Marcel,
thank you so much for replying. I actually turned off the firewall for private network on the kpacs workstation and I have the same issue. It just keeps saying that the workstation did not accept the connection.
I really don't know what's preventing the images from being retrieved.
Querying studies from the workstation is never a problem. When I start requesting studies the process starts but then it fails....I am attaching some logs and snapshots to show what's happening.
Below is the log from the kpacs:
*******************************************************
1/8/2021 12:28:44 PM ThreadID[1884] : Retrieving images from CONQUESTSRV1 started
*******************************************************
[1884] >> : c-find request for StudyInstanceUID: 1.2.840.113619.6.106.2294885607438502020627296462822907098891 initated
[DICOM SCU] >> : Connected to: 192.168.1.10:5678
[DICOM SCU] >> : Item Type: 2
[DICOM SCU] >> : Association Accept::ReadDynamic
[DICOM SCU] >> : Presentation Context Accept, Transfer Syntax: 1.2.840.10008.1.2
[1884] >> : c-find result: 1 sequences in study with UID 1.2.840.113619.6.106.2294885607438502020627296462822907098891
[1884] >> : c-move request for sequence 1.2.840.113619.2.364.65140653520588611400.1610028167.7890.0 initiated.
[DICOM SCU] >> : Connected to: 192.168.1.10:5678
[DICOM SCU] >> : Item Type: 2
[DICOM SCU] >> : Association Accept::ReadDynamic
[DICOM SCU] >> : Presentation Context Accept, Transfer Syntax: 1.2.840.10008.1.2
[DICOM SCU] >> : KPServer15 >> CONQUESTSRV1
[DICOM SCU] >> : (0000,0000), 4, UL, "94" - Group Length
[DICOM SCU] >> : (0000,0002), 28, UI, "1.2.840.10008.5.1.4.1.2.2.2" - Affected SOP Class UID
[DICOM SCU] >> : (0000,0100), 2, US, "33" - Command Field
[DICOM SCU] >> : (0000,0110), 2, US, "3" - Message ID
[DICOM SCU] >> : (0000,0600), 10, AE, "KPServer15" - Move Destination
[DICOM SCU] >> : (0000,0700), 2, US, "0" - Priority
[DICOM SCU] >> : (0000,0800), 2, US, "258" - Data Set Type
[DICOM SCU] >> :
[DICOM SCU] >> : KPServer15 >> CONQUESTSRV1
[DICOM SCU] >> : (0008,0021), 0, DA, "" - Series Date
[DICOM SCU] >> : (0008,0031), 0, TM, "" - Series Time
[DICOM SCU] >> : (0008,0052), 6, CS, "SERIES" - Query/Retrieve Level
[DICOM SCU] >> : (0008,0060), 0, CS, "" - Modality
[DICOM SCU] >> : (0008,103E), 0, LO, "" - Series Description
[DICOM SCU] >> : (0018,0015), 0, CS, "" - Body Part Examined
[DICOM SCU] >> : (0018,1030), 0, LO, "" - Protocol Name
[DICOM SCU] >> : (0018,5100), 0, CS, "" - Patient Position
[DICOM SCU] >> : (0020,000D), 62, UI, "1.2.840.113619.6.106.2294885607438502020627296462822907098891" - Study Instance UID
[DICOM SCU] >> : (0020,000E), 60, UI, "1.2.840.113619.2.364.65140653520588611400.1610028167.7890.0" - Series Instance UID
[DICOM SCU] >> : (0020,0011), 0, IS, "" - Series Number
[DICOM SCU] >> :
[DICOM SCU] >> : KPServer15 << CONQUESTSRV1
[DICOM SCU] >> : (0000,0000), 4, UL, "116" - Group Length
[DICOM SCU] >> : (0000,0002), 28, UI, "1.2.840.10008.5.1.4.1.2.2.2" - Affected SOP Class UID
[DICOM SCU] >> : (0000,0100), 2, US, "32801" - Command Field
[DICOM SCU] >> : (0000,0120), 2, US, "3" - Message ID Being Responded To
[DICOM SCU] >> : (0000,0800), 2, US, "257" - Data Set Type
[DICOM SCU] >> : (0000,0900), 2, US, "49157" - Status
[DICOM SCU] >> : (0000,1020), 2, US, "0" - Number of Remaining Sub-Operations
[DICOM SCU] >> : (0000,1021), 2, US, "0" - Number of Completed Sub-Operations
[DICOM SCU] >> : (0000,1022), 2, US, "0" - Number of Failed Sub-Operations
[DICOM SCU] >> : (0000,1023), 2, US, "0" - Number of Warning Sub-Operations
[DICOM SCU] >> :
[1884] >> : Error: retreive failed for Doe, Jane , Study: LE Angio/LE PTA , Sequence: 1
[1884] >> : c-move request for sequence 1.2.840.113619.2.364.65140653520588611400.1610028167.7890.1 initiated.
[DICOM SCU] >> : Connected to: 192.168.1.10:5678
[DICOM SCU] >> : Item Type: 2
[DICOM SCU] >> : Association Accept::ReadDynamic
[DICOM SCU] >> : Presentation Context Accept, Transfer Syntax: 1.2.840.10008.1.2
[DICOM SCU] >> : KPServer15 >> CONQUESTSRV1
[DICOM SCU] >> : (0000,0000), 4, UL, "94" - Group Length
[DICOM SCU] >> : (0000,0002), 28, UI, "1.2.840.10008.5.1.4.1.2.2.2" - Affected SOP Class UID
[DICOM SCU] >> : (0000,0100), 2, US, "33" - Command Field
[DICOM SCU] >> : (0000,0110), 2, US, "3" - Message ID
[DICOM SCU] >> : (0000,0600), 10, AE, "KPServer15" - Move Destination
[DICOM SCU] >> : (0000,0700), 2, US, "0" - Priority
[DICOM SCU] >> : (0000,0800), 2, US, "258" - Data Set Type
[DICOM SCU] >> :
[DICOM SCU] >> : KPServer15 >> CONQUESTSRV1
[DICOM SCU] >> : (0008,0021), 0, DA, "" - Series Date
[DICOM SCU] >> : (0008,0031), 0, TM, "" - Series Time
[DICOM SCU] >> : (0008,0052), 6, CS, "SERIES" - Query/Retrieve Level
[DICOM SCU] >> : (0008,0060), 0, CS, "" - Modality
[DICOM SCU] >> : (0008,103E), 0, LO, "" - Series Description
[DICOM SCU] >> : (0018,0015), 0, CS, "" - Body Part Examined
[DICOM SCU] >> : (0018,1030), 0, LO, "" - Protocol Name
[DICOM SCU] >> : (0018,5100), 0, CS, "" - Patient Position
[DICOM SCU] >> : (0020,000D), 62, UI, "1.2.840.113619.6.106.2294885607438502020627296462822907098891" - Study Instance UID
[DICOM SCU] >> : (0020,000E), 60, UI, "1.2.840.113619.2.364.65140653520588611400.1610028167.7890.1" - Series Instance UID
[DICOM SCU] >> : (0020,0011), 0, IS, "" - Series Number
[DICOM SCU] >> :
[DICOM SCU] >> : KPServer15 << CONQUESTSRV1
[DICOM SCU] >> : (0000,0000), 4, UL, "116" - Group Length
[DICOM SCU] >> : (0000,0002), 28, UI, "1.2.840.10008.5.1.4.1.2.2.2" - Affected SOP Class UID
[DICOM SCU] >> : (0000,0100), 2, US, "32801" - Command Field
[DICOM SCU] >> : (0000,0120), 2, US, "3" - Message ID Being Responded To
[DICOM SCU] >> : (0000,0800), 2, US, "257" - Data Set Type
[DICOM SCU] >> : (0000,0900), 2, US, "49157" - Status
[DICOM SCU] >> : (0000,1020), 2, US, "0" - Number of Remaining Sub-Operations
[DICOM SCU] >> : (0000,1021), 2, US, "0" - Number of Completed Sub-Operations
[DICOM SCU] >> : (0000,1022), 2, US, "0" - Number of Failed Sub-Operations
[DICOM SCU] >> : (0000,1023), 2, US, "0" - Number of Warning Sub-Operations
[DICOM SCU] >> :
[1884] >> : Error: retreive failed for Doe, Jane , Study: LE Angio/LE PTA , Sequence: 1
*******************************************************
1/8/2021 12:28:49 PM ThreadID[1884] : Closed.
I attached a snapshot of the server side.