Hi all,
I have a DX modality which is sending a study with (for example) 4 images in one serie.
Is it possible, f.e. using an ImportConverter to store this study with 1 image in 4 different seriesfor this study?
Chris
Hi all,
I have a DX modality which is sending a study with (for example) 4 images in one serie.
Is it possible, f.e. using an ImportConverter to store this study with 1 image in 4 different seriesfor this study?
Chris
Hi,
Seems a problem with this modality.
Upload to other SC's also fails.
Sorry for the inconvenience,
Chris
Hi,
When I try to upload from a modality
This the output from the server status screen:
[CONQUESTDRAAK] User interface test: local server is running!
[CONQUESTDRAAK]
[CONQUESTDRAAK] UPACS THREAD 0: STARTED AT: Fri Apr 11 16:58:59 2008
[CONQUESTDRAAK] Calling Application Title : "KPServerDraak"
[CONQUESTDRAAK] Called Application Title : "CONQUESTDRAAK"
[CONQUESTDRAAK] Application Context : "1.2.840.10008.3.1.1.1"
[CONQUESTDRAAK] Presentation Context 0 "1.2.840.10008.5.1.4.1.1.1.1"
[CONQUESTDRAAK] *** Association rejected: you may need to add the listed presentation context as sop to dgatesop.lst
[CONQUESTDRAAK] UPACS THREAD 0: ENDED AT: Fri Apr 11 16:59:00 2008
[CONQUESTDRAAK] UPACS THREAD 0: TOTAL RUNNING TIME: 1 SECONDS
However, my dgatesop.list shows:
#
# DICOM Application / sop / transfer UID list.
#
# This list is used by the CheckedPDU_Service ( "filename" ) service
# class. All incoming associations will be verified against this
# file.
#
# Revision 2: disabled GEMRStorage and GECTStorage
# Revision 3: extended with new sops and with JPEG transfer syntaxes
# Revision 4: added Modality Worklist query
#
#None none RemoteAE
#None none LocalAE
#DICOM 1.2.840.10008.3.1.1.1 application
Verification 1.2.840.10008.1.1 sop
StoredPrintStorage 1.2.840.10008.5.1.1.27 sop
HardcopyGrayscaleImageStorage 1.2.840.10008.5.1.1.29 sop
HardcopyColorImageStorage 1.2.840.10008.5.1.1.30 sop
CRStorage 1.2.840.10008.5.1.4.1.1.1 sop
DXStorageForPresentation 1.2.840.10008.5.1.4.1.1.1.1 sop
^
^
What's going wrong?
Chris
Hi,
RTFM, in this case very essential:
also running on my Conquest server: KPacs with same AET as on the remote machines: KPServer.
Obvious you have to use unique AET's within your Conquest server and also configure on your remote clients.
Sorry for the inconvenience.
Chris
Hi,
When using "Verbose Mode", no output is seen in the commandline screen.
Should this be then a network problem?
Chris
Hi,
Loading a study, it seems to take some time to load (severals seconds, up to 6 or more, according the Conquest log), but nothing seems to be saved in the local KPacs database. I suppose that's the reason not able to see any images.
It's quite confusing, sometimes it works on a particular machine, sometimes it won't.
Even more frustrating if any other apps are working from anywere.
Chris
Hi all,
After a fresh installation of KPacs 1.5.0 seem to perform well and without errors both on the remote LAN Conquest side and KPacs side. However, the images are not displayed an not put in the KPacs Imagebox folder.
When retrieving from Conquest on the same PC as KPacs, everything goes well.
Retrieving images from the remote Conquest server with another Dicom-viewer (i2view) goes also well.
Switching off firewalls and antivirus makes no difference.
Below the KPacs log.
*******************************************************
5-4-2008 11:26:41 ThreadID[888] : Retrieving images from CONQUESTSRV1 started
*******************************************************
[888] >> : c-find request for StudyInstanceUID: 1.3.51.0.7.11756374874.14557.24652.35757.59959.51423.12942 initated
[DICOM SCU] >> : Connected to: 192.168.1.36:5678
[DICOM SCU] >> : Item Type: 2
[DICOM SCU] >> : Association Accept::ReadDynamic
[DICOM SCU] >> : Presentation Context Accept, Transfer Syntax: 1.2.840.10008.1.2
[888] >> : c-find result: 1 sequences in study with UID 1.3.51.0.7.11756374874.14557.24652.35757.59959.51423.12942
[888] >> : c-move request for sequence 1.3.51.0.7.11395779729.58038.26437.43473.32644.20109.39996 initiated.
[DICOM SCU] >> : Connected to: 192.168.1.36:5678
[DICOM SCU] >> : Item Type: 2
[DICOM SCU] >> : Association Accept::ReadDynamic
[DICOM SCU] >> : Presentation Context Accept, Transfer Syntax: 1.2.840.10008.1.2
[DICOM SCU] >> : KPServer >> CONQUESTSRV1
[DICOM SCU] >> : (0000,0000), 4, UL, "92" - Group Length
[DICOM SCU] >> : (0000,0002), 28, UI, "1.2.840.10008.5.1.4.1.2.2.2" - Affected SOP Class UID
[DICOM SCU] >> : (0000,0100), 2, US, "33" - Command Field
[DICOM SCU] >> : (0000,0110), 2, US, "3" - Message ID
[DICOM SCU] >> : (0000,0600), 8, AE, "KPServer" - Move Destination
[DICOM SCU] >> : (0000,0700), 2, US, "0" - Priority
[DICOM SCU] >> : (0000,0800), 2, US, "258" - Data Set Type
[DICOM SCU] >> :
[DICOM SCU] >> : KPServer >> CONQUESTSRV1
[DICOM SCU] >> : (0008,0021), 0, DA, "" - Series Date
[DICOM SCU] >> : (0008,0031), 0, TM, "" - Series Time
[DICOM SCU] >> : (0008,0052), 6, CS, "SERIES" - Query/Retrieve Level
[DICOM SCU] >> : (0008,0060), 0, CS, "" - Modality
[DICOM SCU] >> : (0008,103E), 0, LO, "" - Series Description
[DICOM SCU] >> : (0018,0015), 0, CS, "" - Body Part Examined
[DICOM SCU] >> : (0018,1030), 0, LO, "" - Protocol Name
[DICOM SCU] >> : (0018,5100), 0, CS, "" - Patient Position
[DICOM SCU] >> : (0020,000D), 58, UI, "1.3.51.0.7.11756374874.14557.24652.35757.59959.51423.12942" - Study Instance UID
[DICOM SCU] >> : (0020,000E), 58, UI, "1.3.51.0.7.11395779729.58038.26437.43473.32644.20109.39996" - Series Instance UID
[DICOM SCU] >> : (0020,0011), 0, IS, "" - Series Number
[DICOM SCU] >> :
[DICOM SCU] >> : KPServer <<CONQUESTSRV1>> : (0000,0000), 4, UL, "116" - Group Length
[DICOM SCU] >> : (0000,0002), 28, UI, "1.2.840.10008.5.1.4.1.2.2.2" - Affected SOP Class UID
[DICOM SCU] >> : (0000,0100), 2, US, "32801" - Command Field
[DICOM SCU] >> : (0000,0120), 2, US, "3" - Message ID Being Responded To
[DICOM SCU] >> : (0000,0800), 2, US, "257" - Data Set Type
[DICOM SCU] >> : (0000,0900), 2, US, "65280" - Status
[DICOM SCU] >> : (0000,1020), 2, US, "0" - Number of Remaining Sub-Operations
[DICOM SCU] >> : (0000,1021), 2, US, "1" - Number of Completed Sub-Operations
[DICOM SCU] >> : (0000,1022), 2, US, "0" - Number of Failed Sub-Operations
[DICOM SCU] >> : (0000,1023), 2, US, "0" - Number of Warning Sub-Operations
[DICOM SCU] >> :
[DICOM SCU] >> : KPServer <<CONQUESTSRV1>> : (0000,0000), 4, UL, "116" - Group Length
[DICOM SCU] >> : (0000,0002), 28, UI, "1.2.840.10008.5.1.4.1.2.2.2" - Affected SOP Class UID
[DICOM SCU] >> : (0000,0100), 2, US, "32801" - Command Field
[DICOM SCU] >> : (0000,0120), 2, US, "3" - Message ID Being Responded To
[DICOM SCU] >> : (0000,0800), 2, US, "257" - Data Set Type
[DICOM SCU] >> : (0000,0900), 2, US, "0" - Status
[DICOM SCU] >> : (0000,1020), 2, US, "0" - Number of Remaining Sub-Operations
[DICOM SCU] >> : (0000,1021), 2, US, "1" - Number of Completed Sub-Operations
[DICOM SCU] >> : (0000,1022), 2, US, "0" - Number of Failed Sub-Operations
[DICOM SCU] >> : (0000,1023), 2, US, "0" - Number of Warning Sub-Operations
[DICOM SCU] >> :
[888] >> : c-move request for Bassa Th, Luca , Study: VOORHAND middel , Sequence: 1 successful.
[888] >> : c-move request for sequence 1.3.51.0.7.12094452639.38947.57671.40278.42073.13958.16455 initiated.
[DICOM SCU] >> : Connected to: 192.168.1.36:5678
[DICOM SCU] >> : Item Type: 2
[DICOM SCU] >> : Association Accept::ReadDynamic
[DICOM SCU] >> : Presentation Context Accept, Transfer Syntax: 1.2.840.10008.1.2
[DICOM SCU] >> : KPServer >> CONQUESTSRV1
[DICOM SCU] >> : (0000,0000), 4, UL, "92" - Group Length
[DICOM SCU] >> : (0000,0002), 28, UI, "1.2.840.10008.5.1.4.1.2.2.2" - Affected SOP Class UID
[DICOM SCU] >> : (0000,0100), 2, US, "33" - Command Field
[DICOM SCU] >> : (0000,0110), 2, US, "3" - Message ID
[DICOM SCU] >> : (0000,0600), 8, AE, "KPServer" - Move Destination
[DICOM SCU] >> : (0000,0700), 2, US, "0" - Priority
[DICOM SCU] >> : (0000,0800), 2, US, "258" - Data Set Type
[DICOM SCU] >> :
[DICOM SCU] >> : KPServer >> CONQUESTSRV1
[DICOM SCU] >> : (0008,0021), 0, DA, "" - Series Date
[DICOM SCU] >> : (0008,0031), 0, TM, "" - Series Time
[DICOM SCU] >> : (0008,0052), 6, CS, "SERIES" - Query/Retrieve Level
[DICOM SCU] >> : (0008,0060), 0, CS, "" - Modality
[DICOM SCU] >> : (0008,103E), 0, LO, "" - Series Description
[DICOM SCU] >> : (0018,0015), 0, CS, "" - Body Part Examined
[DICOM SCU] >> : (0018,1030), 0, LO, "" - Protocol Name
[DICOM SCU] >> : (0018,5100), 0, CS, "" - Patient Position
[DICOM SCU] >> : (0020,000D), 58, UI, "1.3.51.0.7.11756374874.14557.24652.35757.59959.51423.12942" - Study Instance UID
[DICOM SCU] >> : (0020,000E), 58, UI, "1.3.51.0.7.12094452639.38947.57671.40278.42073.13958.16455" - Series Instance UID
[DICOM SCU] >> : (0020,0011), 0, IS, "" - Series Number
[DICOM SCU] >> :
[DICOM SCU] >> : KPServer <<CONQUESTSRV1>> : (0000,0000), 4, UL, "116" - Group Length
[DICOM SCU] >> : (0000,0002), 28, UI, "1.2.840.10008.5.1.4.1.2.2.2" - Affected SOP Class UID
[DICOM SCU] >> : (0000,0100), 2, US, "32801" - Command Field
[DICOM SCU] >> : (0000,0120), 2, US, "3" - Message ID Being Responded To
[DICOM SCU] >> : (0000,0800), 2, US, "257" - Data Set Type
[DICOM SCU] >> : (0000,0900), 2, US, "65280" - Status
[DICOM SCU] >> : (0000,1020), 2, US, "0" - Number of Remaining Sub-Operations
[DICOM SCU] >> : (0000,1021), 2, US, "1" - Number of Completed Sub-Operations
[DICOM SCU] >> : (0000,1022), 2, US, "0" - Number of Failed Sub-Operations
[DICOM SCU] >> : (0000,1023), 2, US, "0" - Number of Warning Sub-Operations
[DICOM SCU] >> :
[DICOM SCU] >> : KPServer <<CONQUESTSRV1>> : (0000,0000), 4, UL, "116" - Group Length
[DICOM SCU] >> : (0000,0002), 28, UI, "1.2.840.10008.5.1.4.1.2.2.2" - Affected SOP Class UID
[DICOM SCU] >> : (0000,0100), 2, US, "32801" - Command Field
[DICOM SCU] >> : (0000,0120), 2, US, "3" - Message ID Being Responded To
[DICOM SCU] >> : (0000,0800), 2, US, "257" - Data Set Type
[DICOM SCU] >> : (0000,0900), 2, US, "0" - Status
[DICOM SCU] >> : (0000,1020), 2, US, "0" - Number of Remaining Sub-Operations
[DICOM SCU] >> : (0000,1021), 2, US, "1" - Number of Completed Sub-Operations
[DICOM SCU] >> : (0000,1022), 2, US, "0" - Number of Failed Sub-Operations
[DICOM SCU] >> : (0000,1023), 2, US, "0" - Number of Warning Sub-Operations
[DICOM SCU] >> :
[888] >> : c-move request for Bassa Th, Luca , Study: VOORHAND middel , Sequence: 2 successful.
*******************************************************
5-4-2008 11:26:45 ThreadID[888] : Closed.
Task processed successfully
*******************************************************
Any clues from someone?
Chris
Hi Marcel,
Works fine with IE but Firefox asks me to open or download the dgate.exe file.
The same with Konqueror under Linux.
Any tips?
Chris
Hi Andreas,
Tnx for your very quick reply and the tip. It works well now for any "normal" user.
Chris
Hi,
I just checked with an iQ-view evaluation install, just the same!
Chris
Hi all,
I've just installed K-Pacs 1.5 which works great together with Conquest.
There is only one problem.
When I log in as Administrator I can burn CD/DVD as K-Pacs finds the burning device without problem.
But as I login in as a normal user (with restricted privileges), K-Pacs cannot find the burning device ("CD/DVD writer" and the button "Burn CD/DVD" are greyed out).
Does anyone knows a solution for this (as I don't want normal users to log in with administrator privileges)?
Chris