    Yves,



    There is RealTime Image company's PACS, I will try to find out the version today. The hospital have no support form the vendor anymore. So we can't change nothing in the PACS. Anyway somehow we must to solve the issue because we are expecting to sell more iQ licences for this customer and this customer is very important to us. Can you just change needful line in program code to right one and compile the program? I think this is easy job for you. The situation - in the hospital we installed DEMO unit. Next month will be tender for for software and we still have a problem with modality filtration. Could you give the help for us?


    I enclose an attachment where you can find print screen of administrator window in PACS. We tried to add TAG [0008,0061], but there isn't describtion of this TAG and PACS doesn't give answer of queries. So Vendor didn't describe this TAG.


    P.S. I can not attache image. You will get it by email (support@mage-systems.biz) immediately.
    Sarunas

    Quote from Yves Neumann

    Technically, a DICOM Query on study level (highest level in iQ-VIEW when querying) does not contain the modality information (DICOM 0008,0060) but the "modalities in studies" (DICOM 0008,0061). The modality information is on series level. An individual study can contain series of different modalities.


    It is up to the PACS (or better queried entity) to support the "Modalities in study" as query tag or not. As Sabine said, the DICOM conformance statement should give an answer. DICOM does not define the support of that tag as mandatory.


    I just talk with guy from hospital and he told that PACS is using the TAG [0008,0060] and other workstations work properly. Can we change series modality tag [0008, 0061] to [0008,0060] in iQ-View?

    Hello,


    We have installed iQ-View 2.5.0c Trial version. We put images from archive and on every patient in modality thumbnail is question-mark. When I extract patient then I can see modality in modality column but software can not able to identify modality. Is it bug or something wrong with settings of software?


    Thanks for help