GE Centricity problem

  • Hi,


    We've just installed K-PACS on a Win2k box and are trying to query-retrieve from a GE Centricity server on the local network.


    We can see the studies on the server, but when trying to retrieve, there's a delay and a DICOM error message about the headers.


    Has anyone had problems or otherwise with Centricity?


    It doesn';t seem to be a modality problem, as CR (Agfa) and MR (Siemens) studies gave the same result.


    Thanks,


    William

  • You have to verify that K-PACS AET, IP address and Port number are correctly configured on the Centricity machine for the retrieve operation to work. Query will work without this parameters, Retrieve needs them to know where to send the images.


    Regards,
    Rafael Sanguinetti

  • I am having same problem with a GE AW machine...it simply will not transfer from k-pacs to GE period.
    I tested echo and works fine, we've tried with and without VPN and every time the transfer fails at 1%.


    Here's a log of two attempts to transfer from K-PACS/CONQUEST to GE AW:


    {\rtf1\ansi\ansicpg1252\deff0\deflang1033{\fonttbl{\f0\fnil MS Sans Serif;}{\f1\fnil\fcharset0 MS Sans Serif;}}
    {\colortbl ;\red0\green0\blue0;\red0\green128\blue0;\red0\green0\blue128;\red255\green0\blue0;}
    \viewkind4\uc1\pard\cf1\f0\fs16 *******************************************************
    \par \cf2 5/13/2009 8:23:19 AM ThreadID[700] : Moving images from CTSCAN to AW-RADS started
    \par \cf1 *******************************************************
    \par [700] >> : c-move request for StudyInstanceUID: 1.3.12.2.1107.5.1.4.54984.30000009042718090967100000016 initated
    \par \cf3 [DICOM SCU] >> : Connected to: 192.168.1.11:5678
    \par [DICOM SCU] >> : Item Type: 2
    \par [DICOM SCU] >> : Association Accept::ReadDynamic
    \par [DICOM SCU] >> : Presentation Context Accept, Transfer Syntax: 1.2.840.10008.1.2
    \par [DICOM SCU] >> : VIEW1 >> CTSCAN
    \par [DICOM SCU] >> : (0000,0000), 4,\tab UL, "92" - Group Length
    \par [DICOM SCU] >> : (0000,0002), 28,\tab UI, "1.2.840.10008.5.1.4.1.2.2.2" - Affected SOP Class UID
    \par [DICOM SCU] >> : (0000,0100), 2,\tab US, "33" - Command Field
    \par [DICOM SCU] >> : (0000,0110), 2,\tab US, "3" - Message ID
    \par [DICOM SCU] >> : (0000,0600), 8,\tab AE, "AW-RADS" - Move Destination
    \par [DICOM SCU] >> : (0000,0700), 2,\tab US, "0" - Priority
    \par [DICOM SCU] >> : (0000,0800), 2,\tab US, "258" - Data Set Type
    \par [DICOM SCU] >> :
    \par [DICOM SCU] >> : VIEW1 >> CTSCAN
    \par [DICOM SCU] >> : (0008,0020), 0,\tab DA, "" - Study Date
    \par [DICOM SCU] >> : (0008,0030), 0,\tab TM, "" - Study Time
    \par [DICOM SCU] >> : (0008,0050), 0,\tab SH, "" - Accession Number
    \par [DICOM SCU] >> : (0008,0052), 6,\tab CS, "STUDY " - Query/Retrieve Level
    \par [DICOM SCU] >> : (0008,0061), 0,\tab CS, "" - Modalities in Study
    \par [DICOM SCU] >> : (0008,0080), 0,\tab LO, "" - Institution Name
    \par [DICOM SCU] >> : (0008,0090), 0,\tab PN, "" - Referring Physician\f1 's Name
    \par \f0 [DICOM SCU] >> : (0008,1030), 0,\tab LO, "" - Study Description
    \par [DICOM SCU] >> : (0010,0010), 0,\tab PN, "" - Patient\f1 's Name
    \par \f0 [DICOM SCU] >> : (0010,0020), 0,\tab LO, "" - Patient ID
    \par [DICOM SCU] >> : (0010,0030), 0,\tab DA, "" - Patient's Birth Date
    \par [DICOM SCU] >> : (0010,0040), 0,\tab CS, "" - Patient's Sex
    \par [DICOM SCU] >> : (0010,1010), 0,\tab AS, "" - Patient's Age
    \par [DICOM SCU] >> : (0020,000D), 56,\tab UI, "1.3.12.2.1107.5.1.4.54984.30000009042718090967100000016" - Study Instance UID
    \par [DICOM SCU] >> : (0020,0010), 0,\tab SH, "" - Study ID
    \par [DICOM SCU] >> :
    \par [DICOM SCU] >> : VIEW1 << CTSCAN
    \par [DICOM SCU] >> : (0000,0000), 4,\tab UL, "116" - Group Length
    \par [DICOM SCU] >> : (0000,0002), 28,\tab UI, "1.2.840.10008.5.1.4.1.2.2.2" - Affected SOP Class UID
    \par [DICOM SCU] >> : (0000,0100), 2,\tab US, "32801" - Command Field
    \par [DICOM SCU] >> : (0000,0120), 2,\tab US, "3" - Message ID Being Responded To
    \par [DICOM SCU] >> : (0000,0800), 2,\tab US, "257" - Data Set Type
    \par [DICOM SCU] >> : (0000,0900), 2,\tab US, "49157" - Status
    \par [DICOM SCU] >> : (0000,1020), 2,\tab US, "0" - Number of Remaining Sub-Operations
    \par [DICOM SCU] >> : (0000,1021), 2,\tab US, "0" - Number of Completed Sub-Operations
    \par [DICOM SCU] >> : (0000,1022), 2,\tab US, "0" - Number of Failed Sub-Operations
    \par [DICOM SCU] >> : (0000,1023), 2,\tab US, "0" - Number of Warning Sub-Operations
    \par [DICOM SCU] >> :
    \par \cf4 [700] >> : Error: move failed for DOE, JOHN, Study: Thorax^CHEST_ROUTINE (Adult)
    \par \cf1 [700] >> : c-move request for StudyInstanceUID: 1.3.12.2.1107.5.1.4.54984.30000009042918233234300000010 initated
    \par \cf3 [DICOM SCU] >> : Connected to: 192.168.1.11:5678
    \par [DICOM SCU] >> : Item Type: 2
    \par [DICOM SCU] >> : Association Accept::ReadDynamic
    \par [DICOM SCU] >> : Presentation Context Accept, Transfer Syntax: 1.2.840.10008.1.2
    \par [DICOM SCU] >> : VIEW1 >> CTSCAN
    \par [DICOM SCU] >> : (0000,0000), 4,\tab UL, "92" - Group Length
    \par [DICOM SCU] >> : (0000,0002), 28,\tab UI, "1.2.840.10008.5.1.4.1.2.2.2" - Affected SOP Class UID
    \par [DICOM SCU] >> : (0000,0100), 2,\tab US, "33" - Command Field
    \par [DICOM SCU] >> : (0000,0110), 2,\tab US, "3" - Message ID
    \par [DICOM SCU] >> : (0000,0600), 8,\tab AE, "AW-RADS" - Move Destination
    \par [DICOM SCU] >> : (0000,0700), 2,\tab US, "0" - Priority
    \par [DICOM SCU] >> : (0000,0800), 2,\tab US, "258" - Data Set Type
    \par [DICOM SCU] >> :
    \par [DICOM SCU] >> : VIEW1 >> CTSCAN
    \par [DICOM SCU] >> : (0008,0020), 0,\tab DA, "" - Study Date
    \par [DICOM SCU] >> : (0008,0030), 0,\tab TM, "" - Study Time
    \par [DICOM SCU] >> : (0008,0050), 0,\tab SH, "" - Accession Number
    \par [DICOM SCU] >> : (0008,0052), 6,\tab CS, "STUDY " - Query/Retrieve Level
    \par [DICOM SCU] >> : (0008,0061), 0,\tab CS, "" - Modalities in Study
    \par [DICOM SCU] >> : (0008,0080), 0,\tab LO, "" - Institution Name
    \par [DICOM SCU] >> : (0008,0090), 0,\tab PN, "" - Referring Physician\f1 's Name
    \par \f0 [DICOM SCU] >> : (0008,1030), 0,\tab LO, "" - Study Description
    \par [DICOM SCU] >> : (0010,0010), 0,\tab PN, "" - Patient\f1 's Name
    \par \f0 [DICOM SCU] >> : (0010,0020), 0,\tab LO, "" - Patient ID
    \par [DICOM SCU] >> : (0010,0030), 0,\tab DA, "" - Patient's Birth Date
    \par [DICOM SCU] >> : (0010,0040), 0,\tab CS, "" - Patient's Sex
    \par [DICOM SCU] >> : (0010,1010), 0,\tab AS, "" - Patient's Age
    \par [DICOM SCU] >> : (0020,000D), 56,\tab UI, "1.3.12.2.1107.5.1.4.54984.30000009042918233234300000010" - Study Instance UID
    \par [DICOM SCU] >> : (0020,0010), 0,\tab SH, "" - Study ID
    \par [DICOM SCU] >> :
    \par [DICOM SCU] >> : VIEW1 << CTSCAN
    \par [DICOM SCU] >> : (0000,0000), 4,\tab UL, "116" - Group Length
    \par [DICOM SCU] >> : (0000,0002), 28,\tab UI, "1.2.840.10008.5.1.4.1.2.2.2" - Affected SOP Class UID
    \par [DICOM SCU] >> : (0000,0100), 2,\tab US, "32801" - Command Field
    \par [DICOM SCU] >> : (0000,0120), 2,\tab US, "3" - Message ID Being Responded To
    \par [DICOM SCU] >> : (0000,0800), 2,\tab US, "257" - Data Set Type
    \par [DICOM SCU] >> : (0000,0900), 2,\tab US, "49157" - Status
    \par [DICOM SCU] >> : (0000,1020), 2,\tab US, "0" - Number of Remaining Sub-Operations
    \par [DICOM SCU] >> : (0000,1021), 2,\tab US, "0" - Number of Completed Sub-Operations
    \par [DICOM SCU] >> : (0000,1022), 2,\tab US, "0" - Number of Failed Sub-Operations
    \par [DICOM SCU] >> : (0000,1023), 2,\tab US, "0" - Number of Warning Sub-Operations
    \par [DICOM SCU] >> :
    \par \cf4 [700] >> : Error: move failed for DOE, JANE, Study: Abdomen^ABD_PELVIS_WO_W (Adult)
    \par \cf1 *******************************************************
    \par \cf4 5/13/2009 8:23:26 AM ThreadID[700] : Closed.
    \par Task processed with errors
    \par \cf1 *******************************************************
    \par \cf0
    \par }


    transfer has been successful to other servers but this one simply doesn't want to work.


    any thoughts?

  • Hello Georg,
    thanks for your reply.


    I tried to make the changes suggested in your thread and it still didn't work, the changes actually prevent me from retrieving studies from the conquest server.


    this is what the situation is:


    conquest and k-pacs work fine. if k-pacs wants to send the study to aw-rads (remote location) the transfer fails, first time ever, we've always been able to transfer with different providers.


    k-pacs can retrieve studies from aw-rads (not needed but it's possible) so there is connectivity.


    identical k-pacs installations are able to send to aw-rads from different locations


    dicom transfer from k-pacs has been attempted to internet ip as well as lan ip (via vpn tunnel)...it fails at 1% in both cases


    network testing and packet inspection has been performed, the echos report success but transfer fails.


    efilm is able to send studies


    any thoughts?

Participate now!