mammo problem

  • I am getting images from a Emajon PACS unit. The mammo films almost always fail. The error message is (debug 4)


    [FLTSIDNEY] 0000,0002 28 UI AffectedSOPClassUID "1.2.840.10008.5.1.4.1.1.1.2"
    [FLTSIDNEY] 0000,0100 2 US CommandField 1
    [FLTSIDNEY] 0000,0110 2 US MessageID 0
    [FLTSIDNEY] 0000,0700 2 US Priority 0
    [FLTSIDNEY] 0000,0800 2 US DataSetType 258
    [FLTSIDNEY] 0000,1000 54 UI AffectedSOPInstanceU "1.2.392.200036.9125.4.0.135221179.879958528.3701491744"
    [FLTSIDNEY] Failed STORAGE
    [FLTSIDNEY] ***Client Error: Unknown Command: 0001**
    [FLTSIDNEY] ***Connection Terminated


    I think this may be a Emajon problem but it is somewhat sporadic, occasionally an image goes through. Does this error message give me any useful information or is it just a nonspecific failure? The mammo films are coming over a T1 line and the sending pacs is not specified in the conquest known DICOM provider list. I get a number of CTs, MRIs, CR images from the same source without a problem.
    Conquest is v14, 32bit, native mySql, no other problems really. I get digital mammos from several other sources, some slower than the T1...


    Any ideas anyone?

  • Hi,


    I have no real clue - I see no obvious path to "FAILED Storage" without another error message - apart from a network error (maybe there is a too short timeout on the sender, MAMMO images are BIG). To exclude a network error on the sending side, you could use, e.g., K-PACS as a reciever and see if that fails as well.


    If it is not a network error, you can maybe try sending the image that fails again to see if it reproduces? And if so, try to send the same image from a conquest workstation to see if that gives the same message?


    After you have tried that I could get you a test version with extra error messages to try and debug it further.


    Marcel