Hi,
We sometimes get DICOM images that Conquest can't import.
Error with original DICOM file:
***Explicit_ParseRaw encountered an invalid element length during load of DCM file (in 49534d45)
***Length = 8275
***Continuing parsing
***(Exp) Encountered an invalid group order during load of DCM file (after 49534d45)
I had some trouble making an anonymized version because some anonymizers corrected the error in the DICOM file
The error changes to:
***Length exceeds remaining file size: 00002053
When i dump the original DICOM header with DCMTK i get the following warning that i believe is the problem in these files:
W: Found element (0029,1140) with VR UN and undefined length, reading a sequence with transfer syntax LittleEndianImplicit (CP-246)
if i remove this tag (sequence) the problem goes away.
Is there anything that can be done on the Conquest side to make Conquest read these DICOM files?
Attached the anonymized version.
Regards,
Robert