Modality Worklist Not Work in CRVita ImageSuite

  • Hi,


    I've been setting correctly but does not work well in Carestream CRVita ImageSuite. Though setting PACS works well. For comparation with CT99. This is log :


    Thank's



    [CONQUESTSRV2]
    [CONQUESTSRV2] UPACS THREAD 39203: STARTED AT: Thu Nov 20 12:54:24 2014
    [CONQUESTSRV2] A-ASSOCIATE-RQ Packet Dump
    [CONQUESTSRV2] Calling Application Title : "ImageSuite "
    [CONQUESTSRV2] Called Application Title : "conquestsrv2 "
    [CONQUESTSRV2] Application Context : "1.2.840.10008.3.1.1.1", PDU length: 65536
    [CONQUESTSRV2] Number of Proposed Presentation Contexts: 1
    [CONQUESTSRV2] Presentation Context 0 "1.2.840.10008.5.1.4.31" 1
    [CONQUESTSRV2] Server Command := 0020
    [CONQUESTSRV2] Message ID := 0000
    [CONQUESTSRV2] 0000,0002 22 UI AffectedSOPClassUID "1.2.840.10008.5.1.4.31"
    [CONQUESTSRV2] 0000,0100 2 US CommandField 32
    [CONQUESTSRV2] 0000,0110 2 US MessageID 0
    [CONQUESTSRV2] 0000,0700 2 US Priority 0
    [CONQUESTSRV2] 0000,0800 2 US DataSetType 258
    [CONQUESTSRV2] 0002,0010 17 UI TransferSyntaxUID "1.2.840.10008.1.2"
    [CONQUESTSRV2] (ModalityWorkListQuery) search level:
    [CONQUESTSRV2] 0002,0010 17 UI TransferSyntaxUID "1.2.840.10008.1.2"
    [CONQUESTSRV2] 0008,0050 0 SH AccessionNumber (empty)
    [CONQUESTSRV2] 0008,0090 0 PN ReferringPhysicianNa (empty)
    [CONQUESTSRV2] 0010,0010 0 PN PatientName (empty)
    [CONQUESTSRV2] 0010,0020 0 LO PatientID (empty)
    [CONQUESTSRV2] 0010,0030 0 DA PatientBirthDate (empty)
    [CONQUESTSRV2] 0010,0040 0 CS PatientSex (empty)
    [CONQUESTSRV2] 0010,1000 0 LO OtherPatientIDs (empty)
    [CONQUESTSRV2] 0020,000d 0 UI StudyInstanceUID (empty)
    [CONQUESTSRV2] 0032,1060 0 LO RequestedProcedureDe (empty)
    [CONQUESTSRV2] 0032,1064 0 SQ RequestedProcedureCo Sequence Item
    [CONQUESTSRV2] >0008,0100 0 SH CodeValue (empty)
    [CONQUESTSRV2] >0008,0104 0 LO CodeMeaning (empty)
    [CONQUESTSRV2] 0038,0008 0 CS VisitStatusID (empty)
    [CONQUESTSRV2] 0038,0300 0 LO CurrentPatientLocati (empty)
    [CONQUESTSRV2] 0040,0100 0 SQ ScheduledProcedureSt Sequence Item
    [CONQUESTSRV2] >0008,0060 0 CS Modality (empty)
    [CONQUESTSRV2] >0032,1070 0 LO RequestedContrastAge (empty)
    [CONQUESTSRV2] >0040,0002 18 DA ScheduledProcedureSt "20141120-20141120 "
    [CONQUESTSRV2] >0040,0003 14 TM ScheduledProcedureSt "115338-125338 "
    [CONQUESTSRV2] >0040,0004 0 DA ScheduledProcedureSt (empty)
    [CONQUESTSRV2] >0040,0008 0 SQ ScheduledProtocolCod Sequence Item
    [CONQUESTSRV2] >>0008,0100 0 SH CodeValue (empty)
    [CONQUESTSRV2] >0040,0009 0 SH ScheduledProcedureSt (empty)
    [CONQUESTSRV2] 0040,1001 0 SH RequestedProcedureID (empty)
    [CONQUESTSRV2] 0040,1005 0 LO RequestedProcedureLo (empty)
    [CONQUESTSRV2] Query On Modality WorkList
    [CONQUESTSRV2] Queried item 0008 0090 is not in the database
    [CONQUESTSRV2] Queried item 0010 1000 is not in the database
    [CONQUESTSRV2] Queried item 0008 0100 is not in the database
    [CONQUESTSRV2] Queried item 0008 0104 is not in the database
    [CONQUESTSRV2] Queried item 0038 0008 is not in the database
    [CONQUESTSRV2] Queried item 0038 0300 is not in the database
    [CONQUESTSRV2] Queried item 0040 0004 is not in the database
    [CONQUESTSRV2] Queried item 0008 0100 is not in the database
    [CONQUESTSRV2] Queried item 0040 1005 is not in the database
    [CONQUESTSRV2] Issue Query on Columns: DICOMWorkList.AccessionN, DICOMWorkList.PatientNam, DICOMWorkList.PatientID, DICOMWorkList.PatientBir, DICOMWorkList.PatientSex, DICOMWorkList.StudyInsta, DICOMWorkList.ReqProcDes, DICOMWorkList.Modality, DICOMWorkList.ReqContras, DICOMWorkList.StartDate, DICOMWorkList.StartTime, DICOMWorkList.SchedPSID, DICOMWorkList.ReqProcID
    [CONQUESTSRV2] Values: DICOMWorkList.StartDate >= '20141120' and DICOMWorkList.StartDate <= '20141120' and DICOMWorkList.StartTime >= '115338' and DICOMWorkList.StartTime <= '125338'
    [CONQUESTSRV2] Tables: DICOMWorkList
    [CONQUESTSRV2] Sorting (DICOMWorkList.PatientNam) DoSort := 1
    [CONQUESTSRV2] Query Distinct Tables: DICOMWorkList
    [CONQUESTSRV2] Columns : DICOMWorkList.AccessionN, DICOMWorkList.PatientNam, DICOMWorkList.PatientID, DICOMWorkList.PatientBir, DICOMWorkList.PatientSex, DICOMWorkList.StudyInsta, DICOMWorkList.ReqProcDes, DICOMWorkList.Modality, DICOMWorkList.ReqContras, DICOMWorkList.StartDate, DICOMWorkList.StartTime, DICOMWorkList.SchedPSID, DICOMWorkList.ReqProcID
    [CONQUESTSRV2] Where : DICOMWorkList.StartDate >= '20141120' and DICOMWorkList.StartDate <= '20141120' and DICOMWorkList.StartTime >= '115338' and DICOMWorkList.StartTime <= '125338'
    [CONQUESTSRV2] Order : DICOMWorkList.PatientNam
    [CONQUESTSRV2] Records = 0
    [CONQUESTSRV2] C-Find (Modality Work List) located 0 records
    [CONQUESTSRV2] UPACS THREAD 39203: ENDED AT: Thu Nov 20 12:54:25 2014
    [CONQUESTSRV2] UPACS THREAD 39203: TOTAL RUNNING TIME: 1 SECONDS



    [CONQUESTSRV2] Connected by address: 0f02a8c0
    [CONQUESTSRV2] Testing transfer: '1.2.840.10008.1.2' against list #0 = '1.2.840.10008.1.2'
    [CONQUESTSRV2]
    [CONQUESTSRV2] UPACS THREAD 39236: STARTED AT: Thu Nov 20 13:04:16 2014
    [CONQUESTSRV2] A-ASSOCIATE-RQ Packet Dump
    [CONQUESTSRV2] Calling Application Title : "ct99 "
    [CONQUESTSRV2] Called Application Title : "conquestsrv2 "
    [CONQUESTSRV2] Application Context : "1.2.840.10008.3.1.1.1", PDU length: 65464
    [CONQUESTSRV2] Number of Proposed Presentation Contexts: 1
    [CONQUESTSRV2] Presentation Context 0 "1.2.840.10008.5.1.4.31" 1
    [CONQUESTSRV2] Server Command := 0020
    [CONQUESTSRV2] Message ID := 0000
    [CONQUESTSRV2] 0000,0002 22 UI AffectedSOPClassUID "1.2.840.10008.5.1.4.31"
    [CONQUESTSRV2] 0000,0100 2 US CommandField 32
    [CONQUESTSRV2] 0000,0110 2 US MessageID 0
    [CONQUESTSRV2] 0000,0700 2 US Priority 0
    [CONQUESTSRV2] 0000,0800 2 US DataSetType 258
    [CONQUESTSRV2] 0002,0010 17 UI TransferSyntaxUID "1.2.840.10008.1.2"
    [CONQUESTSRV2] (ModalityWorkListQuery) search level:
    [CONQUESTSRV2] 0002,0010 17 UI TransferSyntaxUID "1.2.840.10008.1.2"
    [CONQUESTSRV2] 0008,0005 10 CS SpecificCharacterSet "ISO_IR 100"
    [CONQUESTSRV2] 0008,0050 0 SH AccessionNumber (empty)
    [CONQUESTSRV2] 0008,0080 0 LO InstitutionName (empty)
    [CONQUESTSRV2] 0008,0090 0 PN ReferringPhysicianNa (empty)
    [CONQUESTSRV2] 0008,1110 0 SQ ReferencedStudySeque Sequence Item
    [CONQUESTSRV2] >0008,1150 0 UI ReferencedSOPClassUI (empty)
    [CONQUESTSRV2] >0008,1155 0 UI ReferencedSOPInstanc (empty)
    [CONQUESTSRV2] 0008,1120 0 SQ ReferencedPatientSeq Sequence Item
    [CONQUESTSRV2] >0008,1150 0 UI ReferencedSOPClassUI (empty)
    [CONQUESTSRV2] >0008,1155 0 UI ReferencedSOPInstanc (empty)
    [CONQUESTSRV2] 0010,0010 0 PN PatientName (empty)
    [CONQUESTSRV2] 0010,0020 0 LO PatientID (empty)
    [CONQUESTSRV2] 0010,0030 0 DA PatientBirthDate (empty)
    [CONQUESTSRV2] 0010,0040 0 CS PatientSex (empty)
    [CONQUESTSRV2] 0010,1000 0 LO OtherPatientIDs (empty)
    [CONQUESTSRV2] 0010,1020 0 DS PatientSize (empty)
    [CONQUESTSRV2] 0010,1030 0 DS PatientWeight (empty)
    [CONQUESTSRV2] 0010,1040 0 LO PatientAddress (empty)
    [CONQUESTSRV2] 0010,2000 0 LO MedicalAlerts (empty)
    [CONQUESTSRV2] 0010,2110 0 LO Allergies (empty)
    [CONQUESTSRV2] 0010,2154 0 SH PatientTelephoneNumb (empty)
    [CONQUESTSRV2] 0010,2160 0 SH EthnicGroup (empty)
    [CONQUESTSRV2] 0010,21b0 0 LT AdditionalPatientHis (empty)
    [CONQUESTSRV2] 0010,21c0 0 US PregnancyStatus (empty)
    [CONQUESTSRV2] 0010,4000 0 LT PatientComments (empty)
    [CONQUESTSRV2] 0020,000d 0 UI StudyInstanceUID (empty)
    [CONQUESTSRV2] 0032,1032 0 PN RequestingPhysician (empty)
    [CONQUESTSRV2] 0032,1033 0 LO RequestingService (empty)
    [CONQUESTSRV2] 0032,1060 0 LO RequestedProcedureDe (empty)
    [CONQUESTSRV2] 0032,1064 0 SQ RequestedProcedureCo Sequence Item
    [CONQUESTSRV2] >0008,0100 0 SH CodeValue (empty)
    [CONQUESTSRV2] >0008,0102 0 SH CodingSchemeDesignat (empty)
    [CONQUESTSRV2] >0008,0104 0 LO CodeMeaning (empty)
    [CONQUESTSRV2] 0038,0010 0 LO AdmissionID (empty)
    [CONQUESTSRV2] 0038,0050 0 LO SpecialNeeds (empty)
    [CONQUESTSRV2] 0038,0300 0 LO CurrentPatientLocati (empty)
    [CONQUESTSRV2] 0038,0500 0 LO PatientState (empty)
    [CONQUESTSRV2] 0040,0100 0 SQ ScheduledProcedureSt Sequence Item
    [CONQUESTSRV2] >0008,0060 2 CS Modality "CT"
    [CONQUESTSRV2] >0032,1070 0 LO RequestedContrastAge (empty)
    [CONQUESTSRV2] >0040,0001 0 AE ScheduledStationAETi (empty)
    [CONQUESTSRV2] >0040,0002 18 DA ScheduledProcedureSt "20141120-20141120 "
    [CONQUESTSRV2] >0040,0003 0 TM ScheduledProcedureSt (empty)
    [CONQUESTSRV2] >0040,0006 0 PN ScheduledPerformingP (empty)
    [CONQUESTSRV2] >0040,0007 0 LO ScheduledProcedureSt (empty)
    [CONQUESTSRV2] >0040,0008 0 SQ ScheduledProtocolCod Sequence Item
    [CONQUESTSRV2] >>0008,0100 0 SH CodeValue (empty)
    [CONQUESTSRV2] >>0008,0102 0 SH CodingSchemeDesignat (empty)
    [CONQUESTSRV2] >>0008,0104 0 LO CodeMeaning (empty)
    [CONQUESTSRV2] >0040,0009 0 SH ScheduledProcedureSt (empty)
    [CONQUESTSRV2] >0040,0010 0 SH ScheduledStationName (empty)
    [CONQUESTSRV2] >0040,0011 0 SH ScheduledProcedureSt (empty)
    [CONQUESTSRV2] >0040,0012 0 LO PreMedication (empty)
    [CONQUESTSRV2] 0040,1001 0 SH RequestedProcedureID (empty)
    [CONQUESTSRV2] 0040,1003 0 SH RequestedProcedurePr (empty)
    [CONQUESTSRV2] 0040,1004 0 LO PatientTransportArra (empty)
    [CONQUESTSRV2] 0040,1005 0 LO RequestedProcedureLo (empty)
    [CONQUESTSRV2] 0040,1008 0 LO ConfidentialityCode (empty)
    [CONQUESTSRV2] 0040,1010 0 PN NamesOfIntendedRecip (empty)
    [CONQUESTSRV2] 0040,1400 0 LT RequestedProcedureCo (empty)
    [CONQUESTSRV2] 0040,2400 0 LT ImagingServiceReques (empty)
    [CONQUESTSRV2] 0040,3001 0 LO ConfidentialityConst (empty)
    [CONQUESTSRV2] Query On Modality WorkList
    [CONQUESTSRV2] Queried item 0008 0005 is not in the database
    [CONQUESTSRV2] Queried item 0008 0080 is not in the database
    [CONQUESTSRV2] Queried item 0008 0090 is not in the database
    [CONQUESTSRV2] Queried item 0008 1150 is not in the database
    [CONQUESTSRV2] Queried item 0008 1155 is not in the database
    [CONQUESTSRV2] Queried item 0008 1150 is not in the database
    [CONQUESTSRV2] Queried item 0008 1155 is not in the database
    [CONQUESTSRV2] Queried item 0010 1000 is not in the database
    [CONQUESTSRV2] Queried item 0010 1020 is not in the database
    [CONQUESTSRV2] Queried item 0010 1030 is not in the database
    [CONQUESTSRV2] Queried item 0010 1040 is not in the database
    [CONQUESTSRV2] Queried item 0010 2154 is not in the database
    [CONQUESTSRV2] Queried item 0010 2160 is not in the database
    [CONQUESTSRV2] Queried item 0010 21b0 is not in the database
    [CONQUESTSRV2] Queried item 0010 21c0 is not in the database
    [CONQUESTSRV2] Queried item 0010 4000 is not in the database
    [CONQUESTSRV2] Queried item 0032 1033 is not in the database
    [CONQUESTSRV2] Queried item 0008 0100 is not in the database
    [CONQUESTSRV2] Queried item 0008 0102 is not in the database
    [CONQUESTSRV2] Queried item 0008 0104 is not in the database
    [CONQUESTSRV2] Queried item 0038 0010 is not in the database
    [CONQUESTSRV2] Queried item 0038 0050 is not in the database
    [CONQUESTSRV2] Queried item 0038 0300 is not in the database
    [CONQUESTSRV2] Queried item 0038 0500 is not in the database
    [CONQUESTSRV2] Queried item 0008 0100 is not in the database
    [CONQUESTSRV2] Queried item 0008 0102 is not in the database
    [CONQUESTSRV2] Queried item 0008 0104 is not in the database
    [CONQUESTSRV2] Queried item 0040 1004 is not in the database
    [CONQUESTSRV2] Queried item 0040 1005 is not in the database
    [CONQUESTSRV2] Queried item 0040 1008 is not in the database
    [CONQUESTSRV2] Queried item 0040 1010 is not in the database
    [CONQUESTSRV2] Queried item 0040 1400 is not in the database
    [CONQUESTSRV2] Queried item 0040 2400 is not in the database
    [CONQUESTSRV2] Queried item 0040 3001 is not in the database
    [CONQUESTSRV2] Issue Query on Columns: DICOMWorkList.AccessionN, DICOMWorkList.PatientNam, DICOMWorkList.PatientID, DICOMWorkList.PatientBir, DICOMWorkList.PatientSex, DICOMWorkList.MedicalAle, DICOMWorkList.ContrastAl, DICOMWorkList.StudyInsta, DICOMWorkList.ReqPhysici, DICOMWorkList.ReqProcDes, DICOMWorkList.Modality, DICOMWorkList.ReqContras, DICOMWorkList.ScheduledA, DICOMWorkList.StartDate, DICOMWorkList.StartTime, DICOMWorkList.PerfPhysic, DICOMWorkList.SchedPSDes, DICOMWorkList.SchedPSID, DICOMWorkList.SchedStati, DICOMWorkList.SchedPSLoc, DICOMWorkList.PreMedicat, DICOMWorkList.ReqProcID, DICOMWorkList.ReqProcPri
    [CONQUESTSRV2] Values: DICOMWorkList.Modality = 'CT' and DICOMWorkList.StartDate >= '20141120' and DICOMWorkList.StartDate <= '20141120'
    [CONQUESTSRV2] Tables: DICOMWorkList
    [CONQUESTSRV2] Sorting (DICOMWorkList.PatientNam) DoSort := 1
    [CONQUESTSRV2] Query Distinct Tables: DICOMWorkList
    [CONQUESTSRV2] Columns : DICOMWorkList.AccessionN, DICOMWorkList.PatientNam, DICOMWorkList.PatientID, DICOMWorkList.PatientBir, DICOMWorkList.PatientSex, DICOMWorkList.MedicalAle, DICOMWorkList.ContrastAl, DICOMWorkList.StudyInsta, DICOMWorkList.ReqPhysici, DICOMWorkList.ReqProcDes, DICOMWorkList.Modality, DICOMWorkList.ReqContras, DICOMWorkList.ScheduledA, DICOMWorkList.StartDate, DICOMWorkList.StartTime, DICOMWorkList.PerfPhysic, DICOMWorkList.SchedPSDes, DICOMWorkList.SchedPSID, DICOMWorkList.SchedStati, DICOMWorkList.SchedPSLoc, DICOMWorkList.PreMedicat, DICOMWorkList.ReqProcID, DICOMWorkList.ReqProcPri
    [CONQUESTSRV2] Where : DICOMWorkList.Modality = 'CT' and DICOMWorkList.StartDate >= '20141120' and DICOMWorkList.StartDate <= '20141120'
    [CONQUESTSRV2] Order : DICOMWorkList.PatientNam
    [CONQUESTSRV2] 0008,0050 10 SH AccessionNumber "1411108037"
    [CONQUESTSRV2] Coding sequence (0008,1110), level=1
    [CONQUESTSRV2] (sequence is empty)
    [CONQUESTSRV2] Coding sequence (0008,1120), level=1
    [CONQUESTSRV2] (sequence is empty)
    [CONQUESTSRV2] 0010,0010 6 PN PatientName "XXXXXX"
    [CONQUESTSRV2] 0010,0020 16 LO PatientID "1411108037C0095 "
    [CONQUESTSRV2] 0010,0030 8 DA PatientBirthDate "19630209"
    [CONQUESTSRV2] 0010,0040 2 CS PatientSex "M "
    [CONQUESTSRV2] 0010,2000 0 LO MedicalAlerts (empty)
    [CONQUESTSRV2] 0010,2110 0 LO Allergies (empty)
    [CONQUESTSRV2] 0020,000d 0 UI StudyInstanceUID (empty)
    [CONQUESTSRV2] First record of cleaned response:
    [CONQUESTSRV2] UPACS THREAD 39236: ENDED AT: Thu Nov 20 13:04:16 2014
    [CONQUESTSRV2] UPACS THREAD 39236: TOTAL RUNNING TIME: 0 SECONDS

  • Hi,


    I guess '' is not between two valid times. To make this work you could add to dicom.ini:


    [lua]
    QueryWorklistConverter0 = Data.ScheduledProcedureStep[1].ScheduledProcedureStepTime = ''


    If this is correct (untested) the query will ignore the time that is requested.


    Marcel

  • Hi, Marcel


    I problem again, but connect from DR Siemens.
    I think success this worklist conquest, but database worklist no visible in DR Siemens


    [PACSRSPMI] Values: DICOMWorkList.Modality = E'CR'
    [PACSRSPMI] Tables: DICOMWorkList
    [PACSRSPMI] Sorting (DICOMWorkList.PatientNam) DoSort := 1
    [PACSRSPMI] Query Distinct Tables: DICOMWorkList
    [PACSRSPMI] Columns : DICOMWorkList.AccessionN, DICOMWorkList.PatientNam, DICOMWorkList.PatientID, DICOMWorkList.PatientBir, DICOMWorkList.PatientSex, DICOMWorkList.MedicalAle, DICOMWorkList.ContrastAl, DICOMWorkList.StudyInsta, DICOMWorkList.ReqPhysici, DICOMWorkList.ReqProcDes, DICOMWorkList.Modality, DICOMWorkList.ReqContras, DICOMWorkList.ScheduledA, DICOMWorkList.StartDate, DICOMWorkList.StartTime, DICOMWorkList.PerfPhysic, DICOMWorkList.SchedPSDes, DICOMWorkList.SchedPSID, DICOMWorkList.SchedPSLoc, DICOMWorkList.PreMedicat, DICOMWorkList.ReqProcID, DICOMWorkList.ReqProcPri
    [PACSRSPMI] Where : DICOMWorkList.Modality = E'CR'
    [PACSRSPMI] Order : DICOMWorkList.PatientNam
    [PACSRSPMI] 0008,0050 10 SH AccessionNumber "123456789 "
    [PACSRSPMI] Coding sequence (0008,1110), level=1
    [PACSRSPMI] (sequence is empty)
    [PACSRSPMI] Coding sequence (0008,1120), level=1
    [PACSRSPMI] (sequence is empty)
    [PACSRSPMI] 0010,0010 14 PN PatientName "----"
    [PACSRSPMI] 0010,0020 10 LO PatientID "R1608.0501"
    [PACSRSPMI] 0010,0030 8 DA PatientBirthDate "19781229"
    [PACSRSPMI] 0010,0040 2 CS PatientSex "M "
    [PACSRSPMI] 0010,2000 0 LO MedicalAlerts (empty)
    [PACSRSPMI] 0010,2110 0 LO Allergies (empty)
    [PACSRSPMI] 0020,000d 0 UI StudyInstanceUID (empty)
    [PACSRSPMI] 0032,1032 14 PN RequestingPhysician "dr. And Lomata"
    [PACSRSPMI] 0032,1060 0 LO RequestedProcedureDe (empty)
    [PACSRSPMI] Coding sequence (0032,1064), level=1
    [PACSRSPMI] (sequence is empty)
    [PACSRSPMI] Coding sequence (0040,0100), level=1
    [PACSRSPMI] 0008,0060 2 CS Modality "CR"
    [PACSRSPMI] 0032,1070 0 LO RequestedContrastAge (empty)
    [PACSRSPMI] 0040,0001 0 AE ScheduledStationAETi (empty)
    [PACSRSPMI] 0040,0002 8 DA ScheduledProcedureSt "20160802"
    [PACSRSPMI] 0040,0003 6 TM ScheduledProcedureSt "010000"
    [PACSRSPMI] 0040,0006 0 PN ScheduledPerformingP (empty)
    [PACSRSPMI] 0040,0007 0 LO ScheduledProcedureSt (empty)
    [PACSRSPMI] Coding sequence (0040,0008), level=2
    [PACSRSPMI] (sequence is empty)
    [PACSRSPMI] 0040,0009 0 SH ScheduledProcedureSt (empty)
    [PACSRSPMI] 0040,0011 0 SH ScheduledProcedureSt (empty)
    [PACSRSPMI] 0040,0012 0 LO PreMedication (empty)
    [PACSRSPMI] 0040,1001 0 SH RequestedProcedureID (empty)
    [PACSRSPMI] 0040,1003 0 SH RequestedProcedurePr (empty)
    [PACSRSPMI] 0008,0050 10 SH AccessionNumber "4423123123"
    [PACSRSPMI] Coding sequence (0008,1110), level=1
    [PACSRSPMI] (sequence is empty)
    [PACSRSPMI] Coding sequence (0008,1120), level=1
    [PACSRSPMI] (sequence is empty)
    [PACSRSPMI] 0010,0010 18 PN PatientName "...."
    [PACSRSPMI] 0010,0020 10 LO PatientID "C1608.0010"
    [PACSRSPMI] 0010,0030 8 DA PatientBirthDate "19781229"
    [PACSRSPMI] 0010,0040 2 CS PatientSex "M "
    [PACSRSPMI] 0010,2000 0 LO MedicalAlerts (empty)
    [PACSRSPMI] 0010,2110 0 LO Allergies (empty)
    [PACSRSPMI] 0020,000d 0 UI StudyInstanceUID (empty)
    [PACSRSPMI] 0032,1032 16 PN RequestingPhysician "dr. Ari Soetoto "
    [PACSRSPMI] 0032,1060 0 LO RequestedProcedureDe (empty)
    [PACSRSPMI] Coding sequence (0032,1064), level=1
    [PACSRSPMI] (sequence is empty)
    [PACSRSPMI] Coding sequence (0040,0100), level=1
    [PACSRSPMI] 0008,0060 2 CS Modality "CR"
    [PACSRSPMI] 0032,1070 0 LO RequestedContrastAge (empty)
    [PACSRSPMI] 0040,0001 0 AE ScheduledStationAETi (empty)
    [PACSRSPMI] 0040,0002 8 DA ScheduledProcedureSt "20160802"
    [PACSRSPMI] 0040,0003 6 TM ScheduledProcedureSt "010000"
    [PACSRSPMI] 0040,0006 0 PN ScheduledPerformingP (empty)
    [PACSRSPMI] 0040,0007 0 LO ScheduledProcedureSt (empty)
    [PACSRSPMI] Coding sequence (0040,0008), level=2
    [PACSRSPMI] (sequence is empty)
    [PACSRSPMI] 0040,0009 0 SH ScheduledProcedureSt (empty)
    [PACSRSPMI] 0040,0011 0 SH ScheduledProcedureSt (empty)
    [PACSRSPMI] 0040,0012 0 LO PreMedication (empty)
    [PACSRSPMI] 0040,1001 0 SH RequestedProcedureID (empty)
    [PACSRSPMI] 0040,1003 0 SH RequestedProcedurePr (empty)
    [PACSRSPMI] Records = 2
    [PACSRSPMI] First record of cleaned response:
    [PACSRSPMI] 0008,0005 10 CS SpecificCharacterSet "ISO_IR 100"
    [PACSRSPMI] 0008,0050 10 SH AccessionNumber "123456789 "
    [PACSRSPMI] 0008,1110 0 SQ ReferencedStudySeque Sequence Item
    [PACSRSPMI] 0008,1120 0 SQ ReferencedPatientSeq Sequence Item
    [PACSRSPMI] 0010,0010 14 PN PatientName "...."
    [PACSRSPMI] 0010,0020 10 LO PatientID "R1608.0501"
    [PACSRSPMI] 0010,0030 8 DA PatientBirthDate "19781229"
    [PACSRSPMI] 0010,0040 2 CS PatientSex "M "
    [PACSRSPMI] 0010,2000 0 LO MedicalAlerts (empty)
    [PACSRSPMI] 0010,2110 0 LO Allergies (empty)
    [PACSRSPMI] 0020,000d 0 UI StudyInstanceUID (empty)
    [PACSRSPMI] 0032,1032 14 PN RequestingPhysician "dr. And Lomata"
    [PACSRSPMI] 0032,1060 0 LO RequestedProcedureDe (empty)
    [PACSRSPMI] 0032,1064 0 SQ RequestedProcedureCo Sequence Item
    [PACSRSPMI] 0040,0100 0 SQ ScheduledProcedureSt Sequence Item
    [PACSRSPMI] >0008,0060 2 CS Modality "CR"
    [PACSRSPMI] >0032,1070 0 LO RequestedContrastAge (empty)
    [PACSRSPMI] >0040,0001 0 AE ScheduledStationAETi (empty)
    [PACSRSPMI] >0040,0002 8 DA ScheduledProcedureSt "20160802"
    [PACSRSPMI] >0040,0003 6 TM ScheduledProcedureSt "010000"
    [PACSRSPMI] >0040,0006 0 PN ScheduledPerformingP (empty)
    [PACSRSPMI] >0040,0007 0 LO ScheduledProcedureSt (empty)
    [PACSRSPMI] >0040,0008 0 SQ ScheduledProtocolCod Sequence Item
    [PACSRSPMI] >0040,0009 0 SH ScheduledProcedureSt (empty)
    [PACSRSPMI] >0040,0011 0 SH ScheduledProcedureSt (empty)
    [PACSRSPMI] >0040,0012 0 LO PreMedication (empty)
    [PACSRSPMI] 0040,1001 0 SH RequestedProcedureID (empty)
    [PACSRSPMI] 0040,1003 0 SH RequestedProcedurePr (empty)
    [PACSRSPMI] C-Find (Modality Work List) located 2 records
    [PACSRSPMI] UPACS THREAD 117: ENDED AT: Tue Aug 02 15:55:16 2016
    [PACSRSPMI] UPACS THREAD 117: TOTAL RUNNING TIME: 0 SECONDS


    Thanks for any suggestions

  • Hi,


    have a look at the forum; somebody may have fixed it (if I remember correctly). Apparently the Siemens system needs some information that is not provided. But for me it is impossible to guess what that is. If would try to add a studyinstanceuid.


    Marcel

Participate now!

Don’t have an account yet? Register yourself now and be a part of our community!