing transfer: '1.2.840.10008.1.2' against list #0 = '1.2.840.10008.1.2'
20.09.2012 22:16:51 [CONQUESTSRV] 0000,0002 18 UI AffectedSOPClassUID "1.2.840.10008.1.1"
20.09.2012 22:16:51 [CONQUESTSRV] 0000,0100 2 US CommandField 48
20.09.2012 22:16:51 [CONQUESTSRV] 0000,0110 2 US MessageID 7
20.09.2012 22:16:51 [CONQUESTSRV] 0000,0800 2 US DataSetType 257
20.09.2012 22:16:51 [CONQUESTSRV] 0002,0010 17 UI TransferSyntaxUID "1.2.840.10008.1.2"
20.09.2012 22:16:51 [CONQUESTSRV] 9999,0400 6 UN "silent"
20.09.2012 22:17:51 [CONQUESTSRV] Connected by address: 0100007f
20.09.2012 22:17:51 [CONQUESTSRV] Testing transfer: '1.2.840.10008.1.2' against list #0 = '1.2.840.10008.1.2'
20.09.2012 22:17:51 [CONQUESTSRV] 0000,0002 18 UI AffectedSOPClassUID "1.2.840.10008.1.1"
20.09.2012 22:17:51 [CONQUESTSRV] 0000,0100 2 US CommandField 48
20.09.2012 22:17:51 [CONQUESTSRV] 0000,0110 2 US MessageID 7
20.09.2012 22:17:51 [CONQUESTSRV] 0000,0800 2 US DataSetType 257
20.09.2012 22:17:51 [CONQUESTSRV] 0002,0010 17 UI TransferSyntaxUID "1.2.840.10008.1.2"
20.09.2012 22:17:51 [CONQUESTSRV] 9999,0400 6 UN "silent"
20.09.2012 22:18:51 [CONQUESTSRV] Connected by address: 0100007f
20.09.2012 22:18:51 [CONQUESTSRV] Testing transfer: '1.2.840.10008.1.2' against list #0 = '1.2.840.10008.1.2'
20.09.2012 22:18:51 [CONQUESTSRV] 0000,0002 18 UI AffectedSOPClassUID "1.2.840.10008.1.1"
20.09.2012 22:18:51 [CONQUESTSRV] 0000,0100 2 US CommandField 48
20.09.2012 22:18:51 [CONQUESTSRV] 0000,0110 2 US MessageID 7
20.09.2012 22:18:51 [CONQUESTSRV] 0000,0800 2 US DataSetType 257
20.09.2012 22:18:51 [CONQUESTSRV] 0002,0010 17 UI TransferSyntaxUID "1.2.840.10008.1.2"
20.09.2012 22:18:51 [CONQUESTSRV] 9999,0400 6 UN "silent"
20.09.2012 22:19:51 [CONQUESTSRV] Connected by address: 0100007f
20.09.2012 22:19:51 [CONQUESTSRV] Testing transfer: '1.2.840.10008.1.2' against list #0 = '1.2.840.10008.1.2'
20.09.2012 22:19:51 [CONQUESTSRV] 0000,0002 18 UI AffectedSOPClassUID "1.2.840.10008.1.1"
20.09.2012 22:19:51 [CONQUESTSRV] 0000,0100 2 US CommandField 48
20.09.2012 22:19:51 [CONQUESTSRV] 0000,0110 2 US MessageID 7
20.09.2012 22:19:51 [CONQUESTSRV] 0000,0800 2 US DataSetType 257
20.09.2012 22:19:51 [CONQUESTSRV] 0002,0010 17 UI TransferSyntaxUID "1.2.840.10008.1.2"
20.09.2012 22:19:51 [CONQUESTSRV] 9999,0400 6 UN "silent"
20.09.2012 22:19:55 [CONQUESTSRV] Connected by address: 0100007f
20.09.2012 22:19:55 [CONQUESTSRV] Testing transfer: '1.2.840.10008.1.2' against list #0 = '1.2.840.10008.1.2'
20.09.2012 22:19:55 [CONQUESTSRV] 0000,0002 18 UI AffectedSOPClassUID "1.2.840.10008.1.1"
20.09.2012 22:19:55 [CONQUESTSRV] 0000,0100 2 US CommandField 48
20.09.2012 22:19:55 [CONQUESTSRV] 0000,0110 2 US MessageID 7
20.09.2012 22:19:55 [CONQUESTSRV] 0000,0800 2 US DataSetType 257
20.09.2012 22:19:55 [CONQUESTSRV] 0002,0010 17 UI TransferSyntaxUID "1.2.840.10008.1.2"
20.09.2012 22:19:55 [CONQUESTSRV] 9999,0300 24 UN "set debug level from GUI"
20.09.2012 22:19:55 [CONQUESTSRV] 9999,0400 12 UN "debuglevel:3"
20.09.2012 22:19:55 [CONQUESTSRV] set debug level from GUI
Thanks for your help.
works fine except one problem. I must admit having some problems with the script language.
I do have to insert a stop command in the forward expression, when incoming images have the patient name "Konstanz Mammo". Otherwise the system will send all quality images (which are made every day ) to the destination.
so I need something like : ifequal "patientname","KONSTANZ Mammo"; stop; forward .....
So what is the correct parameter for patientname and will this work ?
Thanks for any help
Uli
Hello
I searched the forum and manual unfortunatly without success for my problem :
After incoming of images from a given modality, e.g. Mammography, i would like to activate an export converter, which forwards the new study together with all older studies of that patient with the same modality to another destination (like forward Patient (only Mammography) to Dest).
Is there a filter or anything like that to implement in the export converters or
is there a way to realize this with a dgate movestudy command ?
Thanks for any help
Uli Barkow
Hello
we are searching for a new NAS with a capacity about 10 TB. We used Thecus 5200 the last years, but those are running out of space.
As we experienced some perfomance problems (eg. database rebuild) with thecus 5200, i ´d like to know if anyone can give a recommendation for a perfect running system.
Thanks for any Help
Uli Barkow