Conquest and Ultrasound

  • Hello, We have used Conquest for several years for many applications, but have had poor results using it with Ultrasound. I have difficulty storing and forwarding color US, Cine US, and "Freestyle" US images. Some images store fine but then it stops and locks up. Is there anything I can try? We use Acuson US machines.

  • I also found problems with certain Accuson US images when using Conquest with NKI private compression. This is most certainly caused by the fact that NKI uses private tag 7fdf,0010 for pixel data whereas Accuson uses the same private tag for other data.


    In my opinion, the desicion to use a private tag for pixel data was not a good one in the first place. A more appropriate way would have been to use a private transfer syntax and store images with a meta header.


    And even with a private tag for pixel data, you should not have used 0010 because the first unused element in a private group is reserved for a private code mark to specify the creator of this group (in order to avoid the above problems).


    Regards,
    Andreas

  • Hello Lambert, Hello Andreas,
    Thanks for the replies.
    We have only stored as uncompressed. I tried JPEG Lossless, but had trouble with 12bit CR images and of course US Color JPEG/cine. We have never used NKI Private. Andreas have you ever had trouble storing uncompressed Acuson images in Conquest, like I am describing? KPACS will handle the Acuson well for viewing and storing, but I can not use it for longterm store and autorouting.
    I try to send some of the images. Again, thanks for all you guys do.

  • I am using Conquest 1.4.11 and 1.4.12alpha. When I get back to the office, I will update to 1.4.12. Should I put in the patch also?


    Not sure if it makes a difference, but the images I sent were not directly from the Acuson US machine. They were first stored in GE Centricity, and I retrieved them from there.


    I have put a cine image cine image loop at:http://rapidshare.com/files/11091367/CINE.zip
    They errored on Conquest even coming from GE. I got "Client error:Unknown Command 1" "connection terminated" as soon as the cine images started. .
    Also, I tried to run the cine in KPACS, and though they stored fine, but I couldn't actually play the cine.

  • I checked the first set of US images and they displayed correctly in K-PACS. Sending to Conquest (1.4.12) and receiving from Conquest worked as well. You should activate advanced debuglog in Conquest in order to catch the stops and lock ups.


    The multiframe (cine) US image is strange. Although it says 43 frames, K-PACS seems to fail to find the frame offsets. I will check what is going wrong.


    Regards,
    Andreas

  • The Cine series is indeed strange. There are 43 files, although the header says 43 frames, every file contains only one frame. This is in violation of the DICOM standard.
    Even more problematic is the fact that every file has the same SOPInstUID which is definitively a violation of the standard.


    Regards,
    Andreas

  • Hi Lambert,


    Quote

    AndreasKnopke wrote:
    In my opinion, the desicion to use a private tag for pixel data was not a good one in the first place. A more appropriate way would have been to use a private transfer syntax and store images with a meta header.


    Quote

    And then hope that every vendor will support that transfer syntax; not very realistic...


    No, there is no nead at all that this transfer syntax is supported by a third party. These file are for internal use only and this could remain that way.
    With a private TS you could avoid the use of private tags (which might be occupied) and give a third party application a chance to reject the image without having to find out what the transfer syntax is by advanced investigation. When sending the images via C-Store, Conquest uncompresses the image and recreates the metaheader.


    You could even propose your private TS in a C-Store association (along with LittleEndianImplicit). Marcel once implemented NKI support in K-PACS's viewer class. We could use NKI for a faster image transfer.


    Regards,
    Andreas

  • Andreas brought this discussion thread to my attention.


    I would concur with his assessment that if a private form of pixel data compression is to be used then a private transfer syntax is required.


    It is not legal in DICOM to transfer a standard SOP Class with the image pixel data encoded inconsistently with the standard Transfer Syntax, whether that pixel data be buried in a private attribute or in the standard (7FE0,0010) attribute; it is required that (7FE0,0010) be present and encoded in a manner consistent with the transfer syntax.


    I have observed this before with Conquest and the NKI compression, but decided not to make an issue of it, but since the question was asked ... I would strongly advise that this inappropriate NKI behavior either be removed from the code completely or implemented "properly" using a private transfer syntax. It would be extremely improper for folks to start accumulating archives full of images that claimed to be in a standard transfer syntax but were actually not.


    Also, I did take a look at those very wierd ultrasound images in the CINE.zip file that was posted ... some system (presumably Centricity) has really screwed those up, by splitting them into single frames in separate files; as Andreas noticed, the UIDs are the same ... actually the entire header for each file is the same, and if you cut out the pixel data and concatenate it and reassemble them into a single file with the same header it makes a not perfect but displayable US multi-frame image; the remaining problem is that it has an incorrect Photometric Interpretation - specifically if claims to be YBR_FULL_422 but is really YBR_FULL.


    Anyway, I mention this file because I would be interested in knowing the exact heritage of the file (how ti was passed from system to system), primarily so as to know whether or not I should start ragging on the Centricity guys about incorrect handling of US multiframe images or not.


    David

  • Because of limited knowledge on this, I will not be able to detail the system to system heritage, but I will give a history of what I did to get you the images.
    1. Acuson US sent images to GE Centricty in dicom (to my knowledge.) I think the cine are actually jpeg lossy.
    2. Centricity DAS recieved images and images are converted from dicom to GE proprietory progressive wavelet format for handling within Centricity STS. Images are then "reconverted" to JPEG lossless and stored in long term storage.
    3. Within centricity, I exported the series to be stored on my desktop in dicom. I zipped it and uploaded it.
    Within centricity, The images show as a separate series even if they were taken amidst the other images. It also appears this way on the WEB viewer. In the link below, I recorded the Dicom info that I took off of a Centrictiy workstation on the mutliframe cine.


    The images will transfer and view fine on the GE RA600 "dicom" workstation. Even the cine will start once I get to that image. I tried once more to send to the Kpacs workstation. The transfer stops when it gets to the cine set...then errors... without sending the rest of the study. I put the Kpacs communications log, and the proccess log in the link also. This probably means that we have been sending out Patient cd's with only partial exams on them...we use kpacs/IQview for patient cd's. The study burns well on cd's using the Centricity viewer, but everyone prefers to use IQview/Kpacs. Telerading will probably be out of the question also since we use IQview for this.


    http://rapidshare.com/files/11147350/Cine_Info_and_logs.zip

  • Quote

    Within centricity, I exported the series to be stored on my desktop in dicom. I zipped it and uploaded it.


    This step certainly does not seem to be producing valid DICOM files as per the previous discussion - when you performed this export, did the application offer you any options with respect to how the export was to be performed (e.g., as a single DICOM multi-frame file or seperate single frame files ?). Regardless, whatever errors are present in this export process may not be relevant to your primary question, which is how to transfer these over the network. You should report this incorrect export as a bug to your GE field engineer.


    You might also try burning a CD of the exam from within Centricity, and then passing along the files that are stored on that CD - that might work better than using the "export" function (or maybe not).


    As far as the network transfer is concerned, the dump from the GE workstation would seem to imply that the images is a JPEG lossy YBR_FULL_422 image with all the frames in a single object, which should be fine - perhaps the receiver you are using just doesn't support this correctly; have you tried sending the images to another open source tool, say the OFFIS dcmtk storescp utility - it allows you to be selective about what transfer syntaxes are accepted so you could explore what works and does not work with Centricity. You could probably also obtain a "correctly exported" version of the multi-frame JPEG file (use +B bit-preserving option in storescp to get this) that you could share for others to test with, which would be a more accurate representation of what Centricity is trying to send over the network.


    David

  • Quote

    when you performed this export, did the application offer you any options with respect to how the export was to be performed


    When I exported the images, I was only given the options of Dicom, JPEG, Tiff, or PNG. I could either save to disk or send to a workstation. If I sent to a workstation, I would get errors like mentioned.


    Here is an entire CD with the burned patient and viewer on it.:
    http://rapidshare.com/files/11225542/Burned_CD.zip


    Quote

    the dump from the GE workstation would seem to imply that the images is a JPEG lossy YBR_FULL_422 image with all the frames in a single object, which should be fine - perhaps the receiver you are using just doesn't support this correctly


    So maybe KPACS and Conquest do not support recieving CINE in this fashion:
    Which brings us back to the original problem. I would get the same errors when sending directly from the Acuson machine to Conquest without ever hitting centricity. So with all of this, my dilema is:


    1. I cannot autoroute or store images in ConQuest for telerad purposes. Whether sending directly from the machine, or Cmove out of Centricity.


    2. I cannot send a complete study, whether directly or indirectly, to KPACS/IQView for telerad and CD burning purposes.


    Doesnt this mean that the images are messed up from the beginning, and Centricity is somehow able to make sense of it? Has anyone ever successfully recieved Acuson Cine images on either Kpacs or Conquest?[/quote]

  • I dont know if this helps, but I experimented with this.
    1. I updated Conquest to latest version.
    2. I sent the cine study to it, and it stored, but there was an error:
    [CONQUESTMRI] [DecompressImage]: internal decompressor does not support color jpeg, using external decompressor
    [CONQUESTMRI] ***[DecompressImage]: Error on load after external decompression, image not decompressed

    If I forced display, I could see the first image and the other images above 13,.....
    3. When I queried Conquest with IQView, the exam tree showed all images and that there was 43 frames in image 12. A query of GE will only let me see to the study level.
    4. As I went to retrieve from Conquest with either GE Workstation or IQView, only the first 11 images showed up. and lots of beeps... errors.
    5. As I retrieve directly from GE, with IQview, only 11 images show up, and lots of beeps with an error...... though... the GE Dicom Workstation will retireve fine.

    In the case of Conquest....Could the failure of the color jpeg to decompress be causing the Conquest failure. That being said, IQview still cannot view images retrieved from either Conquest or PACS or cine being sent from Acuson...

  • in order to find out why the iQServer fails to store the images you can start it manually in verbose mode.


    To do so open a windows command shell (start->run->"cmd") and go to the server's directory ("cd" for change directory)
    run it like


    iqserver.exe -d -dw -od "c:\program files\iq-View\server\database" -aet iQSERVER 104


    Make sure that iQSERVER is not already running


    Please post the command shell logs.


    Regards,
    Andreas

  • I got the CD image thanks.


    It contains no problems with the encoding of the multi-frame images; further all of the images (formerly JPEG presumably) have been decompressed, since the CD is presumably written to the General Purpose CD Profile, which does not allow compression. The color space had also been transformed to RGB, which means anything can view them, which is nice. There are some ultrasound-specific header attributes that are bad but that is presumably propagated from the original Acuson, and they are relatively harmless.


    So obviously the Centricity has good data inside and is capable of processing it "properly" when writing a CD.


    Why the other "export" mechanism you described screws up remains a mystery (you should file a bug report about that with your local GE FE, if I didn't already mention that).


    As for transferring to Conquest, it would seem expedient to try and find a way to have Conquest refuse the JPEG transfer syntaxes from this source (or at all) ... that would force the Centricity to do the decompression and hopefully color space transformation as well. Since I don't use Conquest, I do not know off hand how to do that, but I am sure there is a way.


    David

  • Yes, the server complains about a lot of non standard VR's. The parser probably looses offset because auf an invalid group length.


    What you could try is the following:
    since iQSERVER is a modified OFFIS STORESCP.exe, all OFFIS parameters work for it as well.


    run the server manually again and try different options, for instance


    iqserver.exe -d +B +xi -dw -od "c:\program files\iq-View\server\database" -aet iQSERVER 104


    for bit preserving (write dataset exactly as received) and implicit only transfer.


    You could also try the +g option (create group length)


    If it still fails to receive the images, please mail me the logs with the different options again. If not, then tell me the parameter you used that I can implement it into the server administration as a compatibility option.


    Regards,
    Andreas