K-PACS provides standard conformance to the
XRayAngiographicImageStorage -1.2.840.10008.5.1.4.1.1.12.1,
XRayFluoroscopyImageStorage -1.2.840.10008.5.1.4.1.1.12.2,
as well as
RETIRED_XRayAngiographicBiPlaneImageStorage -1.2.840.10008.5.1.4.1.1.12.3
SOP Classes as an SCP.
I personally do not now the Thosiba Infinix system but the DICOM conformance statement downloadable at Thoshiba's homepage claims standard conformance for XRayAngiographicImageStorage as an SCU.
Maybe there is a mistake in the AET/IP/Port settings or an unsupported transfer syntax is used although the default value for the Infinix system is "Implicit VR little endian" which is supported by K-PACS.
regards,
Andreas