Conquest problem with Philips Ingenia 3T

  • We are trying to set up Conquest as a data server for 3 Philips scanners. Most of the images are currently being generated by a Philips Ingenia 3T MRI scanner. These images were previously stored and viewed using KPACS without any problem. dcmtk seems to handle them as well (storescp, storescu, dcmconv). However, transfers to Conquest failed with the message:


    [CONQUESTSRV2] ***Implicit_Parse encountered an invalid element length during load of DCM file (in 68506c69)
    [CONQUESTSRV2] ***Length = 544436329


    Other images (from a GE scanner) transferred fine. Looking through the forum, I tried changing enabling NoDICOMCheck in the ini file, also commenting out MRImageStorageEnhanced and GrayscaleSoftcopyPresentationStateStorage. The first change just resulted in a silent fail. The latter changes did nothing.


    FYI, a series of scans from this scanner is often bundled into a single DICOM file. I don't know if this is common or not. Any suggestions are welcome.

  • Changing NoDICOMCheck = 1 got rid of the error message, but the store still fails. I can't decipher any error from the Conquest debug (see below)


    Brian


    [CONQUESTSRV2] UPACS THREAD 3: STARTED AT: Tue Oct 29 16:53:30 2013
    [CONQUESTSRV2] A-ASSOCIATE-RQ Packet Dump
    [CONQUESTSRV2] Calling Application Title : "KPServer52"
    [CONQUESTSRV2] Called Application Title : "CONQUESTSRV2"
    [CONQUESTSRV2] Application Context : "1.2.840.10008.3.1.1.1", PDU length: 16384
    [CONQUESTSRV2] Number of Proposed Presentation Contexts: 1
    [CONQUESTSRV2] Presentation Context 0 "1.2.840.10008.5.1.4.1.1.4.1" 1
    [CONQUESTSRV2] Server Command := 0001
    [CONQUESTSRV2] Message ID := 0003
    [CONQUESTSRV2] Failed STORAGE
    [CONQUESTSRV2] 0002,0010 17 UI TransferSyntaxUID "1.2.840.10008.1.2"
    [CONQUESTSRV2] 0008,0005 10 CS SpecificCharacterSet "ISO_IR 100"
    [CONQUESTSRV2] 0008,0008 24 CS ImageType "ORIGINAL\\PRIMARY\\T2\\NONE"
    [CONQUESTSRV2] 0008,0012 8 DA InstanceCreationDate "20111205"
    [CONQUESTSRV2] 0008,0013 6 TM InstanceCreationTime "124928"
    [CONQUESTSRV2] 0008,0014 24 UI InstanceCreatorUID "1.3.46.670589.11.42034.5"
    [CONQUESTSRV2] 0008,0016 28 UI SOPClassUID "1.2.840.10008.5.1.4.1.1.4.1"
    [CONQUESTSRV2] 0008,0018 56 UI SOPInstanceUID "1.3.46.670589.11.42034.5.20.1.1.4584.2011120511572859385"
    [CONQUESTSRV2] 0008,0020 8 DA StudyDate "20111205"
    [CONQUESTSRV2] 0008,0021 8 DA SeriesDate "20111205"
    [CONQUESTSRV2] 0008,0023 8 DA ContentDate "20111205"
    [CONQUESTSRV2] 0008,002a 20 DT AcquisitionDateTime "20111205115728.59000"
    [CONQUESTSRV2] 0008,0030 6 TM StudyTime "112906"
    [CONQUESTSRV2] 0008,0031 12 TM SeriesTime "115728.59000"
    [CONQUESTSRV2] 0008,0033 12 TM ContentTime "115728.59000"
    [CONQUESTSRV2] 0008,0050 0 SH AccessionNumber (empty)
    [CONQUESTSRV2] 0008,0060 2 CS Modality "MR"
    [CONQUESTSRV2] 0008,0070 18 LO Manufacturer "Philips Healthcare"
    [CONQUESTSRV2] 0008,0080 18 LO InstitutionName "Methodist Hospital"
    [CONQUESTSRV2] 0008,0090 0 PN ReferringPhysicianNa (empty)
    [CONQUESTSRV2] 0008,0100 0 SH CodeValue (empty)
    [CONQUESTSRV2] 0008,0102 4 SH CodingSchemeDesignat "DCM "
    [CONQUESTSRV2] 0008,0104 0 LO CodeMeaning (empty)
    [CONQUESTSRV2] 0008,1010 8 SH StationName "tmhri3t1"
    [CONQUESTSRV2] 0008,1030 0 LO StudyDescription (empty)
    [CONQUESTSRV2] 0008,103e 8 LO SeriesDescription "T2W_TSE "
    [CONQUESTSRV2] 0008,1040 8 LO InstitutionalDepartm "Research"
    [CONQUESTSRV2] 0008,1070 0 PN OperatorsName (empty)
    [CONQUESTSRV2] 0008,1080 0 LO AdmittingDiagnosesDe (empty)
    [CONQUESTSRV2] 0008,1090 8 LO ManufacturerModelNam "Ingenia "
    [CONQUESTSRV2] 0008,1111 0 SQ ReferencedPerformedP Sequence Item
    [CONQUESTSRV2] >0008,0012 8 DA InstanceCreationDate "20111205"
    [CONQUESTSRV2] >0008,0013 6 TM InstanceCreationTime "124924"
    [CONQUESTSRV2] >0008,0014 24 UI InstanceCreatorUID "1.3.46.670589.11.42034.5"
    [CONQUESTSRV2] >0008,1150 24 UI ReferencedSOPClassUI "1.2.840.10008.3.1.2.3.3"
    [CONQUESTSRV2] >0008,1155 52 UI ReferencedSOPInstanc "1.3.46.670589.11.42034.5.0.10800.2011120511290618001"
    [CONQUESTSRV2] >0020,0013 2 IS InstanceNumber "0 "
    [CONQUESTSRV2] >2005,0014 26 LO Unknown "Philips MR Imaging DD 005 "
    [CONQUESTSRV2] >2005,1406 2 US Unknown 0
    [CONQUESTSRV2] 0008,9092 0 SQ ReferencedImageEvide Sequence Item
    [CONQUESTSRV2] >0008,1115 0 SQ ReferencedSeriesSequ Sequence Item
    [CONQUESTSRV2] >>0008,1199 0 SQ ReferencedSOPSequenc Sequence Item
    [CONQUESTSRV2] >>>0008,1150 28 UI ReferencedSOPClassUI "1.2.840.10008.5.1.4.1.1.4.1"
    [CONQUESTSRV2] >>>0008,1155 56 UI ReferencedSOPInstanc "1.3.46.670589.11.42034.5.20.1.1.4584.2011120511483603320"
    [CONQUESTSRV2] >>0020,000e 52 UI SeriesInstanceUID "1.3.46.670589.11.42034.5.0.4584.2011120511483603320"
    [CONQUESTSRV2] >0020,000d 52 UI StudyInstanceUID "1.3.46.670589.11.42034.5.0.10800.2011120511290615000"
    [CONQUESTSRV2] -
    [CONQUESTSRV2]
    [CONQUESTSRV2] >0008,1115 0 SQ ReferencedSeriesSequ Sequence Item
    [CONQUESTSRV2] >>0008,1199 0 SQ ReferencedSOPSequenc Sequence Item
    [CONQUESTSRV2] >>>0008,1150 28 UI ReferencedSOPClassUI "1.2.840.10008.5.1.4.1.1.4.1"
    [CONQUESTSRV2] >>>0008,1155 56 UI ReferencedSOPInstanc "1.3.46.670589.11.42034.5.20.1.1.4584.2011120511483603320"
    [CONQUESTSRV2] >>0020,000e 52 UI SeriesInstanceUID "1.3.46.670589.11.42034.5.0.4584.2011120511483603320"
    [CONQUESTSRV2] >0020,000d 52 UI StudyInstanceUID "1.3.46.670589.11.42034.5.0.10800.2011120511290615000"
    [CONQUESTSRV2] -
    [CONQUESTSRV2]
    [CONQUESTSRV2] >0008,1115 0 SQ ReferencedSeriesSequ Sequence Item
    [CONQUESTSRV2] >>0008,1199 0 SQ ReferencedSOPSequenc Sequence Item
    [CONQUESTSRV2] >>>0008,1150 28 UI ReferencedSOPClassUI "1.2.840.10008.5.1.4.1.1.4.1"
    [CONQUESTSRV2] >>>0008,1155 56 UI ReferencedSOPInstanc "1.3.46.670589.11.42034.5.20.1.1.4584.2011120511483603320"
    [CONQUESTSRV2] >>0020,000e 52 UI SeriesInstanceUID "1.3.46.670589.11.42034.5.0.4584.2011120511483603320"
    [CONQUESTSRV2] >0020,000d 52 UI StudyInstanceUID "1.3.46.670589.11.42034.5.0.10800.2011120511290615000"
    [CONQUESTSRV2] 0008,9123 16 UI CreatorVersionUID "1.3.46.670589.11"
    [CONQUESTSRV2] 0008,9205 10 CS PixelPresentation "MONOCHROME"
    [CONQUESTSRV2] 0008,9206 6 CS VolumetricProperties "VOLUME"
    [CONQUESTSRV2] 0008,9207 4 CS VolumeBasedCalculati "NONE"
    [CONQUESTSRV2] 0008,9208 10 CS ComplexImageComponen "MAGNITUDE "
    [CONQUESTSRV2] 0008,9209 2 CS AcquisitionContrast "T2"
    [CONQUESTSRV2] 0010,0010 16 PN PatientName "bunnicula22L-02 "
    [CONQUESTSRV2] 0010,0020 6 LO PatientID "B22L-2"
    [CONQUESTSRV2] 0010,0030 8 DA PatientBirthDate "19700101"
    [CONQUESTSRV2] 0010,0040 2 CS PatientSex "M "
    [CONQUESTSRV2] 0010,1000 0 LO OtherPatientIDs (empty)
    [CONQUESTSRV2] 0010,1030 4 DS PatientWeight "100 "
    [CONQUESTSRV2] 0010,21c0 2 US PregnancyStatus 4
    [CONQUESTSRV2] 0018,0015 6 CS BodyPartExamined "BRAIN "
    [CONQUESTSRV2] 0018,0023 2 CS MRAcquisitionType "2D"
    [CONQUESTSRV2] 0018,0087 2 DS MagneticFieldStrengt "3 "
    [CONQUESTSRV2] 0018,0088 2 DS SpacingBetweenSlices "2 "
    [CONQUESTSRV2] 0018,0095 4 DS PixelBandwidth "219 "
    [CONQUESTSRV2] 0018,1000 6 LO DeviceSerialNumber "42034 "
    [CONQUESTSRV2] 0018,1020 14 LO SoftwareVersions "4.1.1\\4.1.1.1 "
    [CONQUESTSRV2] 0018,1030 14 LO ProtocolName "T2W_TSE SENSE "
    [CONQUESTSRV2] 0018,5100 4 CS PatientPosition "HFS "
    [CONQUESTSRV2] 0018,9004 8 CS ContentQualification "RESEARCH"
    [CONQUESTSRV2] 0018,9005 4 SH PulseSequenceName "TSE "
    [CONQUESTSRV2] 0018,9008 4 CS EchoPulseSequence "SPIN"
    [CONQUESTSRV2] 0018,9011 4 CS MultipleSpinEcho "YES "
    [CONQUESTSRV2] 0018,9012 2 CS MultiPlanarExcitatio "NO"
    [CONQUESTSRV2] 0018,9014 2 CS PhaseContrast "NO"
    [CONQUESTSRV2] 0018,9015 2 CS TimeOfFlightContrast "NO"
    [CONQUESTSRV2] 0018,9016 4 CS Spoiling "NONE"
    [CONQUESTSRV2] 0018,9017 4 CS SteadyStatePulseSequ "NONE"
    [CONQUESTSRV2] 0018,9018 2 CS EchoPlanarPulseSeque "NO"
    [CONQUESTSRV2] 0018,9020 4 CS MagnetizationTransfe "NONE"
    [CONQUESTSRV2] 0018,9021 2 CS T2Preparation "NO"
    [CONQUESTSRV2] 0018,9022 2 CS BloodSignalNulling "NO"
    [CONQUESTSRV2] 0018,9024 2 CS SaturationRecovery "NO"
    [CONQUESTSRV2] 0018,9025 4 CS SpectrallySelectedSu "NONE"
    [CONQUESTSRV2] 0018,9026 6 CS SpectrallySelectedEx "WATER "
    [CONQUESTSRV2] 0018,9027 4 CS SpatialPresaturation "NONE"
    [CONQUESTSRV2] 0018,9028 4 CS Tagging "NONE"
    [CONQUESTSRV2] 0018,9029 2 CS OversamplingPhase "2D"
    [CONQUESTSRV2] 0018,9032 12 CS GeometryOfKSpaceTrav "RECTILINEAR "
    [CONQUESTSRV2] 0018,9033 8 CS SegmentedKSpaceTrave "PARTIAL "
    [CONQUESTSRV2] 0018,9034 8 CS RectilinearPhaseEnco "UNKNOWN "
    [CONQUESTSRV2] 0018,9035 8 FD TagThickness
    [CONQUESTSRV2] 0018,9036 10 CS PartialFourierDirect "FREQUENCY "
    [CONQUESTSRV2] 0018,9037 4 CS CardiacSynchronizati "NONE"
    [CONQUESTSRV2] 0018,9051 8 CS TransmitCoilType "SURFACE "
    [CONQUESTSRV2] 0018,9053 8 FD ChemicalShiftReferen
    [CONQUESTSRV2] 0018,9058 2 US MRAcquisitionFrequen 428
    [CONQUESTSRV2] 0018,9059 2 CS Decoupling "NO"
    [CONQUESTSRV2] 0018,9064 6 CS KSpaceFiltering "RIESZ "
    [CONQUESTSRV2] >0020,9165 4 AT DimensionIndexPointe (0020, 9056)
    [CONQUESTSRV2] UPACS THREAD 3: ENDED AT: Tue Oct 29 16:53:30 2013

  • With GUI debuglevel:4, I get a little more data, but it doesn't seem any more helpful. Are there some other logs to look at? I also don't mind sending some data but the files are large.


    Brian




    [CONQUESTSRV2] UPACS THREAD 16: STARTED AT: Tue Oct 29 18:03:57 2013
    [CONQUESTSRV2] A-ASSOCIATE-RQ Packet Dump
    [CONQUESTSRV2] Calling Application Title : "KPServer52"
    [CONQUESTSRV2] Called Application Title : "CONQUESTSRV2"
    [CONQUESTSRV2] Application Context : "1.2.840.10008.3.1.1.1", PDU length: 16384
    [CONQUESTSRV2] Number of Proposed Presentation Contexts: 1
    [CONQUESTSRV2] Presentation Context 0 "1.2.840.10008.5.1.4.1.1.4.1" 1
    [CONQUESTSRV2] Server Command := 0001
    [CONQUESTSRV2] Message ID := 0003
    [CONQUESTSRV2] 0000,0002 28 UI AffectedSOPClassUID "1.2.840.10008.5.1.4.1.1.4.1"
    [CONQUESTSRV2] 0000,0100 2 US CommandField 1
    [CONQUESTSRV2] 0000,0110 2 US MessageID 3
    [CONQUESTSRV2] 0000,0700 2 US Priority 0
    [CONQUESTSRV2] 0000,0800 2 US DataSetType 258
    [CONQUESTSRV2] 0000,1000 56 UI AffectedSOPInstanceU "1.3.46.670589.11.42034.5.20.1.1.4584.2011120511572859385"
    [CONQUESTSRV2] 0002,0010 17 UI TransferSyntaxUID "1.2.840.10008.1.2"
    [CONQUESTSRV2] Failed STORAGE
    [CONQUESTSRV2] 0002,0010 17 UI TransferSyntaxUID "1.2.840.10008.1.2"
    [CONQUESTSRV2] 0008,0005 10 CS SpecificCharacterSet "ISO_IR 100"
    [CONQUESTSRV2] 0008,0008 24 CS ImageType "ORIGINAL\\PRIMARY\\T2\\NONE"
    [CONQUESTSRV2] 0008,0012 8 DA InstanceCreationDate "20111205"
    [CONQUESTSRV2] 0008,0013 6 TM InstanceCreationTime "124928"
    [CONQUESTSRV2] 0008,0014 24 UI InstanceCreatorUID "1.3.46.670589.11.42034.5"
    [CONQUESTSRV2] 0008,0016 28 UI SOPClassUID "1.2.840.10008.5.1.4.1.1.4.1"
    [CONQUESTSRV2] 0008,0018 56 UI SOPInstanceUID "1.3.46.670589.11.42034.5.20.1.1.4584.2011120511572859385"
    [CONQUESTSRV2] 0008,0020 8 DA StudyDate "20111205"
    [CONQUESTSRV2] 0008,0021 8 DA SeriesDate "20111205"
    [CONQUESTSRV2] 0008,0023 8 DA ContentDate "20111205"
    [CONQUESTSRV2] 0008,002a 20 DT AcquisitionDateTime "20111205115728.59000"
    [CONQUESTSRV2] 0008,0030 6 TM StudyTime "112906"
    [CONQUESTSRV2] 0008,0031 12 TM SeriesTime "115728.59000"
    [CONQUESTSRV2] 0008,0033 12 TM ContentTime "115728.59000"
    [CONQUESTSRV2] 0008,0050 0 SH AccessionNumber (empty)
    [CONQUESTSRV2] 0008,0060 2 CS Modality "MR"
    [CONQUESTSRV2] 0008,0070 18 LO Manufacturer "Philips Healthcare"
    [CONQUESTSRV2] 0008,0080 18 LO InstitutionName "Methodist Hospital"
    [CONQUESTSRV2] 0008,0090 0 PN ReferringPhysicianNa (empty)
    [CONQUESTSRV2] 0008,0100 0 SH CodeValue (empty)
    [CONQUESTSRV2] 0008,0102 4 SH CodingSchemeDesignat "DCM "
    [CONQUESTSRV2] 0008,0104 0 LO CodeMeaning (empty)
    [CONQUESTSRV2] 0008,1010 8 SH StationName "tmhri3t1"
    [CONQUESTSRV2] 0008,1030 0 LO StudyDescription (empty)
    [CONQUESTSRV2] 0008,103e 8 LO SeriesDescription "T2W_TSE "
    [CONQUESTSRV2] 0008,1040 8 LO InstitutionalDepartm "Research"
    [CONQUESTSRV2] 0008,1070 0 PN OperatorsName (empty)
    [CONQUESTSRV2] 0008,1080 0 LO AdmittingDiagnosesDe (empty)
    [CONQUESTSRV2] 0008,1090 8 LO ManufacturerModelNam "Ingenia "
    [CONQUESTSRV2] 0008,1111 0 SQ ReferencedPerformedP Sequence Item
    [CONQUESTSRV2] >0008,0012 8 DA InstanceCreationDate "20111205"
    [CONQUESTSRV2] >0008,0013 6 TM InstanceCreationTime "124924"
    [CONQUESTSRV2] >0008,0014 24 UI InstanceCreatorUID "1.3.46.670589.11.42034.5"
    [CONQUESTSRV2] >0008,1150 24 UI ReferencedSOPClassUI "1.2.840.10008.3.1.2.3.3"
    [CONQUESTSRV2] >0008,1155 52 UI ReferencedSOPInstanc "1.3.46.670589.11.42034.5.0.10800.2011120511290618001"
    [CONQUESTSRV2] >0020,0013 2 IS InstanceNumber "0 "
    [CONQUESTSRV2] >2005,0014 26 LO Unknown "Philips MR Imaging DD 005 "
    [CONQUESTSRV2] >2005,1406 2 US Unknown 0
    [CONQUESTSRV2] 0008,9092 0 SQ ReferencedImageEvide Sequence Item
    [CONQUESTSRV2] >0008,1115 0 SQ ReferencedSeriesSequ Sequence Item
    [CONQUESTSRV2] >>0008,1199 0 SQ ReferencedSOPSequenc Sequence Item
    [CONQUESTSRV2] >>>0008,1150 28 UI ReferencedSOPClassUI "1.2.840.10008.5.1.4.1.1.4.1"
    [CONQUESTSRV2] >>>0008,1155 56 UI ReferencedSOPInstanc "1.3.46.670589.11.42034.5.20.1.1.4584.2011120511483603320"
    [CONQUESTSRV2] >>0020,000e 52 UI SeriesInstanceUID "1.3.46.670589.11.42034.5.0.4584.2011120511483603320"
    [CONQUESTSRV2] >0020,000d 52 UI StudyInstanceUID "1.3.46.670589.11.42034.5.0.10800.2011120511290615000"
    [CONQUESTSRV2] -
    [CONQUESTSRV2]
    [CONQUESTSRV2] >0008,1115 0 SQ ReferencedSeriesSequ Sequence Item
    [CONQUESTSRV2] >>0008,1199 0 SQ ReferencedSOPSequenc Sequence Item
    [CONQUESTSRV2] >>>0008,1150 28 UI ReferencedSOPClassUI "1.2.840.10008.5.1.4.1.1.4.1"
    [CONQUESTSRV2] >>>0008,1155 56 UI ReferencedSOPInstanc "1.3.46.670589.11.42034.5.20.1.1.4584.2011120511483603320"
    [CONQUESTSRV2] >>0020,000e 52 UI SeriesInstanceUID "1.3.46.670589.11.42034.5.0.4584.2011120511483603320"
    [CONQUESTSRV2] >0020,000d 52 UI StudyInstanceUID "1.3.46.670589.11.42034.5.0.10800.2011120511290615000"
    [CONQUESTSRV2] -
    [CONQUESTSRV2]
    [CONQUESTSRV2] >0008,1115 0 SQ ReferencedSeriesSequ Sequence Item
    [CONQUESTSRV2] >>0008,1199 0 SQ ReferencedSOPSequenc Sequence Item
    [CONQUESTSRV2] >>>0008,1150 28 UI ReferencedSOPClassUI "1.2.840.10008.5.1.4.1.1.4.1"
    [CONQUESTSRV2] >>>0008,1155 56 UI ReferencedSOPInstanc "1.3.46.670589.11.42034.5.20.1.1.4584.2011120511483603320"
    [CONQUESTSRV2] >>0020,000e 52 UI SeriesInstanceUID "1.3.46.670589.11.42034.5.0.4584.2011120511483603320"
    [CONQUESTSRV2] >0020,000d 52 UI StudyInstanceUID "1.3.46.670589.11.42034.5.0.10800.2011120511290615000"
    [CONQUESTSRV2] 0008,9123 16 UI CreatorVersionUID "1.3.46.670589.11"
    [CONQUESTSRV2] 0008,9205 10 CS PixelPresentation "MONOCHROME"
    [CONQUESTSRV2] 0008,9206 6 CS VolumetricProperties "VOLUME"
    [CONQUESTSRV2] 0008,9207 4 CS VolumeBasedCalculati "NONE"
    [CONQUESTSRV2] 0008,9208 10 CS ComplexImageComponen "MAGNITUDE "
    [CONQUESTSRV2] 0008,9209 2 CS AcquisitionContrast "T2"
    [CONQUESTSRV2] 0010,0010 16 PN PatientName "bunnicula22L-02 "
    [CONQUESTSRV2] 0010,0020 6 LO PatientID "B22L-2"
    [CONQUESTSRV2] 0010,0030 8 DA PatientBirthDate "19700101"
    [CONQUESTSRV2] 0010,0040 2 CS PatientSex "M "
    [CONQUESTSRV2] 0010,1000 0 LO OtherPatientIDs (empty)
    [CONQUESTSRV2] 0010,1030 4 DS PatientWeight "100 "
    [CONQUESTSRV2] 0010,21c0 2 US PregnancyStatus 4
    [CONQUESTSRV2] 0018,0015 6 CS BodyPartExamined "BRAIN "
    [CONQUESTSRV2] 0018,0023 2 CS MRAcquisitionType "2D"
    [CONQUESTSRV2] 0018,0087 2 DS MagneticFieldStrengt "3 "
    [CONQUESTSRV2] 0018,0088 2 DS SpacingBetweenSlices "2 "
    [CONQUESTSRV2] 0018,0095 4 DS PixelBandwidth "219 "
    [CONQUESTSRV2] 0018,1000 6 LO DeviceSerialNumber "42034 "
    [CONQUESTSRV2] 0018,1020 14 LO SoftwareVersions "4.1.1\\4.1.1.1 "
    [CONQUESTSRV2] 0018,1030 14 LO ProtocolName "T2W_TSE SENSE "
    [CONQUESTSRV2] 0018,5100 4 CS PatientPosition "HFS "
    [CONQUESTSRV2] 0018,9004 8 CS ContentQualification "RESEARCH"
    [CONQUESTSRV2] 0018,9005 4 SH PulseSequenceName "TSE "
    [CONQUESTSRV2] 0018,9008 4 CS EchoPulseSequence "SPIN"
    [CONQUESTSRV2] 0018,9011 4 CS MultipleSpinEcho "YES "
    [CONQUESTSRV2] 0018,9012 2 CS MultiPlanarExcitatio "NO"
    [CONQUESTSRV2] 0018,9014 2 CS PhaseContrast "NO"
    [CONQUESTSRV2] 0018,9015 2 CS TimeOfFlightContrast "NO"
    [CONQUESTSRV2] 0018,9016 4 CS Spoiling "NONE"
    [CONQUESTSRV2] 0018,9017 4 CS SteadyStatePulseSequ "NONE"
    [CONQUESTSRV2] 0018,9018 2 CS EchoPlanarPulseSeque "NO"
    [CONQUESTSRV2] 0018,9020 4 CS MagnetizationTransfe "NONE"
    [CONQUESTSRV2] 0018,9021 2 CS T2Preparation "NO"
    [CONQUESTSRV2] 0018,9022 2 CS BloodSignalNulling "NO"
    [CONQUESTSRV2] 0018,9024 2 CS SaturationRecovery "NO"
    [CONQUESTSRV2] 0018,9025 4 CS SpectrallySelectedSu "NONE"
    [CONQUESTSRV2] 0018,9026 6 CS SpectrallySelectedEx "WATER "
    [CONQUESTSRV2] 0018,9027 4 CS SpatialPresaturation "NONE"
    [CONQUESTSRV2] 0018,9028 4 CS Tagging "NONE"
    [CONQUESTSRV2] 0018,9029 2 CS OversamplingPhase "2D"
    [CONQUESTSRV2] 0018,9032 12 CS GeometryOfKSpaceTrav "RECTILINEAR "
    [CONQUESTSRV2] 0018,9033 8 CS SegmentedKSpaceTrave "PARTIAL "
    [CONQUESTSRV2] 0018,9034 8 CS RectilinearPhaseEnco "UNKNOWN "
    [CONQUESTSRV2] 0018,9035 8 FD TagThickness
    [CONQUESTSRV2] 0018,9036 10 CS PartialFourierDirect "FREQUENCY "
    [CONQUESTSRV2] 0018,9037 4 CS CardiacSynchronizati "NONE"
    [CONQUESTSRV2] 0018,9051 8 CS TransmitCoilType "SURFACE "
    [CONQUESTSRV2] 0018,9053 8 FD ChemicalShiftReferen
    [CONQUESTSRV2] 0018,9058 2 US MRAcquisitionFrequen 428
    [CONQUESTSRV2] 0018,9059 2 CS Decoupling "NO"
    [CONQUESTSRV2] 0018,9064 6 CS KSpaceFiltering "RIESZ "
    [CONQUESTSRV2] 2001,105f 0 SQ StackSequence Sequence Item

  • That is basically the entire thing. Setting debuglevel:4 in the GUI gave a few extra lines at the top:


    ...
    [CONQUESTSRV2] Server Command := 0001
    [CONQUESTSRV2] Message ID := 0003
    [CONQUESTSRV2] 0000,0002 28 UI AffectedSOPClassUID "1.2.840.10008.5.1.4.1.1.4.1"
    [CONQUESTSRV2] 0000,0100 2 US CommandField 1
    [CONQUESTSRV2] 0000,0110 2 US MessageID 3
    [CONQUESTSRV2] 0000,0700 2 US Priority 0
    [CONQUESTSRV2] 0000,0800 2 US DataSetType 258
    [CONQUESTSRV2] 0000,1000 56 UI AffectedSOPInstanceU "1.3.46.670589.11.42034.5.20.1.1.4584.2011120511572859385"
    [CONQUESTSRV2] 0002,0010 17 UI TransferSyntaxUID "1.2.840.10008.1.2"
    [CONQUESTSRV2] Failed STORAGE
    [CONQUESTSRV2] 0002,0010 17 UI TransferSyntaxUID "1.2.840.10008.1.2"
    ...
    etc.

  • Hi,


    as you mailed the image is 'corrupted' by kpacs; but the file you sent loads OK with drag and drop. The problem may be because it contains many private sequences. The error states a tag that is not present, I.e., the DICOM parser loses synchronization. It may be that the culprit is an implicit transfer. You might try and edit dgatesop.lst to disable LittleEndianImplicit and see what happens.


    Marcel

  • Okay. Based on your hint, I looked for where to do this and noticed that my dgatesop.lst had:


    LittleEndianImplicit 1.2.840.10008.1.2 transfer
    #LittleEndianExplicit 1.2.840.10008.1.2.1 transfer


    rather than what was in the manual:


    LittleEndianImplicit 1.2.840.10008.1.2 transfer
    LittleEndianExplicit 1.2.840.10008.1.2.1 transfer


    Removing the # seems to have solved the problem. Don't know how it got there in the first place.... but thanks for your help!


    Brian

  • Hi,
    I think this post might be related to my current question. We are transferring multi-frame Dicom from a Philips Achieva (R.3.2.1) MR scanner to Conquest running on a Windows 7 PC. I am finding that the Private attributes are written as bytes and seem to loose their VR. This VR issue does not happen if a file is drag-and-dropped into Conquest (at least not on my laptop version). My dgatesop.lst has the entries:
    LittleEndianImplicit 1.2.840.10008.1.2 transfer
    #LittleEndianExplicit 1.2.840.10008.1.2.1 transfer
    #BigEndianExplicit 1.2.840.10008.1.2.2 transfer


    which looks similar to what was seen in this thread and different from the manual. Would you recommend removing the "#" before the LittleEndianExplicit?


    Regards,


    David

  • Hi Marcel,


    I am having the same issue here. When I try to send a study to Conquest 1.4.17 from a Philips Achieva 3T I have this message:
    [CompressJPEGImage]: JPEG compression not allowed from PR/SR/RT


    And then the server do not accept all images, just the first image of the series.


    In other server I have this message, after some configuration by the engineer trying to solve the problem:
    ***no UID associated with PCID 159 during load of DCM file in ParseRawVRIntoDCM


    After it I have this message in the PacsTrouble.log:
    Inconsistent StudyModal in DICOMStudies:......., Old='MR', New='PR//MR'


    So I tried al I found to solve this problem like:
    Removing the commentary of LittleEndianExplicit on dgatesop.lst;
    Changing the outgoig compression to 'ul';
    Commented the lines MRIStorageEnhanced and GrayscaleSoftcopyPresentationStateStorage too;


    Nothing at all worked. I don't know what can I do more, based on what I have ever found in the forum.


    One thing is important, I could send the studies from an eFilm nicely, but when trying to send from Philips Achieva 3T or Osirix I faced this problem.
    I will grateful if you can help me!


    Thank you!

  • Quote from marcelvanherk

    Hi,


    can you try to turn off jpeg compression and see what happens?


    Marcel


    Hi Marcel,


    I already tried it, but nothing happens. It still accepting just few images, because of database inconsistency. Like in PacsTrouble.log:
    Inconsistent StudyModal in DICOMStudy: PatientID = '123456' StudyInsta = 'xxxx', old= 'MR`, New= 'PR//MR'


    I saw this case too: Inconsistent StudyModal in DICOMStudy: PatientID = '123456' StudyInsta = 'xxxx', old= 'MR`, New= 'SR//MR'


    Any idea?


    Thanks,

  • So why it stop sendding the images inside a series?
    Just the first image of each series is accepted, the others no.


    Acctualy, Osirix seems to send just 17 images from a study whith more than 100.


    Other viewers or equipements send all images from other modalities.


    Do you have any idea to help me? Or something that I can do to help with this problem?


    I am very new inthis world!


    Thank you again Marcel

  • Quote from marcelvanherk

    Hi,


    Are the image types for the other images maybe not in dgatesop.lst? Is there any error message?


    Marcel


    Hi Marcel, where can I find if there is any error message?
    In the PacsTrouble.log there is this error message: ***no UID associated with PCID 159 during load of DCM file in ParseRawVRIntoDCM


    Thank you again!

  • Quote from robertvanommen

    Philips Private MR Spectrum Storage 1.3.46.670589.11.0.0.12.1
    Philips Private MR Series Data Storage 1.3.46.670589.11.0.0.12.2
    Philips Private MR Examcard Storage 1.3.46.670589.11.0.0.12.4


    Philips often forgets to unconfigure the above private SOP Classes for other than Philips PACS systems. Add the above to the dgatesop.lst


    Regards,
    Robert


    Hi Robert,


    I added this line into my dgatesop.lst and I will test it again and write the results here latter.
    Can you, please, send me your dgatesol.lst for I check with mine?


    Thank you,
    Thiago

  • Hello Marcel,


    After I include those SOP Classe sent by Robert, my Conquest start accepting the studies from Osirix, but there is something strange happening that I did not understand yet. On my Osirix, some study has 120 images, e.g, when I try to send it to Conquest, Osirix tells me that just 34 archives will be send and it sends all 34 files. On my Conquest Web (Wado interface) I can see just 1 image from each serie on the study and sometimes I can't view these images on serversideviewer, but other studies it is normal.


    When I try to get this exam from Conquest to Osirix (after deleting the exam from Osirix) everything runs ok and Osirix gets the same 34 archives and shows me 120 images again. Do you know why it happens?


    Sometimes, even Osirix telling me that he have to send 34 archives to Conquest, Conquest only accept certain files (for example 26 files). Is there a way to discover what type of file Conquest is missing? Or SOP class?
    Then copying back the study to Osirix is normally, even though downloading just the 26 files.


    Best regards,
    Thiago

  • Quote from robertvanommen

    Just add:


    PhilipsPrivateMRSpectrumStorage 1.3.46.670589.11.0.0.12.1 sop
    PhilipsPrivateMRSerieDataStorage 1.3.46.670589.11.0.0.12.2 sop
    PhilipsPrivateMRExamcardStorage 1.3.46.670589.11.0.0.12.4 sop


    between the other "storage" sopclasses


    Thank you Robert, I added these SOP Classes an the RawStorage too, and it worked!

Participate now!