Andreas:
I have a problem with K-PACS that does NOT occur with iQ-VIEW.
I am using the lastest version, but i also tried with an earlier version.
I have 4 sites using K-PACS as a way to transfer studies from their location to my iQ-WEB PACS.
If there is data in the "referring physican" field, K-PACS reports a "storage error", but there is no error. The studies store just fine, with no error on the iQ-WEB side.
if there is NO data in the "referring physican" field, no error is reported.
I did the exact same study, same system, with iQ-VIEW, with and without data in the "referring physican" field, no storage error reported and everything worked fine.
We use the field to determine viewing rights on the iQ-WEB, so we have to use it. The techs now have to log into the iQ-WEB to verify that the study went ok since the K-PACS reports the error, even though they are aware of the issue. Can't take a chance on it now storing properly.
Below is the log (since I could not attach the it as this site will not allow .txt files to attach).
Can you find out the bug for me please. These people bought 3 iQ-WEB license's from me via John Mazur in Philadelphia and neither he nor I can see what is causing this issue.
Thanks.
[Stop Process] >> : writing configurations to C:\KPacs\k-pacs.INI
[Stop Process] >> : writing configurations to C:\KPacs\k-pacs.INI
*******************************************************
8/27/2009 9:17:45 AM ThreadID[972] : Sending images to TIAPACS started
*******************************************************
[972] >> : c-store request for Series C:\KPacs\Imagebox\1.2.826.0.1.3680043.1.2.119.707110073.20090826.182541.0\1.2.826.0.1.3680043.1.2.119.707110073.20090826.182541.0.0.1 initiated.
[DICOM SCU] >> : Connected to: 10.0.15.100:105
[DICOM SCU] >> : Item Type: 2
[DICOM SCU] >> : Association Accept::ReadDynamic
[DICOM SCU] >> : Presentation Context Accept, Transfer Syntax: 1.2.840.10008.1.2.4.70
[DICOM SCU] >> : WNY1C-DVS >> TIAPACS
[DICOM SCU] >> : (0000,0000), 4, UL, "142" - Group Length
[DICOM SCU] >> : (0000,0002), 26, UI, "1.2.840.10008.5.1.4.1.1.1" - Affected SOP Class UID
[DICOM SCU] >> : (0000,0100), 2, US, "1" - Command Field
[DICOM SCU] >> : (0000,0110), 2, US, "3" - Message ID
[DICOM SCU] >> : (0000,0700), 2, US, "0" - Priority
[DICOM SCU] >> : (0000,0800), 2, US, "258" - Data Set Type
[DICOM SCU] >> : (0000,1000), 60, UI, "1.2.826.0.1.3680043.1.2.119.707110073.20090826.182541.0.0.1" - Affected SOP Instance UID
[DICOM SCU] >> :
[DICOM SCU] >> : WNY1C-DVS >> TIAPACS
[DICOM SCU] >> : (0008,0016), 26, UI, "1.2.840.10008.5.1.4.1.1.1" - SOP Class UID
[DICOM SCU] >> : (0008,0018), 60, UI, "1.2.826.0.1.3680043.1.2.119.707110073.20090826.182541.0.0.1" - SOP Instance UID
[DICOM SCU] >> : (0008,0020), 8, DA, "20090826" - Study Date
[DICOM SCU] >> : (0008,0022), 8, DA, "20090826" - Acquisition Date
[DICOM SCU] >> : (0008,0030), 6, TM, "182541" - Study Time
[DICOM SCU] >> : (0008,0032), 6, TM, "182649" - Acquisition Time
[DICOM SCU] >> : (0008,0050), 0, SH, "" - Accession Number
[DICOM SCU] >> : (0008,0060), 2, CS, "CR" - Modality
[DICOM SCU] >> : (0008,0070), 30, LO, "Imaging Dynamics Company Ltd. " - Manufacturer
[DICOM SCU] >> : (0008,0080), 48, LO, "Western New York Immediate Care (NIAGARA FALLS) " - Institution Name
[DICOM SCU] >> : (0008,0090), 6, PN, "WNYIC " - Referring Physician’s Name
[DICOM SCU] >> : (0008,1010), 10, SH, "WNYIC-XRAY" - Station Name
[DICOM SCU] >> : (0008,1030), 8, LO, "TOE (S) " - Study Description
[DICOM SCU] >> : (0008,1070), 6, PN, "^KN^^^" - Operators’ Name
[DICOM SCU] >> : (0008,1090), 2, LO, "X3" - Manufacturer’s Model Name
[DICOM SCU] >> : (0008,2112), -1, SQ
[DICOM SCU] >> : (FFFE,E000), -1, DL
[DICOM SCU] >> : (0008,1150), 26, UI, "1.2.840.10008.5.1.4.1.1.1" - Referenced SOP Class UID
[DICOM SCU] >> : (0008,1155), 60, UI, "1.2.826.0.1.3680043.1.2.119.707110073.20090826.182541.0.0.0" - Referenced SOP Instance UID
[DICOM SCU] >> : (FFFE,E00D), 0, DL, "" - Item Delimitation Item
[DICOM SCU] >> : (FFFE,E0DD), 0, DL, "" - Sequence Delimitation Item
[DICOM SCU] >> : (0010,0010), 18, PN, "ZAMBOTTI^JENNA^^^ " - Patient’s Name
[DICOM SCU] >> : (0010,0020), 8, LO, "W1178990" - Patient ID
[DICOM SCU] >> : (0010,0030), 8, DA, "19681222" - Patient's Birth Date
[DICOM SCU] >> : (0010,0040), 2, CS, "F " - Patient's Sex
[DICOM SCU] >> : (0010,2000), 0, LO, "" - Medical Alerts
[DICOM SCU] >> : (0010,2110), 0, LO, "" - Contrast Allergies
[DICOM SCU] >> : (0018,0015), 4, CS, "TOE " - Body Part Examined
[DICOM SCU] >> : (0018,0060), 2, DS, "60" - KVP
[DICOM SCU] >> : (0018,1000), 12, LO, "D7E07110073 " - Device Serial Number
[DICOM SCU] >> : (0018,1020), 8, LO, "3.3.257 " - Software Version(s)
[DICOM SCU] >> : (0018,1030), 8, LO, "TOE (S) " - Protocol Name
[DICOM SCU] >> : (0018,1150), 2, IS, "20" - Exposure Time
[DICOM SCU] >> : (0018,1151), 4, IS, "100 " - X-ray Tube Current
[DICOM SCU] >> : (0018,1152), 2, IS, "2 " - Exposure
[DICOM SCU] >> : (0018,1153), 4, IS, "2000" - Exposure in uAs
[DICOM SCU] >> : (0018,1164), 12, DS, "0.144\0.144 " - Imager Pixel Spacing
[DICOM SCU] >> : (0018,1405), 4, IS, "0.6 " - Relative X-ray Exposure
[DICOM SCU] >> : (0018,5101), 16, CS, "ANTERO_POSTERIOR" - View Position
[DICOM SCU] >> : (0018,6000), 4, DS, "300 " - Sensitivity
[DICOM SCU] >> : (0018,8150), 6, DS, "20000 " - Exposure Time in µS
[DICOM SCU] >> : (0018,8151), 6, DS, "100000" - X-Ray Tube Current in µA
[DICOM SCU] >> : (0020,000D), 56, UI, "1.2.826.0.1.3680043.1.2.119.707110073.20090826.182541.0" - Study Instance UID
[DICOM SCU] >> : (0020,000E), 60, UI, "1.2.826.0.1.3680043.1.2.119.707110073.20090826.182541.0.0.1" - Series Instance UID
[DICOM SCU] >> : (0020,0010), 4, SH, "8948" - Study ID
[DICOM SCU] >> : (0020,0011), 2, IS, "1 " - Series Number
[DICOM SCU] >> : (0020,0013), 6, IS, "54204 " - Instance Number
[DICOM SCU] >> : (0028,0002), 2, US, "1" - Samples per Pixel
[DICOM SCU] >> : (0028,0004), 12, CS, "MONOCHROME2 " - Photometric Interpretation
[DICOM SCU] >> : (0028,0010), 2, US, "1208" - Rows
[DICOM SCU] >> : (0028,0011), 2, US, "752" - Columns
[DICOM SCU] >> : (0028,0030), 12, DS, "0.144\0.144 " - Pixel Spacing
[DICOM SCU] >> : (0028,0100), 2, US, "16" - Bits Allocated
[DICOM SCU] >> : (0028,0101), 2, US, "12" - Bits Stored
[DICOM SCU] >> : (0028,0102), 2, US, "11" - High Bit
[DICOM SCU] >> : (0028,0103), 2, US, "0" - Pixel Representation
[DICOM SCU] >> : (0028,1050), 4, DS, "1689" - Window Center
[DICOM SCU] >> : (0028,1051), 4, DS, "1893" - Window Width
[DICOM SCU] >> : (0028,1052), 2, DS, "0 " - Rescale Intercept
[DICOM SCU] >> : (0028,1053), 2, DS, "1 " - Rescale Slope
[DICOM SCU] >> : (0028,1054), 2, LO, "US" - Rescale Type
[DICOM SCU] >> : (0040,0254), 8, LO, "TOE (S) " - Performed Procedure Step Description
[DICOM SCU] >> : (0040,0275), -1, SQ
[DICOM SCU] >> : (FFFE,E000), -1, DL
[DICOM SCU] >> : (0008,0050), 0, SH, "" - Accession Number
[DICOM SCU] >> : (0020,000D), 56, UI, "1.2.826.0.1.3680043.1.2.119.707110073.20090826.182541.0" - Study Instance UID
[DICOM SCU] >> : (FFFE,E00D), 0, DL, "" - Item Delimitation Item
[DICOM SCU] >> : (FFFE,E0DD), 0, DL, "" - Sequence Delimitation Item
[DICOM SCU] >> : (4321,0010), 30, UN, "Imaging Dynamics Company Ltd. " - *** Unidentified ***
[DICOM SCU] >> : (4321,1041), 2, UN, "N " - *** Unidentified ***
[DICOM SCU] >> : (4321,1042), 2, UN, "´" - *** Unidentified ***
[DICOM SCU] >> : (4321,1064), 2, UN, "HR" - *** Unidentified ***
[DICOM SCU] >> : (7FE0,0010), -1, OW
[DICOM SCU] >> : (FFFE,E000), 0, DL, "" - Item
[DICOM SCU] >> : (FFFE,E000), 770348, DL
[DICOM SCU] >> : (FFFE,E0DD), 0, DL, "" - Sequence Delimitation Item
[DICOM SCU] >> :
[DICOM SCU] >> : WNY1C-DVS << TIAPACS
[DICOM SCU] >> : (0000,0000), 4, UL, "142" - Group Length
[DICOM SCU] >> : (0000,0002), 26, UI, "1.2.840.10008.5.1.4.1.1.1" - Affected SOP Class UID
[DICOM SCU] >> : (0000,0100), 2, US, "32769" - Command Field
[DICOM SCU] >> : (0000,0120), 2, US, "3" - Message ID Being Responded To
[DICOM SCU] >> : (0000,0800), 2, US, "257" - Data Set Type
[DICOM SCU] >> : (0000,0900), 2, US, "45056" - Status
[DICOM SCU] >> : (0000,1000), 60, UI, "1.2.826.0.1.3680043.1.2.119.707110073.20090826.182541.0.0.1" - Affected SOP Instance UID
[DICOM SCU] >> :
[972] >> : Error: store failed for ZAMBOTTI, JENNA, , , , Study: TOE (S) Series: C:\KPacs\Imagebox\1.2.826.0.1.3680043.1.2.119.707110073.20090826.182541.0
\1.2.826.0.1.3680043.1.2.119.707110073.20090826.182541.0.0.1
[972] >> : c-store request for Series C:\KPacs\Imagebox\1.2.826.0.1.3680043.1.2.119.707110073.20090826.182541.0\1.2.826.0.1.3680043.1.2.119.707110073.20090826.182541.0.0.3 initiated.
[DICOM SCU] >> : Connected to: 10.0.15.100:105
[DICOM SCU] >> : Item Type: 2
[DICOM SCU] >> : Association Accept::ReadDynamic
[DICOM SCU] >> : Presentation Context Accept, Transfer Syntax: 1.2.840.10008.1.2.4.70
[DICOM SCU] >> : WNY1C-DVS >> TIAPACS
[DICOM SCU] >> : (0000,0000), 4, UL, "142" - Group Length
[DICOM SCU] >> : (0000,0002), 26, UI, "1.2.840.10008.5.1.4.1.1.1" - Affected SOP Class UID
[DICOM SCU] >> : (0000,0100), 2, US, "1" - Command Field
[DICOM SCU] >> : (0000,0110), 2, US, "3" - Message ID
[DICOM SCU] >> : (0000,0700), 2, US, "0" - Priority
[DICOM SCU] >> : (0000,0800), 2, US, "258" - Data Set Type
[DICOM SCU] >> : (0000,1000), 60, UI, "1.2.826.0.1.3680043.1.2.119.707110073.20090826.182541.0.0.3" - Affected SOP Instance UID
[DICOM SCU] >> :
[DICOM SCU] >> : WNY1C-DVS >> TIAPACS
[DICOM SCU] >> : (0008,0016), 26, UI, "1.2.840.10008.5.1.4.1.1.1" - SOP Class UID
[DICOM SCU] >> : (0008,0018), 60, UI, "1.2.826.0.1.3680043.1.2.119.707110073.20090826.182541.0.0.3" - SOP Instance UID
[DICOM SCU] >> : (0008,0020), 8, DA, "20090826" - Study Date
[DICOM SCU] >> : (0008,0022), 8, DA, "20090826" - Acquisition Date
[DICOM SCU] >> : (0008,0030), 6, TM, "182541" - Study Time
[DICOM SCU] >> : (0008,0032), 6, TM, "182709" - Acquisition Time
[DICOM SCU] >> : (0008,0050), 0, SH, "" - Accession Number
[DICOM SCU] >> : (0008,0060), 2, CS, "CR" - Modality
[DICOM SCU] >> : (0008,0070), 30, LO, "Imaging Dynamics Company Ltd. " - Manufacturer
[DICOM SCU] >> : (0008,0080), 48, LO, "Western New York Immediate Care (NIAGARA FALLS) " - Institution Name
[DICOM SCU] >> : (0008,0090), 6, PN, "WNYIC " - Referring Physician’s Name
[DICOM SCU] >> : (0008,1010), 10, SH, "WNYIC-XRAY" - Station Name
[DICOM SCU] >> : (0008,1030), 8, LO, "TOE (S) " - Study Description
[DICOM SCU] >> : (0008,1070), 6, PN, "^KN^^^" - Operators’ Name
[DICOM SCU] >> : (0008,1090), 2, LO, "X3" - Manufacturer’s Model Name
[DICOM SCU] >> : (0008,2112), -1, SQ
[DICOM SCU] >> : (FFFE,E000), -1, DL
[DICOM SCU] >> : (0008,1150), 26, UI, "1.2.840.10008.5.1.4.1.1.1" - Referenced SOP Class UID
[DICOM SCU] >> : (0008,1155), 60, UI, "1.2.826.0.1.3680043.1.2.119.707110073.20090826.182541.0.0.2" - Referenced SOP Instance UID
[DICOM SCU] >> : (FFFE,E00D), 0, DL, "" - Item Delimitation Item
[DICOM SCU] >> : (FFFE,E0DD), 0, DL, "" - Sequence Delimitation Item
[DICOM SCU] >> : (0010,0010), 18, PN, "ZAMBOTTI^JENNA^^^ " - Patient’s Name
[DICOM SCU] >> : (0010,0020), 8, LO, "W1178990" - Patient ID
[DICOM SCU] >> : (0010,0030), 8, DA, "19681222" - Patient's Birth Date
[DICOM SCU] >> : (0010,0040), 2, CS, "F " - Patient's Sex
[DICOM SCU] >> : (0010,2000), 0, LO, "" - Medical Alerts
[DICOM SCU] >> : (0010,2110), 0, LO, "" - Contrast Allergies
[DICOM SCU] >> : (0018,0015), 4, CS, "TOE " - Body Part Examined
[DICOM SCU] >> : (0018,0060), 2, DS, "60" - KVP
[DICOM SCU] >> : (0018,1000), 12, LO, "D7E07110073 " - Device Serial Number
[DICOM SCU] >> : (0018,1020), 8, LO, "3.3.257 " - Software Version(s)
[DICOM SCU] >> : (0018,1030), 8, LO, "TOE (S) " - Protocol Name
[DICOM SCU] >> : (0018,1150), 2, IS, "20" - Exposure Time
[DICOM SCU] >> : (0018,1151), 4, IS, "100 " - X-ray Tube Current
[DICOM SCU] >> : (0018,1152), 2, IS, "2 " - Exposure
[DICOM SCU] >> : (0018,1153), 4, IS, "2000" - Exposure in uAs
[DICOM SCU] >> : (0018,1164), 12, DS, "0.144\0.144 " - Imager Pixel Spacing
[DICOM SCU] >> : (0018,1405), 4, IS, "0.54" - Relative X-ray Exposure
[DICOM SCU] >> : (0018,5101), 8, CS, "OBLIQUE " - View Position
[DICOM SCU] >> : (0018,6000), 4, DS, "300 " - Sensitivity
[DICOM SCU] >> : (0018,8150), 6, DS, "20000 " - Exposure Time in µS
[DICOM SCU] >> : (0018,8151), 6, DS, "100000" - X-Ray Tube Current in µA
[DICOM SCU] >> : (0020,000D), 56, UI, "1.2.826.0.1.3680043.1.2.119.707110073.20090826.182541.0" - Study Instance UID
[DICOM SCU] >> : (0020,000E), 60, UI, "1.2.826.0.1.3680043.1.2.119.707110073.20090826.182541.0.0.3" - Series Instance UID
[DICOM SCU] >> : (0020,0010), 4, SH, "8948" - Study ID
[DICOM SCU] >> : (0020,0011), 2, IS, "3 " - Series Number
[DICOM SCU] >> : (0020,0013), 6, IS, "54206 " - Instance Number
[DICOM SCU] >> : (0028,0002), 2, US, "1" - Samples per Pixel
[DICOM SCU] >> : (0028,0004), 12, CS, "MONOCHROME2 " - Photometric Interpretation
[DICOM SCU] >> : (0028,0010), 2, US, "1092" - Rows
[DICOM SCU] >> : (0028,0011), 2, US, "752" - Columns
[DICOM SCU] >> : (0028,0030), 12, DS, "0.144\0.144 " - Pixel Spacing
[DICOM SCU] >> : (0028,0100), 2, US, "16" - Bits Allocated
[DICOM SCU] >> : (0028,0101), 2, US, "12" - Bits Stored
[DICOM SCU] >> : (0028,0102), 2, US, "11" - High Bit
[DICOM SCU] >> : (0028,0103), 2, US, "0" - Pixel Representation
[DICOM SCU] >> : (0028,1050), 4, DS, "2039" - Window Center
[DICOM SCU] >> : (0028,1051), 4, DS, "1922" - Window Width
[DICOM SCU] >> : (0028,1052), 2, DS, "0 " - Rescale Intercept
[DICOM SCU] >> : (0028,1053), 2, DS, "1 " - Rescale Slope
[DICOM SCU] >> : (0028,1054), 2, LO, "US" - Rescale Type
[DICOM SCU] >> : (0040,0254), 8, LO, "TOE (S) " - Performed Procedure Step Description
[DICOM SCU] >> : (0040,0275), -1, SQ
[DICOM SCU] >> : (FFFE,E000), -1, DL
[DICOM SCU] >> : (0008,0050), 0, SH, "" - Accession Number
[DICOM SCU] >> : (0020,000D), 56, UI, "1.2.826.0.1.3680043.1.2.119.707110073.20090826.182541.0" - Study Instance UID
[DICOM SCU] >> : (FFFE,E00D), 0, DL, "" - Item Delimitation Item
[DICOM SCU] >> : (FFFE,E0DD), 0, DL, "" - Sequence Delimitation Item
[DICOM SCU] >> : (4321,0010), 30, UN, "Imaging Dynamics Company Ltd. " - *** Unidentified ***
[DICOM SCU] >> : (4321,1041), 2, UN, "N " - *** Unidentified ***
[DICOM SCU] >> : (4321,1042), 2, UN, "´" - *** Unidentified ***
[DICOM SCU] >> : (4321,1064), 2, UN, "HR" - *** Unidentified ***
[DICOM SCU] >> : (7FE0,0010), -1, OW
[DICOM SCU] >> : (FFFE,E000), 0, DL, "" - Item
[DICOM SCU] >> : (FFFE,E000), 733328, DL
[DICOM SCU] >> : (FFFE,E0DD), 0, DL, "" - Sequence Delimitation Item
[DICOM SCU] >> :
[DICOM SCU] >> : WNY1C-DVS << TIAPACS
[DICOM SCU] >> : (0000,0000), 4, UL, "142" - Group Length
[DICOM SCU] >> : (0000,0002), 26, UI, "1.2.840.10008.5.1.4.1.1.1" - Affected SOP Class UID
[DICOM SCU] >> : (0000,0100), 2, US, "32769" - Command Field
[DICOM SCU] >> : (0000,0120), 2, US, "3" - Message ID Being Responded To
[DICOM SCU] >> : (0000,0800), 2, US, "257" - Data Set Type
[DICOM SCU] >> : (0000,0900), 2, US, "45056" - Status
[DICOM SCU] >> : (0000,1000), 60, UI, "1.2.826.0.1.3680043.1.2.119.707110073.20090826.182541.0.0.3" - Affected SOP Instance UID
[DICOM SCU] >> :
[972] >> : Error: store failed for ZAMBOTTI, JENNA, , , , Study: TOE (S) Series: C:\KPacs\Imagebox\1.2.826.0.1.3680043.1.2.119.707110073.20090826.182541.0
\1.2.826.0.1.3680043.1.2.119.707110073.20090826.182541.0.0.3
[972] >> : c-store request for Series C:\KPacs\Imagebox\1.2.826.0.1.3680043.1.2.119.707110073.20090826.182541.0\1.2.826.0.1.3680043.1.2.119.707110073.20090826.182541.0.0.5 initiated.
[DICOM SCU] >> : Connected to: 10.0.15.100:105
[DICOM SCU] >> : Item Type: 2
[DICOM SCU] >> : Association Accept::ReadDynamic
[DICOM SCU] >> : Presentation Context Accept, Transfer Syntax: 1.2.840.10008.1.2.4.70
[DICOM SCU] >> : WNY1C-DVS >> TIAPACS
[DICOM SCU] >> : (0000,0000), 4, UL, "142" - Group Length
[DICOM SCU] >> : (0000,0002), 26, UI, "1.2.840.10008.5.1.4.1.1.1" - Affected SOP Class UID
[DICOM SCU] >> : (0000,0100), 2, US, "1" - Command Field
[DICOM SCU] >> : (0000,0110), 2, US, "3" - Message ID
[DICOM SCU] >> : (0000,0700), 2, US, "0" - Priority
[DICOM SCU] >> : (0000,0800), 2, US, "258" - Data Set Type
[DICOM SCU] >> : (0000,1000), 60, UI, "1.2.826.0.1.3680043.1.2.119.707110073.20090826.182541.0.0.5" - Affected SOP Instance UID
[DICOM SCU] >> :
[DICOM SCU] >> : WNY1C-DVS >> TIAPACS
[DICOM SCU] >> : (0008,0016), 26, UI, "1.2.840.10008.5.1.4.1.1.1" - SOP Class UID
[DICOM SCU] >> : (0008,0018), 60, UI, "1.2.826.0.1.3680043.1.2.119.707110073.20090826.182541.0.0.5" - SOP Instance UID
[DICOM SCU] >> : (0008,0020), 8, DA, "20090826" - Study Date
[DICOM SCU] >> : (0008,0022), 8, DA, "20090826" - Acquisition Date
[DICOM SCU] >> : (0008,0030), 6, TM, "182541" - Study Time
[DICOM SCU] >> : (0008,0032), 6, TM, "182737" - Acquisition Time
[DICOM SCU] >> : (0008,0050), 0, SH, "" - Accession Number
[DICOM SCU] >> : (0008,0060), 2, CS, "CR" - Modality
[DICOM SCU] >> : (0008,0070), 30, LO, "Imaging Dynamics Company Ltd. " - Manufacturer
[DICOM SCU] >> : (0008,0080), 48, LO, "Western New York Immediate Care (NIAGARA FALLS) " - Institution Name
[DICOM SCU] >> : (0008,0090), 6, PN, "WNYIC " - Referring Physician’s Name
[DICOM SCU] >> : (0008,1010), 10, SH, "WNYIC-XRAY" - Station Name
[DICOM SCU] >> : (0008,1030), 8, LO, "TOE (S) " - Study Description
[DICOM SCU] >> : (0008,1070), 6, PN, "^KN^^^" - Operators’ Name
[DICOM SCU] >> : (0008,1090), 2, LO, "X3" - Manufacturer’s Model Name
[DICOM SCU] >> : (0008,2112), -1, SQ
[DICOM SCU] >> : (FFFE,E000), -1, DL
[DICOM SCU] >> : (0008,1150), 26, UI, "1.2.840.10008.5.1.4.1.1.1" - Referenced SOP Class UID
[DICOM SCU] >> : (0008,1155), 60, UI, "1.2.826.0.1.3680043.1.2.119.707110073.20090826.182541.0.0.4" - Referenced SOP Instance UID
[DICOM SCU] >> : (FFFE,E00D), 0, DL, "" - Item Delimitation Item
[DICOM SCU] >> : (FFFE,E0DD), 0, DL, "" - Sequence Delimitation Item
[DICOM SCU] >> : (0010,0010), 18, PN, "ZAMBOTTI^JENNA^^^ " - Patient’s Name
[DICOM SCU] >> : (0010,0020), 8, LO, "W1178990" - Patient ID
[DICOM SCU] >> : (0010,0030), 8, DA, "19681222" - Patient's Birth Date
[DICOM SCU] >> : (0010,0040), 2, CS, "F " - Patient's Sex
[DICOM SCU] >> : (0010,2000), 0, LO, "" - Medical Alerts
[DICOM SCU] >> : (0010,2110), 0, LO, "" - Contrast Allergies
[DICOM SCU] >> : (0018,0015), 4, CS, "TOE " - Body Part Examined
[DICOM SCU] >> : (0018,0060), 2, DS, "60" - KVP
[DICOM SCU] >> : (0018,1000), 12, LO, "D7E07110073 " - Device Serial Number
[DICOM SCU] >> : (0018,1020), 8, LO, "3.3.257 " - Software Version(s)
[DICOM SCU] >> : (0018,1030), 8, LO, "TOE (S) " - Protocol Name
[DICOM SCU] >> : (0018,1150), 2, IS, "20" - Exposure Time
[DICOM SCU] >> : (0018,1151), 4, IS, "100 " - X-ray Tube Current
[DICOM SCU] >> : (0018,1152), 2, IS, "2 " - Exposure
[DICOM SCU] >> : (0018,1153), 4, IS, "2000" - Exposure in uAs
[DICOM SCU] >> : (0018,1164), 12, DS, "0.144\0.144 " - Imager Pixel Spacing
[DICOM SCU] >> : (0018,1405), 6, IS, "-0.64 " - Relative X-ray Exposure
[DICOM SCU] >> : (0018,5101), 8, CS, "LATERAL " - View Position
[DICOM SCU] >> : (0018,6000), 4, DS, "300 " - Sensitivity
[DICOM SCU] >> : (0018,8150), 6, DS, "20000 " - Exposure Time in µS
[DICOM SCU] >> : (0018,8151), 6, DS, "100000" - X-Ray Tube Current in µA
[DICOM SCU] >> : (0020,000D), 56, UI, "1.2.826.0.1.3680043.1.2.119.707110073.20090826.182541.0" - Study Instance UID
[DICOM SCU] >> : (0020,000E), 60, UI, "1.2.826.0.1.3680043.1.2.119.707110073.20090826.182541.0.0.5" - Series Instance UID
[DICOM SCU] >> : (0020,0010), 4, SH, "8948" - Study ID
[DICOM SCU] >> : (0020,0011), 2, IS, "5 " - Series Number
[DICOM SCU] >> : (0020,0013), 6, IS, "54208 " - Instance Number
[DICOM SCU] >> : (0028,0002), 2, US, "1" - Samples per Pixel
[DICOM SCU] >> : (0028,0004), 12, CS, "MONOCHROME2 " - Photometric Interpretation
[DICOM SCU] >> : (0028,0010), 2, US, "748" - Rows
[DICOM SCU] >> : (0028,0011), 2, US, "1104" - Columns
[DICOM SCU] >> : (0028,0030), 12, DS, "0.144\0.144 " - Pixel Spacing
[DICOM SCU] >> : (0028,0100), 2, US, "16" - Bits Allocated
[DICOM SCU] >> : (0028,0101), 2, US, "12" - Bits Stored
[DICOM SCU] >> : (0028,0102), 2, US, "11" - High Bit
[DICOM SCU] >> : (0028,0103), 2, US, "0" - Pixel Representation
[DICOM SCU] >> : (0028,1050), 4, DS, "2020" - Window Center
[DICOM SCU] >> : (0028,1051), 4, DS, "2005" - Window Width
[DICOM SCU] >> : (0028,1052), 2, DS, "0 " - Rescale Intercept
[DICOM SCU] >> : (0028,1053), 2, DS, "1 " - Rescale Slope
[DICOM SCU] >> : (0028,1054), 2, LO, "US" - Rescale Type
[DICOM SCU] >> : (0040,0254), 8, LO, "TOE (S) " - Performed Procedure Step Description
[DICOM SCU] >> : (0040,0275), -1, SQ
[DICOM SCU] >> : (FFFE,E000), -1, DL
[DICOM SCU] >> : (0008,0050), 0, SH, "" - Accession Number
[DICOM SCU] >> : (0020,000D), 56, UI, "1.2.826.0.1.3680043.1.2.119.707110073.20090826.182541.0" - Study Instance UID
[DICOM SCU] >> : (FFFE,E00D), 0, DL, "" - Item Delimitation Item
[DICOM SCU] >> : (FFFE,E0DD), 0, DL, "" - Sequence Delimitation Item
[DICOM SCU] >> : (4321,0010), 30, UN, "Imaging Dynamics Company Ltd. " - *** Unidentified ***
[DICOM SCU] >> : (4321,1041), 2, UN, "N " - *** Unidentified ***
[DICOM SCU] >> : (4321,1042), 2, UN, "" - *** Unidentified ***
[DICOM SCU] >> : (4321,1064), 2, UN, "HR" - *** Unidentified ***
[DICOM SCU] >> : (7FE0,0010), -1, OW
[DICOM SCU] >> : (FFFE,E000), 0, DL, "" - Item
[DICOM SCU] >> : (FFFE,E000), 696126, DL
[DICOM SCU] >> : (FFFE,E0DD), 0, DL, "" - Sequence Delimitation Item
[DICOM SCU] >> :
[DICOM SCU] >> : WNY1C-DVS << TIAPACS
[DICOM SCU] >> : (0000,0000), 4, UL, "142" - Group Length
[DICOM SCU] >> : (0000,0002), 26, UI, "1.2.840.10008.5.1.4.1.1.1" - Affected SOP Class UID
[DICOM SCU] >> : (0000,0100), 2, US, "32769" - Command Field
[DICOM SCU] >> : (0000,0120), 2, US, "3" - Message ID Being Responded To
[DICOM SCU] >> : (0000,0800), 2, US, "257" - Data Set Type
[DICOM SCU] >> : (0000,0900), 2, US, "45056" - Status
[DICOM SCU] >> : (0000,1000), 60, UI, "1.2.826.0.1.3680043.1.2.119.707110073.20090826.182541.0.0.5" - Affected SOP Instance UID
[DICOM SCU] >> :
[972] >> : Error: store failed for ZAMBOTTI, JENNA, , , , Study: TOE (S) Series: C:\KPacs\Imagebox\1.2.826.0.1.3680043.1.2.119.707110073.20090826.182541.0
\1.2.826.0.1.3680043.1.2.119.707110073.20090826.182541.0.0.5
*******************************************************
8/27/2009 9:18:06 AM ThreadID[972] : Closed.
Task processed with errors
*******************************************************